  • TonyBrooks
    Senior Member
    • Jun 2009
    • 303

    #61
    This upgrade really is a pain in the backside. I have no idea why Illumina couldn't make this back compatible and then use the RFID chips to determine whether to process the run as required.

    I wonder what happens if someone comes wanting to compare to an older data set once we've switched to v2. They'll need to pay to re-run those old libraries on v2.

    Comment

    • SA2002
      Junior Member
      • Apr 2015
      • 2

      #62
      Hello,

      Has anyone used v2 reagents with NCS v1.3? If so, what did the results look like?

      Comment

      • TonyBrooks
        Senior Member
        • Jun 2009
        • 303

        #63
        Originally posted by SA2002 View Post
        Hello,

        Has anyone used v2 reagents with NCS v1.3? If so, what did the results look like?
        I'm pretty sure you can't do this as the software would surely check which kit version you are using and tell you it's incompatible.
        It also looks like they changed the dyes, so the lasers would most likely excite at the wrong wavelength.

        Comment

        • GenoMax
          Senior Member
          • Feb 2008
          • 7142

          #64
          Originally posted by SA2002 View Post
          Hello,

          Has anyone used v2 reagents with NCS v1.3? If so, what did the results look like?
          Quote from here:http://support.illumina.com/sequenci...xtseq-500.html

          NCS v1.4—Required for use with the NextSeq 500 Kit v2. NCS v1.4 is optimized for v2 reagents. Earlier kit versions are not compatible.

          NCS v1.3—Compatible NextSeq v1 reagents.

          Comment

          • SA2002
            Junior Member
            • Apr 2015
            • 2

            #65
            Thanks for the replies. Actually we recently ran kits with v1.3 but noticed that the stickers on the flow cells looked slightly different from previous flow cells and we got low density clusters. We're in the process of trying to figure out whether our reagents were v1 or v2 and was just curious if anyone else ran into the same issue.

            Comment

            • rogerzzw
              Member
              • Mar 2015
              • 16

              #66
              WGBS poor index and read2 on Nextseq

              Hi, Guys

              I am new to NGS. We've been trying to do whole genome bisulfite sequencing with Nugen Kit on Nextseq 500 for a few times with 2 samples.

              We always encountered the same problem. After demultiplexing, there is about 50% of unknown index reads (most of them are "GGGGG" and "NNNNNN") . According to other centers using exact same kit and machine, they do not have any problems with adapters and indexes. And we always use single-use illumina kit and reagent for each run. So we do not know what happens.

              We ve also tried to map read1 and read2 (with unknown barcodes). The alignment of read1 is quite good (90% aligniable), but reads2 is not good (~40%).

              We used custom primer for read1 and illumina primers for read2 and index.

              Do you guys have some ideas on what happened?

              Many thanks in advance!

              Roger

              Comment

              • GenoMax
                Senior Member
                • Feb 2008
                • 7142

                #67
                @rogerzzw: I assume you have been in touch with Illumina tech support about this? What was their take?

                In general tag read issues with other illumina sequencers (don't have direct NextSeq experience) happen when a sample is (borderline) overloaded. Has that possibility been eliminated?

                Comment

                • rogerzzw
                  Member
                  • Mar 2015
                  • 16

                  #68
                  Originally posted by GenoMax View Post
                  @rogerzzw: I assume you have been in touch with Illumina tech support about this? What was their take?

                  In general tag read issues with other illumina sequencers (don't have direct NextSeq experience) happen when a sample is (borderline) overloaded. Has that possibility been eliminated?

                  Hi, GenoMax

                  I talked to tech support. We loaded as suggested (1.8pM - 2pM), but our cluster density is always lower than expect( ~140k/mm2 vs ~200k/mm2).

                  I learnt from several posts that the tag reading may be due to overloaded. But in our case, we actually turned out lower density, which makes me very confused.

                  What do you think?

                  BTW, illumina said it is due to low diversity libraries. But I knew there are few groups only spike in 5% Phix and got quite good results.

                  Appreciate!

                  Comment

                  • GenoMax
                    Senior Member
                    • Feb 2008
                    • 7142

                    #69
                    Were your runs using V1 or V2 reagents? There is significant differences between chemistry and run procedures between those versions. Hopefully someone with direct NextSeq experience will be able to chime in soon.

                    Comment

                    • rogerzzw
                      Member
                      • Mar 2015
                      • 16

                      #70
                      Originally posted by GenoMax View Post
                      Were your runs using V1 or V2 reagents? There is significant differences between chemistry and run procedures between those versions. Hopefully someone with direct NextSeq experience will be able to chime in soon.
                      We always use V2.

                      Our machine is brand new, and it had a good 1st run only on Phix. Then I started to see the tag reading problem for 4 times.

                      Once comparing with other people's method of library prep, we found no difference between two. But we always is quite far from expectation.

                      Comment

                      • kentawan
                        Member
                        • Apr 2014
                        • 14

                        #71
                        Originally posted by GenoMax View Post
                        Were your runs using V1 or V2 reagents? There is significant differences between chemistry and run procedures between those versions. Hopefully someone with direct NextSeq experience will be able to chime in soon.
                        V2 is superb! One of our run uses the same library, once on the V1 and another on V2. The V1 run's cluster density was about 170(sorry i forgotten the exact unit, while V2 was 198.

                        Our V1 run gave us q30 passing of about 78% while the V2 gave us 92%! Despite the higher density cluster!

                        Comment

                        • rogerzzw
                          Member
                          • Mar 2015
                          • 16

                          #72
                          Originally posted by kentawan View Post
                          V2 is superb! One of our run uses the same library, once on the V1 and another on V2. The V1 run's cluster density was about 170(sorry i forgotten the exact unit, while V2 was 198.

                          Our V1 run gave us q30 passing of about 78% while the V2 gave us 92%! Despite the higher density cluster!
                          Hi, Kentawan.

                          I am a beginner of NGS on Nextseq 500. I am primarily doing whole genome bisulfite sequencing with Nextseq 500. I saw your post on this forum and would like to learn from you.

                          1. based on what you said, you use V2 kit and got a very good cluster density. we used V2 but never get that high cluster density. Ours is about 140K. Can you share your experience on that? Appreciate!

                          2. We always had a big issue on Barcode reading, other metrics look very good. There are a high portion (35%) reads are reading as "GGGGGG", I think it is due to reading failure, but we do not know how to pin it down, we pooled two samples each time. Unfortunately, there was one time, we pooled with other 4 Chip-seq samples and still the same. Do you some idea how to fix it?

                          Thank you very much

                          Roger

                          Comment

                          • GenoMax
                            Senior Member
                            • Feb 2008
                            • 7142

                            #73
                            Originally posted by rogerzzw View Post
                            There are a high portion (35%) reads are reading as "GGGGGG", I think it is due to reading failure, but we do not know how to pin it down, we pooled two samples each time. Unfortunately, there was one time, we pooled with other 4 Chip-seq samples and still the same.

                            Roger
                            That is not good news. That would mean there is something wrong with your bisulfite libraries since low nucleotide diversity should not have been an issue with 4 additional samples (these additional 4 samples had diverse barcodes?).

                            Comment

                            • rogerzzw
                              Member
                              • Mar 2015
                              • 16

                              #74
                              Originally posted by GenoMax View Post
                              That is not good news. That would mean there is something wrong with your bisulfite libraries since low nucleotide diversity should not have been an issue with 4 additional samples (these additional 4 samples had diverse barcodes?).
                              Yes, those 4 Chip-seq samples barcodes are well selected. But, the problem is I am not sure if it is due to low diversity since it is highly diverse when we combine chip-seq samples. It was still problematic when demultiplex, pretty much similar to the trial when we only have 2 samples for bisulfite sequencing.

                              Comment

                              • rogerzzw
                                Member
                                • Mar 2015
                                • 16

                                #75
                                @GenoMax

                                if it is library issue, 60-70% are demultiplexed very well, I cannot explain that.

                                Comment

                                Latest Articles

                                Collapse

                                • SEQadmin2
                                  Beyond CRISPR/Cas9: Understand, Choose, and Use the Right Genome Editing Tool
                                  by SEQadmin2



                                  CRISPR/Cas9 sparked the gene editing revolution for both research and therapeutics.1 But this system still showed severe issues that limited its applications. The most prominent were the heavy reliance on PAM sequences, delivery limitations, double-stranded breaks that prompt unintended edits and cell death, and editing inefficiency (both in targeting and in knock-in reliability).

                                  Despite this, “CRISPR helped turn genome editing from a specialized technique into
                                  ...
                                  07-31-2026, 11:01 AM
                                • SEQadmin2
                                  Proteomic Platforms: How to Choose the Right Analytical Strategy to Improve Detection and Clinical Applications
                                  by SEQadmin2


                                  Proteomics platforms are evolving rapidly, with advances in mass spectrometry and affinity-based approaches expanding what researchers can detect and at what scale. As the field moves toward deeper proteome coverage and clinical applications, scientists face an increasingly complex landscape of tools. This article will explore how researchers are navigating these choices to find the right platform for their work.

                                  The systematic characterization of the human proteome has
                                  ...
                                  07-20-2026, 11:48 AM
                                • SEQadmin2
                                  Advanced Sequencing Platforms Tackle Neuroscience’s Toughest Genomics Problems
                                  by SEQadmin2



                                  Genomics studies in neuroscience face a special challenge due to the brain’s complexity and scarcity of samples. Mapping changes in cell type and state using conventional next-generation sequencing methods remains challenging. Advances in technologies like single-cell sequencing, spatial transcriptomics, and long-read sequencing have opened the door to deeper studies of the brain and diseases like Alzheimer’s, amyotrophic lateral sclerosis (ALS), and schizophrenia.
                                  ...
                                  07-09-2026, 11:10 AM

                                ad_right_rmr

                                Collapse

                                News

                                Collapse

                                Topics Statistics Last Post
                                Started by SEQadmin2, Today, 10:13 AM
                                0 responses
                                10 views
                                0 reactions
                                Last Post SEQadmin2  
                                Started by SEQadmin2, 07-31-2026, 02:55 AM
                                0 responses
                                23 views
                                0 reactions
                                Last Post SEQadmin2  
                                Started by SEQadmin2, 07-24-2026, 12:17 PM
                                0 responses
                                19 views
                                0 reactions
                                Last Post SEQadmin2  
                                Started by SEQadmin2, 07-23-2026, 11:41 AM
                                0 responses
                                18 views
                                0 reactions
                                Last Post SEQadmin2  
                                Working...