    Inconsitencies with qPCR results

    I've begun quantifying all libraries with qPCR (Kapa kit) but I'm getting massive inconsistencies with the number of reads I get from samples in a multiplex pool.
    I dilute each final library 1:1000000 (three 1:100 serial dilution) mixing at least 10 times by pipette after each dilution. This then puts me somewhere in the middle of the Kapa standard curve. Everything is run in triplicate and the R2 of the standard curve is always great >0.99.
    I then normalise, pool and requantify by both Qubit & qPCR and generally get a pooled concentration of 9-11nM (what I might expect). Cluster density for the overall run is fairly good, but after indexing there's massive variation sample-to-sample (5-fold differences for my last set of TruSeq RNA libraries).
    I've also Qubit'd and Bioanalysed each library (I use the average size between 200-700bp to normalise my Kapa qPCR). There are no dimers or evidence of over-amplification. The Qubit concetrations are slightly better than qPCR but still not great (R2 between qubit and cluster# is 0.52, but for qPCR it was 0.41)
    Does anyone have a fool-proof method for qPCR? I was wondering if it's worth running mulitple dilutions of the same sample on the PCR plate, but I would quickly run out of space on my 96 well plate (I often have to do 24 libraries at a time).
    Has anyone tried a homebrew probe assay?
    Attached Files
  • microgirl123
    Senior Member
    • Jun 2012
    • 199

    #2
    Strange, I pretty much use your approach for our gDNA libraries and don't see such huge inconsistencies in number of reads.

    Are your samples PCR-enriched after library prep? I usually make PCR-free libraries and have found that the structure of the adapters can make the samples run strangely in the Bioanalyzer, resulting in incorrect sizing which affects the KAPA math.

    Are all your pooled libraries approximately the same size?

    Pipetting error?

    The other thing that I've been doing (I'm running a MiSeq) is denaturing the samples with NaOH and then denaturing them again at 96 degrees. Somewhere there's an Illumina bulletin that explains how different adapters/indices denature less effectively than others.

    Comment

    • nucacidhunter
      Jafar Jabbari
      • Jan 2013
      • 1250

      #3
      We use concentration from TapeStation or bioanalyser and normalise to 6 nM. Then dilute it 10000 times in two steps for QPCR using KAPA kit. Our pooled samples gives very close figures specially when they are the same type of library with similar trace.

      Comment

      • TonyBrooks
        Senior Member
        • Jun 2009
        • 303

        #4
        Originally posted by nucacidhunter View Post
        We use concentration from TapeStation or bioanalyser and normalise to 6 nM. Then dilute it 10000 times in two steps for QPCR using KAPA kit. Our pooled samples gives very close figures specially when they are the same type of library with similar trace.
        I was thinking about trying a double normalisation too. First using the Bioanalyser and next by qPCR. Might try that for the next run.
        We'll also work on automating qPCR set up to avoid any pipettting error.

        Comment

        Latest Articles

        Collapse

        • GATTACAT
          Reply to Nine Things a Sample Prep Scientist Thinks About Before Sequencing
          by GATTACAT
          Love this - good data definitely starts from good input, and poor input can only give relatively poor data. I particularly like the mention of Nanodrop/absorbance based methods for quantification. It's such a toss up if you'll get an accurate reading or what amounts to a randomly generated number, and a lot of library/sequencing related issues can be traced back to poor quant.
          07-01-2026, 11:43 AM
        • SEQadmin2
          Nine Things a Sample Prep Scientist Thinks About Before Sequencing
          by SEQadmin2


          I’m not a sequencing expert. I’m a purification scientist who uses NGS to evaluate workflows my group develops. With this perspective, we think about the sample first and the NGS workflow second. The sequencer is an exceptionally honest reporter, but it can only report on what you give it, so whether you get clean, interpretable data from an NGS workflow is largely determined before you begin.

          Here are nine questions we think about, in roughly the order they matter, before...
          06-18-2026, 07:11 AM

        ad_right_rmr

        Collapse

        News

        Collapse

        Topics Statistics Last Post
        Started by SEQadmin2, 07-02-2026, 11:08 AM
        0 responses
        12 views
        0 reactions
        Last Post SEQadmin2  
        Started by SEQadmin2, 06-30-2026, 05:37 AM
        0 responses
        14 views
        0 reactions
        Last Post SEQadmin2  
        Started by SEQadmin2, 06-26-2026, 11:10 AM
        0 responses
        20 views
        0 reactions
        Last Post SEQadmin2  
        Started by SEQadmin2, 06-17-2026, 06:09 AM
        0 responses
        54 views
        0 reactions
        Last Post SEQadmin2  
        Working...