    MiSeq 600bp sequencing problem

    Hi, our Genomics Center sent me a link to the MiSeq run they are doing for us with the new 600 bp chemistry - and I notice something on the run stats that scares me:

    1. Yield perfect read 1: 170.8M
    2. Yield perfect read 4 (second 300 bp read): 9.6M!

    The run is still going- will the second read keep going up? It started at 8.9 M for this statistic, and I panicked because in the first attempt to sequence my amplicons, the second read was a lot lower than the first.

    I'm worried because I added indexes and adapters via PCR (regular primer, then adapters+ regular primers, then indexes+adapters in 3 sequential PCRs) and I'm so afraid I screwed that up.

    Last edited by LizBent; 01-29-2014, 05:43 AM. Reason: wrong info for second read
  • TonyBrooks
    Senior Member
    • Jun 2009
    • 303

    #2
    Originally posted by LizBent View Post
    Hi, our Genomics Center sent me a link to the MiSeq run they are doing for us with the new 600 bp chemistry - and I notice something on the run stats that scares me:

    1. Yield perfect read 1: 170.8M
    2. Yield perfect read 4 (second 300 bp read): 9.6M!

    The run is still going- will the second read keep going up? It started at 8.9 M for this statistic, and I panicked because in the first attempt to sequence my amplicons, the second read was a lot lower than the first.

    I'm worried because I added indexes and adapters via PCR (regular primer, then adapters+ regular primers, then indexes+adapters in 3 sequential PCRs) and I'm so afraid I screwed that up.

    The indexes and adapter sequences should be on the same oligo, so if you messed up the indexing PCR, you'd fail to cluster. What's your cluster density?

    Comment

    • LizBent
      Member
      • Jan 2012
      • 31

      #3
      Originally posted by tonybrooks View Post
      the indexes and adapter sequences should be on the same oligo, so if you messed up the indexing pcr, you'd fail to cluster. What's your cluster density?
      1067 +/- 23

      Comment

      • GenoMax
        Senior Member
        • Feb 2008
        • 7142

        #4
        Is this a 2D barcode or 1D barcode run? V2 or V3 chemistry?

        If this is V2 chemistry and these are amplicons then this run is overloaded at 1000+ clusters/mm^2.

        Are you sure you have 170M reads for Read 1?

        Comment

        • TonyBrooks
          Senior Member
          • Jun 2009
          • 303

          #5
          Optimal loading for amplicon on V2 is around 800K/mm2, but it's very library specific.
          We've had libraries run at 1000k/mm2 with no problem at all.

          Either way, it looks like you library is clustering. Next thing to check is the Phas/Prephas. You're looking for values below 0.1% for your reads, indexes are usually higher. Higher values indicate a possible problem with your reagents.

          The % perfect applies only to the PhiX spiked into the run, hence the low yield.

          Comment

          • AllSeq
            Registered Vendor
            • Oct 2013
            • 138

            #6
            Originally posted by GenoMax View Post
            Are you sure you have 170M reads for Read 1?
            Same question: are you sure you have 170M reads? The spec is 25M.
            AllSeq - The Sequencing Marketplace
            [email protected]
            www.AllSeq.com

            Comment

            • TonyBrooks
              Senior Member
              • Jun 2009
              • 303

              #7
              Originally posted by AllSeq View Post
              Same question: are you sure you have 170M reads? The spec is 25M.
              Pretty sure you're looking at the "Yield Perfect" column in BaseSpace, which is the PhiX reads. M = Mb here, not number of reads
              My last run had 12.78M reads pass filter = 3.2Gb for R1 (250bp * 12.78M)
              I has 10% PhiX spiked in, so 320Mb of that 3.2Gb was PhiX
              Yield Perfect was 240M, so I had ~75% %Pefect (in actual fact, it was 75.4% according to BaseSpace)

              Comment

              • LizBent
                Member
                • Jan 2012
                • 31

                #8
                The run statistics are getting better, so I think I was freaking out for no reason... I have 3.4G of data from read 1, and it's halfway through read 2 and there are 1.7G of data, so this run should (?) be fine...

                Sorry to have bothered you all. I really hate the BaseSpace run summary interface!

                Comment

                • GenoMax
                  Senior Member
                  • Feb 2008
                  • 7142

                  #9
                  At times like these having less (or not continuously updated) information actually turns out to be better :-)

                  Comment

                  • microgirl123
                    Senior Member
                    • Jun 2012
                    • 199

                    #10
                    At times like these having less (or not continuously updated) information actually turns out to be better :-)
                    So true. I just started my first V3 run Tuesday so I was watching it closely on Basespace. First up pops the cluster density at 523 +/- 477! Panic inducing! Turns out to be 983, which is perfect. Then clusters passing filter popped up at 43% +/- 52%! Completely useless information but worrisome. Turns out to be 91%, so I'm happy.

                    Comment

                    Latest Articles

                    Collapse

                    • SEQadmin2
                      Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
                      by SEQadmin2



                      Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

                      Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
                      ...
                      10-14-2026, 09:55 AM
                    • SEQadmin2
                      New Genomics Technologies Take Aim at Long-Standing Limits
                      by SEQadmin2


                      Researchers using sequencing and genomics tools often have to make trade-offs. They can choose between speed or scale, short reads or long-range information, or targeted panels or a view of the whole transcriptome. New technologies that have been released this year are built to address those tough choices.

                      We asked six companies the same four questions to learn about their latest products. The new technologies bring a lot to the table, including rethinking sequencing
                      ...
                      09-28-2026, 10:25 AM

                    ad_right_rmr

                    Collapse

                    News

                    Collapse

                    Topics Statistics Last Post
                    Started by SEQadmin2, Yesterday, 11:13 AM
                    0 responses
                    19 views
                    0 reactions
                    Last Post SEQadmin2  
                    Started by SEQadmin2, Yesterday, 10:48 AM
                    0 responses
                    18 views
                    0 reactions
                    Last Post SEQadmin2  
                    Started by SEQadmin2, 10-15-2026, 10:52 AM
                    0 responses
                    36 views
                    0 reactions
                    Last Post SEQadmin2  
                    Started by SEQadmin2, 10-14-2026, 05:39 AM
                    0 responses
                    42 views
                    0 reactions
                    Last Post SEQadmin2  
                    Working...