  • BioAnalyzer RIN/rRNA ratio disparity

    Hi all,

    I am preparing a few dozen libraries for RNA-sequencing. They all come from stingray embryos in early development. To my knowledge, nobody has extracted RNA from any species within the same family, so we're kind of in uncharted waters. Therefore, when I see something odd in my BioAnalyzer traces, it's hard to know what's normal for these animals. I'm particularly vexed by a consistent disparity between the RIN and rRNA ratios.

    I am extracting RNA using Trizol followed by a column-based purification and DNase I digestion. After each RNA extraction I am diluting the total RNA anywhere from 1:20 - 1:100 (based on Qubit values) and running it on a BioAnalyzer using an RNA Pico kit (FYI: we could only afford one kit, and we went with the Pico instead of the Nano, due to its ability to also detect low levels of mRNA).

    So far I have not had a single sample with a ribosomal ratio > 1.5, but I have also not had a single sample with a RIN < 8.5. In about 50% of the samples, a RIN cannot be calculated because of an 'unexpected ribosomal ratio', which may fall as low as 0.6. When I change the threshold for the ribosomal ratio parameter from 0.7 to 1, thereby forcing a RIN calculation, it is almost always exceptional (see attached traces). Further, all of my samples show similar traces which, besides the low rRNA ratio, show few, if any, other signs of degradation.

    I've read conflicting reports about the reliability of ribosomal ratios and RINs, so I bring my inquiry to you knowledgeable folk in the seqanswers community: do you suspect my low rRNA ratios may just be an artifact of dilution or the suggested 70C for 2 min denaturation right before running (I have heard the 28s rRNA can break down a bit during this step)?

    More importantly, can I trust my BioAnalyzer traces and proceed with RNA-Seq library prep?

    Many thanks for your time,
    John
    Attached Files
    Last edited by JDSwenson; 01-29-2017, 11:50 AM.

  • #2
    28s rRNA breakage caused by heating has been reported in insects and if that was the cause then you would expect it in all your samples. Dilution is also less likely cause because the assigned RIN is independent of sample concentration, instrument and analyst but you may have other species in some of your samples with differing rRNA ratios that affects the overall sample rRNA ratios (I do not know how the embryos have been collected and the possibility of contamination). You can find an example of 28s rRNA breakage upon heating and Bioanalyser run in following paper:

    https://www.ncbi.nlm.nih.gov/pmc/articles/PMC3016993/

    I do not see any signs of RNA degradation and I assume you are going to do polyA selection/priming so it is safe to proceed with library prep because:
    1- If these samples were run on a gel for QC based on gel image I would have proceed with library prep.
    2- The ratios are relatively close in all samples so it does not introduce variability
    3- The ratios can even vary in different tissues of the same species

    Comment


    • #3
      Awesome. Thanks a million, nucacidhunter. I really appreciate the link and the explanation. I am the first on my campus to prepare libraries for RNA-Sequencing, so when I run into a problem, it's helpful to have somebody as kind as yourself who's willing to take the time to help me troubleshoot and explain things I may not have understood.

      So, thank-you.

      Comment


      • #4
        Your are welcome, happy to help.

        Comment

        Latest Articles

        Collapse

        • seqadmin
          Current Approaches to Protein Sequencing
          by seqadmin


          Proteins are often described as the workhorses of the cell, and identifying their sequences is key to understanding their role in biological processes and disease. Currently, the most common technique used to determine protein sequences is mass spectrometry. While still a valuable tool, mass spectrometry faces several limitations and requires a highly experienced scientist familiar with the equipment to operate it. Additionally, other proteomic methods, like affinity assays, are constrained...
          04-04-2024, 04:25 PM
        • seqadmin
          Strategies for Sequencing Challenging Samples
          by seqadmin


          Despite advancements in sequencing platforms and related sample preparation technologies, certain sample types continue to present significant challenges that can compromise sequencing results. Pedro Echave, Senior Manager of the Global Business Segment at Revvity, explained that the success of a sequencing experiment ultimately depends on the amount and integrity of the nucleic acid template (RNA or DNA) obtained from a sample. “The better the quality of the nucleic acid isolated...
          03-22-2024, 06:39 AM

        ad_right_rmr

        Collapse

        News

        Collapse

        Topics Statistics Last Post
        Started by seqadmin, 04-11-2024, 12:08 PM
        0 responses
        31 views
        0 likes
        Last Post seqadmin  
        Started by seqadmin, 04-10-2024, 10:19 PM
        0 responses
        32 views
        0 likes
        Last Post seqadmin  
        Started by seqadmin, 04-10-2024, 09:21 AM
        0 responses
        28 views
        0 likes
        Last Post seqadmin  
        Started by seqadmin, 04-04-2024, 09:00 AM
        0 responses
        53 views
        0 likes
        Last Post seqadmin  
        Working...
        X