Unconfigured Ad

Collapse
X
 
  • Time
  • Show
Clear All
new posts
  • MikeNice
    Junior Member
    • Mar 2010
    • 6

    #1

    HiSeq 2500 v4 chemistry Read 1 quality drop

    We have been seeing a very specific fail mode in about 1 in 10 runs since upgrading to v4 chemistry last year, and we wonder if anyone else is experiencing this or has thoughts to the cause.

    When this happens:
    • the Q30 starts dropping faster than normal, always at about cycle 50,
    • only the center lanes are affected, 3-6 being most severe, 2 and 7 less so, and 1 and 8 never affected,
    • always total recovery in the Index and Read 2, with normal Q30 for those reads
    • accompanied by high pre-phasing.


    Illumina suspects it's the "blue bubble" phenomenon that many users see since the upgrade, which is caused by the peltier being at a slightly higher temperature during the reaction (54 degrees instead of 52.5). But that should cause the "blue bubble" pattern on the intensity chart - higher intensity at the bottom of FC due to the cold reagent entering, and then lower intensity as it passes through. We're not seeing that here. Also, our engineer has made temperature adjustments to no avail.

    Perhaps v4 chemistry is more sensitive and less robust than previous versions. Has anyone else experienced this?

    Thanks!
  • nucacidhunter
    Jafar Jabbari
    • Jan 2013
    • 1250

    #2
    I have not seen anomaly like this. Just out of interest, are there any differences among good and bad lanes in cluster numbers? Could it be related maybe to a fault in C-Bot?

    Comment

    • MikeNice
      Junior Member
      • Mar 2010
      • 6

      #3
      Thanks for the reply, nuc. There is no difference between lanes on any early metrics (cluster density, PF, intensities, etc.). We've seen this problem even when we load the same sample in all 8 lanes at the same concentration. We'll have consistent density and PF, then at about cycle 50, the middle lanes start dropping Q30 and pre-phasing jumps. Fluidics always passes tests before and after runs, too.

      Comment

      • Jessica D.
        Junior Member
        • Feb 2015
        • 7

        #4
        I haven't seen the particular issue you're talking about, either.

        None of the usual suspects (solenoid, pump) I'd look at seem to be making sense, considering your fluidics tests pass. Lane dependent issues just always make me think hardware.

        Can you share some images off SAV? The Q30 chart would help, but also the intensity heat maps off the flow cell for one or two cycles before the issue through to about five cycles after when it more or less remains steadily bad?

        Comment

        • AdriH
          Junior Member
          • Oct 2015
          • 1

          #5
          Did you ever resolve your issue? We have seen similar situations with high prefacing in the middle lanes, where lane 3 and 6 are most significant. Also, when looking at the tile images, it seems as if the clusters are becoming more static with each cycle, as if the the next base fails to incorporate. This has only been evident with the HiSeq 2500 v4 run. Also, it seems that a rehyb will resolve the issue, but this is not a permanent fix. If you had a resolution or any feed back it would be much appreciated.

          Comment

          • MikeNice
            Junior Member
            • Mar 2010
            • 6

            #6
            We never fully confirmed the cause of this, however it was resolved on all but one instrument by having the FSE replace the front manifold. The last instrument was temporarily fixed with a new front manifold, but the problem returned again and again, until we went further upstream and cleaned the VICI selector valve. We haven't seen this issue since. Good luck!

            Comment

            Latest Articles

            Collapse

            • SEQadmin2
              Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
              by SEQadmin2



              Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

              Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
              ...
              10-14-2026, 09:55 AM
            • SEQadmin2
              New Genomics Technologies Take Aim at Long-Standing Limits
              by SEQadmin2


              Researchers using sequencing and genomics tools often have to make trade-offs. They can choose between speed or scale, short reads or long-range information, or targeted panels or a view of the whole transcriptome. New technologies that have been released this year are built to address those tough choices.

              We asked six companies the same four questions to learn about their latest products. The new technologies bring a lot to the table, including rethinking sequencing
              ...
              09-28-2026, 10:25 AM

            ad_right_rmr

            Collapse

            News

            Collapse

            Topics Statistics Last Post
            Started by SEQadmin2, 10-22-2026, 11:13 AM
            0 responses
            23 views
            0 reactions
            Last Post SEQadmin2  
            Started by SEQadmin2, 10-22-2026, 10:48 AM
            0 responses
            20 views
            0 reactions
            Last Post SEQadmin2  
            Started by SEQadmin2, 10-15-2026, 10:52 AM
            0 responses
            36 views
            0 reactions
            Last Post SEQadmin2  
            Started by SEQadmin2, 10-14-2026, 05:39 AM
            0 responses
            42 views
            0 reactions
            Last Post SEQadmin2  
            Working...