    Can we stop the run and re-generated the cluster on flow cell on MiSeq?

    We are using the MiSeq machine (2*250, V2 kit) for a Truseq library sequencing. As we find the cluster density is low (only 239k) on the cycle 3, can we just stop the machine now and add additional libraries and re-run the protocol?

    Because the MiSeq machine used to report an error just after the cluster generation step. We washed and restarted the machine, and all process worked fine. I just wondering if I can do that again on cycle 3.

    add:
    The run continue and yield 2.7G result at last. The normal run with density 800K should yield 7.5G data.

    add:
    We quantified the libraries with real-time PCR using KAPA kit.
    Last edited by runsheng; 04-22-2015, 05:45 PM. Reason: add result
  • microgirl123
    Senior Member
    • Jun 2012
    • 199

    #2
    As far as I know, you cannot stop and restart the MiSeq. You should try to figure out why your cluster density is so low. How did you quantify your libraries before loading them on the MiSeq?

    Comment

    • runsheng
      Junior Member
      • Jul 2014
      • 4

      #3
      Originally posted by microgirl123 View Post
      As far as I know, you cannot stop and restart the MiSeq. You should try to figure out why your cluster density is so low. How did you quantify your libraries before loading them on the MiSeq?
      We quantified the libraries with real-time PCR using KAPA kit. It works well for the RNA-seq libraries with insertion size ~500bp. As for the DNA library (insertion size ~800), the concentration should be lower than the calculated one, adjusted as 5/8.

      Comment

      • microgirl123
        Senior Member
        • Jun 2012
        • 199

        #4
        Interesting. Have you contacted Illumina tech support? The last time I saw such low cluster densities on a properly-quantified run, there was a problem with the optics.

        Comment

        • Jessica_L
          Senior Member
          • Feb 2010
          • 117

          #5
          microgirl is right-- you can't just switch the instrument off and on and re-cluster a flow cell.

          The end of the cluster gen process includes a 3' blocking step that prevents extension of DNA template and flow cell oligos (which would interfere with bridge amplifying any new libraries you might add in an effort to boost cluster counts). The only work around I can think of would be to run a deblock step similar to what happens during the PE turnaround but I don't know how you would do that manually. You would still have to worry about having enough reagents to repeat clustering from the same reagent cartridge, though.

          Comment

          • runsheng
            Junior Member
            • Jul 2014
            • 4

            #6
            Originally posted by microgirl123 View Post
            Interesting. Have you contacted Illumina tech support? The last time I saw such low cluster densities on a properly-quantified run, there was a problem with the optics.
            I guess the machine should be all right cause we just got a normal run for RNAseq one week ago.

            We have analysed the data generated from this run. Though with a low density and yield, the read seems to be fine and even get a little bit higher Q30 percentage than our former runs (with >800 cluster density). The quality, read length, read yield for each index and read mappbility all seem to be OK.

            The only flaw is the read coverage, we expected a 8X coverage but now we only get 3X.
            And that's why I asked if I can re-do the cluster generation process.

            Comment

            • runsheng
              Junior Member
              • Jul 2014
              • 4

              #7
              Originally posted by Jessica_L View Post
              microgirl is right-- you can't just switch the instrument off and on and re-cluster a flow cell.

              The end of the cluster gen process includes a 3' blocking step that prevents extension of DNA template and flow cell oligos (which would interfere with bridge amplifying any new libraries you might add in an effort to boost cluster counts). The only work around I can think of would be to run a deblock step similar to what happens during the PE turnaround but I don't know how you would do that manually. You would still have to worry about having enough reagents to repeat clustering from the same reagent cartridge, though.
              Thank you for reply!
              Indeed we have re-generated the cluster once. At that time, we encountered an error after the cluster generation step and the machine was halted on cycle 1. After wash and restart, the run completed normally and yielded full data.

              The 3' blocking is still a ddNTP blocking for MiSeq?

              Comment

              • HeinKey
                Member
                • May 2009
                • 21

                #8
                hi Runsheng,
                I understood you compared an 800 nt library qPCR result with a 500 nt library, and corrected loading with 5/8.
                This seems logical given both libraries cluster equally efficient.
                This is not the case. Shorter fragments cluster far more efficient.
                That is probably why you got a lower cluster density than expected.

                Comment

                Latest Articles

                Collapse

                • SEQadmin2
                  Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
                  by SEQadmin2



                  Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

                  Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
                  ...
                  10-14-2026, 09:55 AM
                • SEQadmin2
                  New Genomics Technologies Take Aim at Long-Standing Limits
                  by SEQadmin2


                  Researchers using sequencing and genomics tools often have to make trade-offs. They can choose between speed or scale, short reads or long-range information, or targeted panels or a view of the whole transcriptome. New technologies that have been released this year are built to address those tough choices.

                  We asked six companies the same four questions to learn about their latest products. The new technologies bring a lot to the table, including rethinking sequencing
                  ...
                  09-28-2026, 10:25 AM

                ad_right_rmr

                Collapse

                News

                Collapse

                Topics Statistics Last Post
                Started by SEQadmin2, 10-22-2026, 11:13 AM
                0 responses
                26 views
                0 reactions
                Last Post SEQadmin2  
                Started by SEQadmin2, 10-22-2026, 10:48 AM
                0 responses
                20 views
                0 reactions
                Last Post SEQadmin2  
                Started by SEQadmin2, 10-15-2026, 10:52 AM
                0 responses
                36 views
                0 reactions
                Last Post SEQadmin2  
                Started by SEQadmin2, 10-14-2026, 05:39 AM
                0 responses
                43 views
                0 reactions
                Last Post SEQadmin2  
                Working...