    Originally posted by nucacidhunter View Post
    I wonder if you are using custom sequencing primers. Your read2 sequencing primer might not be optimal in addition to other reasons given previously. I think you need to provide more information about the library and sequencing metrics to enable others to help you effectively.
    i am new in this field. can anyone tell me how to check the cluster density. just in case this is casued by amplicon diversity issue, is there any trick to aviod this problem happens again. thank you guys so much!

    I will send more sequencing metrics.

    Comment

    • dalin
      Member
      • Sep 2014
      • 18

      #17
      Read1

      [IMG][/IMG]

      Comment

      • dalin
        Member
        • Sep 2014
        • 18

        #18

        Comment

        • dalin
          Member
          • Sep 2014
          • 18

          #19
          read2

          Comment

          • dalin
            Member
            • Sep 2014
            • 18

            #20
            Originally posted by nucacidhunter View Post
            I wonder if you are using custom sequencing primers. Your read2 sequencing primer might not be optimal in addition to other reasons given previously. I think you need to provide more information about the library and sequencing metrics to enable others to help you effectively.
            in this library,i use the custom sequencing primers
            P5 primer:AATGATACGGCGACCACCGAGATCTACACTCTTTCCCTACACGACGCTCTTCCGATCT
            P7 primer:CAAGCAGAAGACGGCATACGAGATCGTGATGTGACTGGAGTTCAGACGTGTGCTCTTCCGATC

            Comment

            • weigrc
              Member
              • Oct 2011
              • 46

              #21
              Originally posted by dalin View Post
              thanks for your reply, which explained a lot. Yes I am using amplicon. The worse is that my library has only about 20bp diverse sequence in 5 and 3 end. the sequence in between is a linker, means there is no diverse at all. i cannot change my library. but hope there will be a trick to circle this problem.

              After these FastQC reports and your explanation, the issue should not be generated by over-clustering.

              As nucacidhunter said, you may give more info for trouble shoot, like listed below:

              (1) What/which protocol/kit for library construction?
              (2) Adapter system you used? e.g. Nextera or TruSeq LT, if you used custom primers / adaptors, please give intact sequences of them, cannot recognize "the sequencing primers" you provided.
              (3) Sequencing run type you used? eg. V3 high output run of HiSeq PE100.
              (4) % of PhiX spike-in?

              Comment

              • Brian Bushnell
                Super Moderator
                • Jan 2014
                • 2709

                #22
                I wait with bated breath. This is the most terrible run I have ever seen! It looks to me like it ran out of reagent after 100 cycles, but I'm betting on nucacidhunter for diagnosing it correctly.

                Comment

                • dalin
                  Member
                  • Sep 2014
                  • 18

                  #23
                  Originally posted by Brian Bushnell View Post
                  I wait with bated breath. This is the most terrible run I have ever seen! It looks to me like it ran out of reagent after 100 cycles, but I'm betting on nucacidhunter for diagnosing it correctly.
                  Thanks for reply, In this library,i only PCR amplification for about 19 cycles......

                  Comment

                  • dalin
                    Member
                    • Sep 2014
                    • 18

                    #24
                    Originally posted by weigrc View Post
                    After these FastQC reports and your explanation, the issue should not be generated by over-clustering.

                    As nucacidhunter said, you may give more info for trouble shoot, like listed below:

                    (1) What/which protocol/kit for library construction?
                    (2) Adapter system you used? e.g. Nextera or TruSeq LT, if you used custom primers / adaptors, please give intact sequences of them, cannot recognize "the sequencing primers" you provided.
                    (3) Sequencing run type you used? eg. V3 high output run of HiSeq PE100.
                    (4) % of PhiX spike-in?
                    in this library,i use the custom sequencing primers
                    P5 primer:AATGATACGGCGACCACCGAGATCTACACTCTTTCCCTACACGACGCTCTTCCGATCT
                    P7 primer:CAAGCAGAAGACGGCATACGAGATCGTGATGTGACTGGAGTTCAGACGTGTGCTCTTCCGATC

                    With the custom sequencing primers, the library sequence is
                    AATGATACGGCGACCACCGAGATCTACACTCTTTCCCTACACGACGCTCTTCCGATCT--NNNNNNNNNNNNNNNNNNNN------GTCGGAGAATTCCTTACTAGTAGAACTCTGTTCTTGAGCTAGCATCGATGCTAGCTCAAGAACAGAGTTCTACTAGTAAGGAATTCTCCGAC----NNNNNNNNNNNNNNNNNNNN---GATCGGAAGAGCACACGTCTGAACTCCAGTCACATCACGATCTCGTATGCCGTCTTCTGCTTG
                    thanks a lot!

                    Comment

                    • nucacidhunter
                      Jafar Jabbari
                      • Jan 2013
                      • 1250

                      #25
                      From provided information I assume that your library insert size is around 130 bp with middle 90 bases being the same in all inserts and flanking 5' and 3' 20 bases divergent. The library structure contains truncated universal LT P5 and index1 P7 adapter with additional bases. Considering the position of P5 deletion and P7 addition this library should be suitable for sequencing with standard MiSeq reagents. Read1 sequence quality drops below 30 around 60th cycle which is below expected for 100 cycle sequencing. I would suggest that read2 low quality has been caused either by a fault in sequencer or MiSeq reagents. Reagents could have been faulty or used past expiry date or recommended handling has not been followed.

                      You have not mentioned the library prep method, but the data suggests that mid sequences of insert are not exactly the same.

                      Comment

                      • dalin
                        Member
                        • Sep 2014
                        • 18

                        #26
                        Originally posted by nucacidhunter View Post
                        From provided information I assume that your library insert size is around 130 bp with middle 90 bases being the same in all inserts and flanking 5' and 3' 20 bases divergent. The library structure contains truncated universal LT P5 and index1 P7 adapter with additional bases. Considering the position of P5 deletion and P7 addition this library should be suitable for sequencing with standard MiSeq reagents. Read1 sequence quality drops below 30 around 60th cycle which is below expected for 100 cycle sequencing. I would suggest that read2 low quality has been caused either by a fault in sequencer or MiSeq reagents. Reagents could have been faulty or used past expiry date or recommended handling has not been followed.

                        You have not mentioned the library prep method, but the data suggests that mid sequences of insert are not exactly the same.
                        Thank you very much,the 5 and 3 flanking divergent sequence are accually 19-21 bp(probally mainly 20),we used hiseq 2000. other lines in this run looks fine. so it may not likely caused by reagent.
                        you suggest help me a lot, thank you!

                        Comment

                        • ashleylose07
                          Junior Member
                          • Aug 2021
                          • 1

                          #27
                          I'm sorry i don't know the cluster density of the library. The Sequencing company have no information for the cluster density. teatv spice money login
                          Last edited by ashleylose07; 08-21-2021, 02:11 AM.

                          Comment

                          Latest Articles

                          Collapse

                          • SEQadmin2
                            From Collection to Sequencing: Why Sample Preparation and Preservation Define Sequencing Data
                            by SEQadmin2


                            Data variability is still an issue in sequencing technologies despite the advances in reproducibility and accuracy of these platforms. But the problem does not originate in the sequencing itself, but in the previous steps, before the sample reaches the sequencer.


                            The first step is collection, followed by preservation and sample preparation for analysis. Most scientists overlook those steps, but not being careful might just be skewing the experiment’s results.
                            ...
                            06-02-2026, 10:05 AM
                          • SEQadmin2
                            Single-Cell Sequencing at an Inflection Point: Early Impacts of New Platforms and Emerging Trends
                            by SEQadmin2


                            With the launch of new single-cell sequencing platforms in 2026, the field stands at an exciting inflection point. This article surveys the most impactful advances in the field and discusses how they’re reshaping research in cancer, immunology, and beyond.


                            Introduction

                            Single-cell sequencing technologies have undergone remarkable advances over the past decade, transitioning from low-throughput experimental approaches to highly scalable platforms capable of...
                            05-22-2026, 06:42 AM
                          • SEQadmin2
                            Environmental Genomics in the Age of NGS: From Microbes to Conservation Strategies
                            by SEQadmin2

                            Studying ecosystems means dealing with complex, multi-species communities that are hard to observe at scale. This complexity, however, hides many important questions to be answered, from how biogeochemical cycles work and how climate change can affect species distribution to how conservation strategies can work best.


                            Genomics, particularly since the expansion of NGS, has transformed ecosystem ecology. By sequencing environmental DNA, we can now assess biodiversity without direct...
                            05-06-2026, 09:04 AM

                          ad_right_rmr

                          Collapse

                          News

                          Collapse

                          Topics Statistics Last Post
                          Started by SEQadmin2, Today, 08:59 AM
                          0 responses
                          11 views
                          0 reactions
                          Last Post SEQadmin2  
                          Started by SEQadmin2, 06-02-2026, 12:03 PM
                          0 responses
                          21 views
                          0 reactions
                          Last Post SEQadmin2  
                          Started by SEQadmin2, 06-02-2026, 11:40 AM
                          0 responses
                          17 views
                          0 reactions
                          Last Post SEQadmin2  
                          Started by SEQadmin2, 05-28-2026, 11:40 AM
                          0 responses
                          31 views
                          0 reactions
                          Last Post SEQadmin2  
                          Working...