    Anybody here with a 3000/4000 seen poor quality read 2's lately? I have attached example plots from 4 previous runs below. It seems to be library agnostic.

    I've been lead to understand that there is an escalating investigation internal to Illumina into possible chemistry issues. They're not comp-ing runs until lot numbers are given for the failed runs, I know that.

    Just wondering how far the problem extends.
    Attached Files
  • Brian Bushnell
    Super Moderator
    • Jan 2014
    • 2709

    #2
    That is consistent with my analysis of data from the HiSeq 4000 platform, provided to us by Illumina. The chemistry (as of a few months ago) was absolutely not up to 2x150bp runs IMO, and read 2 particularly had very low quality. I also analyzed some 2x100bp PhiX data that had decent quality for both read 1 and read 2. So, the platform may be adequate for 2x100bp runs.

    Comment

    • GenoMax
      Senior Member
      • Feb 2008
      • 7142

      #3
      We have usual "read2-gate" (like the antenna-gate, bend-gate for new apple products) for a new Illumina technology?

      On a serious note, @GW_OK what kind of libraries are represented by these 4 examples?
      Last edited by GenoMax; 12-09-2015, 07:24 AM.

      Comment

      • GW_OK
        Senior Member
        • Sep 2009
        • 411

        #4
        Libraries given above are (in no order) a mix of:
        Kapa Hyper Prep, Bioo Scientific Nextflex PCR-free, Illumina Nextera

        We've got a Bioo Scientific Nextflex flowcell running now. I'll update once it completes.

        Illumina is sending FSE's to pull our entire optics module and check alignments next week. There is hope that it will be less-bad afterwards.

        Comment

        • luc
          Senior Member
          • Dec 2010
          • 469

          #5
          Hi GWOK, thanks for starting this thread!

          We are experiencing the same problems (as is another lab in the neighborhood). The quality of the PE150 bp reads was perfectly fine a few months ago. Illumina likley has reagent quality problems at the moment.
          The quality scores are now dropping considerably faster than a few months ago.
          The % at which the median Q30 score curve ends (e.g. at cycle 310) seems to be a good quantifier for the problems. The curve used to end at about 60%, now often under 20%, even under 10%.

          Illumina tech support did inform us that there are "NO reagent quality problems" - because all reagents individually pass QC. However they are investigating if there is a similar phenomenon ( a "non-problem"?) as with the MiSeq reagents which generate reduced read qualities scores since more than half a year now ( http://seqanswers.com/forums/showthread.php?t=59558 ).

          Please keep us updated.
          Last edited by luc; 12-09-2015, 06:38 PM.

          Comment

          • Brian Bushnell
            Super Moderator
            • Jan 2014
            • 2709

            #6
            Luc, which platform (and chemistry, and software) are you having trouble with?

            FYI, our HiSeq 2000/2500 machines seem fine, which is one of the reasons I have recommended against "upgrading" to HS3000 or 4000 (from which I have never seen good data). As far as I know, our NextSeq is fine too, but our MiSeqs are outputting junk for 2x300bp amplicon runs.
            Last edited by Brian Bushnell; 12-09-2015, 08:03 PM.

            Comment

            • GW_OK
              Senior Member
              • Sep 2009
              • 411

              #7
              Latest run has finished. Two large Bioo NextFlex pools across respective halves of the flowcell.

              Read 2 still looks weak. One thing of note is that the top surface looks better than the bottom. That being said neither surface's read 2 look nearly as good as read 1.

              luc, thanks for posting. I'm glad I'm not the only one seeing this.

              I've also included a Q30 plot from Illumina's "HiSeq 4000: TruSeq PCR-Free (NA12878)" publicly shared run on BaseSpace as an example of what I imagine good runs are supposed to look like.
              Attached Files

              Comment

              • luc
                Senior Member
                • Dec 2010
                • 469

                #8
                Hi Brian,

                We have seen the problems on the HiSeq 4000 the last 6 weeks, and since quite a while on our Miseqs (we are underclustering considerably for these to get any usable data).



                Originally posted by Brian Bushnell View Post
                Luc, which platform (and chemistry, and software) are you having trouble with?

                FYI, our HiSeq 2000/2500 machines seem fine, which is one of the reasons I have recommended against "upgrading" to HS3000 or 4000 (from which I have never seen good data). As far as I know, our NextSeq is fine too, but our MiSeqs are outputting junk for 2x300bp amplicon runs.
                Last edited by luc; 12-10-2015, 06:13 PM.

                Comment

                • GW_OK
                  Senior Member
                  • Sep 2009
                  • 411

                  #9
                  Update.
                  Our FSE has been on site all week working on realigning the optical path and running several tests on the fluidics, hoping that this will mitigate the issue. Based on what the factory people told him from data sent off we were ever so slightly off in some sort of image correction model. I view the instrument tweaking as dubious but have been told that this must be done first before the chemistry is brought into question.

                  We'll have our first test run on Monday. I'll upload performance once it completes.

                  Comment

                  • GW_OK
                    Senior Member
                    • Sep 2009
                    • 411

                    #10
                    Update.
                    After running a mix of PhiX and other libraries Q30 plots are still not great but perhaps less terrible. Attached here are two plots showing a PhiX lane as well as a Nextera lane (PhiX is missing one of the two index reads).

                    While neither are bottoming out as before there's certainly a steep downward trend in read after the intensity jump. This run did "meet spec" for a 3k run (75% >Q30) but Illumina still wants to send out an "instrument expert" to have a look at our machine.
                    Attached Files

                    Comment

                    • GW_OK
                      Senior Member
                      • Sep 2009
                      • 411

                      #11
                      Two more since then have given mixed results. The first was a Bioo NextFlex PCR-free and is perhaps one of the best runs we've had on this machine. The Q30 plot is attached (the one with black boxes).

                      After that run another, with a mix of library types, gave poorer performance. The Q30 plot for this is also attached, this time with blue boxes.
                      Attached Files

                      Comment

                      • luc
                        Senior Member
                        • Dec 2010
                        • 469

                        #12
                        Ours tend to look unfortunately like your last example - we have not seen significant differences between customer prepped libraries our own libraries and also PCR-free libraries.
                        At what concentration are you loading the PCR-free libraries and and how many clusters did you sequence?

                        Thanks!
                        Originally posted by GW_OK View Post
                        Two more since then have given mixed results. The first was a Bioo NextFlex PCR-free and is perhaps one of the best runs we've had on this machine. The Q30 plot is attached (the one with black boxes).

                        After that run another, with a mix of library types, gave poorer performance. The Q30 plot for this is also attached, this time with blue boxes.

                        Comment

                        • GW_OK
                          Senior Member
                          • Sep 2009
                          • 411

                          #13
                          A couple more runs that don't meet spec.

                          Here's the first. It's two pools of NextFlex PCR-free on respective halves. Loaded at 150pM.
                          Q30 tanks towards the end:


                          We have a look at the %base and say "Ah Ha, sneaky adapter dimers slipping in!". Granted one should be able to clearly see adapter dimer traces on a Tapestation or Bioanalyzer but with PCR-free adapters things don't migrate properly and sizes are all off (we have tried denaturing and running it on a RNA tape, with some success).


                          Apparently the ExAmp and ordered flowcells just go bonkers with short libraries and that's what's killing the end of our reads, right?

                          So we take the two pools, give them a 0.8x bead cleanup (as opposed to our typical 0.9x) and load them back up again. We load at 100pM this time to really knock down any possibility of polyclonal wells.

                          The %base plot shows no more adapter dimers:


                          And read 1 Q30 looks pretty good relative to what you can expect on a 3k, but read 2 Q30 is an almost straight slope down:


                          I'm given to understand there's yet another lab out there that's having troubles with a 3k/4k. They've decided to not even bother going out to 150bp and just truncate the runs to 120bp. If that's the way this ends up going here I better not have to pay for those extra cycles.

                          Comment

                          • Brian Bushnell
                            Super Moderator
                            • Jan 2014
                            • 2709

                            #14
                            From what I've seen, you'll get much better results doing 2x100 runs in the first place, rather than truncating the 2x150 runs...

                            Comment

                            • GenoMax
                              Senior Member
                              • Feb 2008
                              • 7142

                              #15
                              Our initial 4000 run (2D x 151) was a mix of random samples from past with a lane of phiX. With the caveat that n=1 things do not look too bad. Still analyzing the data.
                              Attached Files

                              Comment

                              Latest Articles

                              Collapse

                              ad_right_rmr

                              Collapse

                              News

                              Collapse

                              Topics Statistics Last Post
                              Started by SEQadmin2, 06-09-2026, 11:58 AM
                              0 responses
                              22 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 06-05-2026, 10:09 AM
                              0 responses
                              29 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 06-04-2026, 08:59 AM
                              0 responses
                              39 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 06-02-2026, 12:03 PM
                              0 responses
                              61 views
                              0 reactions
                              Last Post SEQadmin2  
                              Working...