    QC control

    Hi all
    Does proton has its own supported QC control measurement?
    My reads quality looks very strange (I get this result by FASTQC software which is Illumina-based).
    In Illumina we know that QC= -10*log10(p)>30 is good reads but Proton seems strange.
    Does anybody know that how to measure it in Proton? Thank you!
    Attached Files
    Last edited by super0925; 04-15-2014, 03:34 AM.
  • nbahlis
    Member
    • May 2013
    • 25

    #2
    That is typical with the proton. We are told by lifetech that a22-24 score on the proton equals the 30 on the illumina platform. It has to do apparently withe the proton more conservative score calling.

    Comment

    • super0925
      Senior Member
      • Feb 2014
      • 206

      #3
      Originally posted by nbahlis View Post
      That is typical with the proton. We are told by lifetech that a22-24 score on the proton equals the 30 on the illumina platform. It has to do apparently withe the proton more conservative score calling.
      Thank you!!! But I want to know the reason and why .
      e.g. How do they calculate the QC score? why lower than Illumina (is that different equation?)
      and, how to compare two measurement ?e,g, is that Illumina score -8 =Proton score?
      Last edited by super0925; 04-15-2014, 05:06 AM.

      Comment

      • snetmcom
        Senior Member
        • Oct 2008
        • 159

        #4
        Originally posted by super0925 View Post
        Thank you!!! But I want to know the reason and why .
        e.g. How do they calculate the QC score? why lower than Illumina (is that different equation?)
        and, how to compare two measurement ?e,g, is that Illumina score -8 =Proton score?
        These are raw quality scores, so only the manufacturers know how they are calculated. That is why platform comparisons at this level are not always the best method.

        Comment

        • Brian Bushnell
          Super Moderator
          • Jan 2014
          • 2709

          #5
          Originally posted by super0925 View Post
          Thank you!!! But I want to know the reason and why .
          e.g. How do they calculate the QC score? why lower than Illumina (is that different equation?)
          and, how to compare two measurement ?e,g, is that Illumina score -8 =Proton score?
          If you want to determine the true quality offset, then map the reads and count the mismatch rate of bases at each quality level. I have a tool that can do that from a sam file but it's not really ready for mainstream use yet. I used it to make this plot for some recent Illumina Hiseq 2000 data, which indicates they were under-calling qualities on that run.

          P.S. This only works if the error model is almost entirely substitutions. If IonTorrent reads have indels then it requires a bit more effort. I'm never used IonTorrent data.
          Attached Files
          Last edited by Brian Bushnell; 04-15-2014, 04:07 PM.

          Comment

          • super0925
            Senior Member
            • Feb 2014
            • 206

            #6
            Originally posted by snetmcom View Post
            These are raw quality scores, so only the manufacturers know how they are calculated. That is why platform comparisons at this level are not always the best method.
            Thank you!
            I agree with you. It is hard to say which machine (HiSeq/MiSeq/Proton) is better from QC score but I want to know the equivalent quality score in Proton .
            Like nbahlis said
            Illumina score 30 corresponding to Proton QC score 22
            So how about other score e.g. 20, 40 ?

            Comment

            • mastal
              Senior Member
              • Mar 2009
              • 666

              #7
              I think the error model for the IonProton would be more like that for 454, and involve homopolymer indels rather than substitutions.

              Comment

              • GenoMax
                Senior Member
                • Feb 2008
                • 7142

                #8
                @super0925: There is a technical note available on the ioncommunity site that describes six Quality Score predictors (and some additional ones that are apparently not listed) used for ion PGM. I did not see one specifically for proton but the one for PGM may work.

                You should be able to create a free account on ioncommunity site and search for that document.

                Comment

                • super0925
                  Senior Member
                  • Feb 2014
                  • 206

                  #9
                  Originally posted by GenoMax View Post
                  @super0925: There is a technical note available on the ioncommunity site that describes six Quality Score predictors (and some additional ones that are apparently not listed) used for ion PGM. I did not see one specifically for proton but the one for PGM may work.

                  You should be able to create a free account on ioncommunity site and search for that document.
                  Thank you very much. I have seen that before.
                  The Quality Score in Proton is based on the 6 predictors which are different from Illumina. Am I right?
                  But I am still confused that the relationship between Proton Q score and Illumina Q score, like Q=22 Proton related to Q=30 MiSeq.
                  So far I could only directly judge that OK Q>22 is good reads and Q<22 is required for trimming...... Am I right?

                  Comment

                  • GenoMax
                    Senior Member
                    • Feb 2008
                    • 7142

                    #10
                    I do not think it would be possible to equate Q-scores from Illumina/Ion by a formula because they are not using the same predictors.

                    Some recent studies seem to suggest that trimming may not be needed unless adapters are present or the raw qualities are poor. I would suggest doing some analysis without trimming to see what happens. BTW, What kind of analysis are you doing (re-seq, RNA-seq etc)?

                    Comment

                    • super0925
                      Senior Member
                      • Feb 2014
                      • 206

                      #11
                      Originally posted by GenoMax View Post
                      I do not think it would be possible to equate Q-scores from Illumina/Ion by a formula because they are not using the same predictors.

                      Some recent studies seem to suggest that trimming may not be needed unless adapters are present or the raw qualities are poor. I would suggest doing some analysis without trimming to see what happens. BTW, What kind of analysis are you doing (re-seq, RNA-seq etc)?
                      My work is RNA-seq.
                      So the trimming step sould be 'gentle' or even not which is suggested by some experts in SeqAnswers and Biostars. Am I right?

                      Comment

                      • GenoMax
                        Senior Member
                        • Feb 2008
                        • 7142

                        #12
                        Originally posted by super0925 View Post
                        My work is RNA-seq.
                        So the trimming step sould be 'gentle' or even not which is suggested by some experts in SeqAnswers and Biostars. Am I right?
                        Basically yes. Need for trimming can be evaluated based on the results you get after the first round of analysis.

                        Comment

                        • Brian Bushnell
                          Super Moderator
                          • Jan 2014
                          • 2709

                          #13
                          I uploaded a new version of BBMap that can tell you the observed quality for claimed qualities, assuming you have a reference. You run it like this:

                          bbmap.sh ref=x.fasta in=reads.fastq nodisk qahist=qahist.txt

                          This will give 7 columns:
                          quality, match, sub, insertion, deletion, observed quality, observed quality (subs only)

                          Since deletions occur between bases, I add deletion events to bases neighboring the deletion. That would over-represent deletions, so all the other columns are multiplied by 2 to compensate. Anyway, if you plot the first column against the 6th column, it will tell you how the observed error rates correlate with quality scores.

                          You can get QC-relevant additional histograms with these flags:

                          bhist=bhist.txt (base composition by read position)
                          qhist=qhist.txt (average quality by read position)
                          mhist=mhist.txt (match, substitution, insertion, deletion events by read position)
                          ihist=ihist.txt (insert size distribution)

                          ...and if you want the actual mapped reads, just add "out=mapped.sam".

                          Comment

                          • Zapages
                            Member
                            • Oct 2012
                            • 98

                            #14
                            ^Is it possible to do quality control without a reference with bbmap?

                            I have some Ion Xpress read data. I have been told the quality control has been done, but when just checked through FastQC, it seems really bad at the end. But then again its my first time working with Ion Torrent/PGM data.

                            Thank you in advance for the help.
                            Attached Files

                            Comment

                            • Brian Bushnell
                              Super Moderator
                              • Jan 2014
                              • 2709

                              #15
                              BBMap requires a reference; all of those histograms (other than bhist) are based on mapping. But it doesn't have to be a good assembly. If you have sufficient coverage, a quick draft assembly is adequate for mapping to generate that QC data.

                              Comment

                              Latest Articles

                              Collapse

                              • SEQadmin2
                                Advanced Sequencing Platforms Tackle Neuroscience’s Toughest Genomics Problems
                                by SEQadmin2



                                Genomics studies in neuroscience face a special challenge due to the brain’s complexity and scarcity of samples. Mapping changes in cell type and state using conventional next-generation sequencing methods remains challenging. Advances in technologies like single-cell sequencing, spatial transcriptomics, and long-read sequencing have opened the door to deeper studies of the brain and diseases like Alzheimer’s, amyotrophic lateral sclerosis (ALS), and schizophrenia.
                                ...
                                07-09-2026, 11:10 AM
                              • SEQadmin2
                                Cancer Drug Resistance: The Lingering Barrier to Rising Survival
                                by SEQadmin2



                                Cancer survival rates have significantly increased in the last few decades in the United States, reaching a combined 70% 5-year survival rate by 2021. Behind this number, there are years of research to find new therapies, drug targets, and early detection methods. But there is one core challenge that keeps slowing down these advances, and it’s about drug resistance.

                                There is no single reason why many patients don’t respond to treatment as expected. Cancer is...
                                07-08-2026, 05:17 AM
                              • GATTACAT
                                Reply to Nine Things a Sample Prep Scientist Thinks About Before Sequencing
                                by GATTACAT
                                Love this - good data definitely starts from good input, and poor input can only give relatively poor data. I particularly like the mention of Nanodrop/absorbance based methods for quantification. It's such a toss up if you'll get an accurate reading or what amounts to a randomly generated number, and a lot of library/sequencing related issues can be traced back to poor quant.
                                07-01-2026, 11:43 AM

                              ad_right_rmr

                              Collapse

                              News

                              Collapse

                              Topics Statistics Last Post
                              Started by SEQadmin2, 07-13-2026, 10:26 AM
                              0 responses
                              27 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 07-09-2026, 10:04 AM
                              0 responses
                              37 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 07-08-2026, 10:08 AM
                              0 responses
                              24 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 07-07-2026, 11:05 AM
                              0 responses
                              34 views
                              0 reactions
                              Last Post SEQadmin2  
                              Working...