    Custom Index Adapters

    How would I verify whether custom adapters have annealed properly? I get odd results when trying to quantify using QuBit, PicoGreen or TapeStation. Since the majority of the adapter is still single stranded, is there even a way to tell?
  • nucacidhunter
    Jafar Jabbari
    • Jan 2013
    • 1250

    #2
    As you have done, Qubit or PicoGreen is the easiest and very informative. With increasing annealing efficiency you should see increasing value in concentration of dsDNA. I regularly use this method as a QC for adapters with full length or partial length complimentary regions to ensure batch to batch consistency. Obviously, adapters with short complimentary region will have lower florescence, but it is sensitive enough to discriminate between annealed and non-annealed oligos. One also can calculate theoretical expected concentration for partially or completely complimentary oligos for any given annealing efficiency.

    Comment

    • jwfoley
      Senior Member
      • Jun 2009
      • 183

      #3
      You could also try denaturing them to see how much that reduces their fluorescence with a dsDNA-specific dye.

      Comment

      • Ingeneious
        Member
        • Dec 2014
        • 30

        #4
        Originally posted by nucacidhunter View Post
        As you have done, Qubit or PicoGreen is the easiest and very informative. With increasing annealing efficiency you should see increasing value in concentration of dsDNA. I regularly use this method as a QC for adapters with full length or partial length complimentary regions to ensure batch to batch consistency. Obviously, adapters with short complimentary region will have lower florescence, but it is sensitive enough to discriminate between annealed and non-annealed oligos. One also can calculate theoretical expected concentration for partially or completely complimentary oligos for any given annealing efficiency.
        What kind of values have you seen with say a full TruSeq adapter?

        Originally posted by jwfoley View Post
        You could also try denaturing them to see how much that reduces their fluorescence with a dsDNA-specific dye.
        That's a good idea. I tried something similar by comparing the annealed(?) product versus an equimolar mixture of unannealed oligos. With a Qubit dsDNA BR kit it did not seem to make a difference - they all seemed to generate a similar value. With PicoGreen I believe I received roughly double the concentration. However, even the individual ssDNA oligos generated a value.

        Comment

        • nucacidhunter
          Jafar Jabbari
          • Jan 2013
          • 1250

          #5
          What kind of values have you seen with say a full TruSeq adapter?
          I have not done exact math. Using 10 uM (each) 40 base long adapter oligos with full complimentary bases across oligos I get around 70 ng/ul with PicoGreen and for 10 uM 30 base long oligos with 14 complimentary nucleotides 30 ng/ul.

          Comment

          Latest Articles

          Collapse

          • SEQadmin2
            Beyond CRISPR/Cas9: Understand, Choose, and Use the Right Genome Editing Tool
            by SEQadmin2



            CRISPR/Cas9 sparked the gene editing revolution for both research and therapeutics.1 But this system still showed severe issues that limited its applications. The most prominent were the heavy reliance on PAM sequences, delivery limitations, double-stranded breaks that prompt unintended edits and cell death, and editing inefficiency (both in targeting and in knock-in reliability).

            Despite this, “CRISPR helped turn genome editing from a specialized technique into
            ...
            07-31-2026, 11:01 AM
          • SEQadmin2
            Proteomic Platforms: How to Choose the Right Analytical Strategy to Improve Detection and Clinical Applications
            by SEQadmin2


            Proteomics platforms are evolving rapidly, with advances in mass spectrometry and affinity-based approaches expanding what researchers can detect and at what scale. As the field moves toward deeper proteome coverage and clinical applications, scientists face an increasingly complex landscape of tools. This article will explore how researchers are navigating these choices to find the right platform for their work.

            The systematic characterization of the human proteome has
            ...
            07-20-2026, 11:48 AM

          ad_right_rmr

          Collapse

          News

          Collapse

          Topics Statistics Last Post
          Started by SEQadmin2, 08-13-2026, 12:22 PM
          0 responses
          25 views
          0 reactions
          Last Post SEQadmin2  
          Started by SEQadmin2, 08-11-2026, 10:35 AM
          0 responses
          20 views
          0 reactions
          Last Post SEQadmin2  
          Started by SEQadmin2, 08-06-2026, 07:41 AM
          0 responses
          33 views
          0 reactions
          Last Post SEQadmin2  
          Started by SEQadmin2, 08-03-2026, 10:13 AM
          0 responses
          51 views
          0 reactions
          Last Post SEQadmin2  
          Working...