  • theduke
    Member
    • Aug 2010
    • 14

    #1

    Uneven reads in pooled sample

    Hi All,

    I've been doing this a while, but have run into a new problem on a batch of RNA-Seq samples (Truseq) that I'm concerned about. We prepped a bunch of RNA samples for sequencing and have a few that are low concentration. The core facility that I have been using for years was unhappy with this, but I didn't really see where the issue was given there was plenty of material for pooling.

    Nevertheless they ran one lane for me (HiSeq), but apparently had to modify their dilution protocol. The result was as follows for the lane:


    Reads % lane % PF cluster Qual score
    Sample*1 31,287,035 12.66 89.81 36.96
    Sample*2 70,103,756 28.36 89.49 36.96
    Sample*3 8,599,888 3.48 90.17 36.97
    Sample*4 3,309,119 1.34 89.80 37.01
    Sample*5 98,775,103 39.95 89.24 36.94
    Sample*6 28,415,646 11.49 90.10 36.94

    Now, clearly the issue is that the distribution of data is very uneven, BUT the number of reads directly and perfectly correlate with the starting concentration of each sample prior to pooling (R2 = 0.99)! From what I can gather, the core did not make an intermediate dilution for each but instead added variable volumes to a pool and used that to cluster, but what happened in between I do not know. What I dont understand is why they did this. Here are the sample concs:

    ng/ul nM
    Sample1 1.6 6.3
    Sample2 2.4 9.6
    Sample3 0.9 3.5
    Sample4 0.6 2.4
    Sample5 2.7 10.8
    Sample6 1.5 6.1

    We are almost certain that given the correlation with the pre-pooled sample concentration the pooling has been done incorrectly and that the variable volumes has perhaps thrown everything off balance. Does that sound reasonable?

    According to all the illumina documentation I have read, the best thing to do would be to normalize each sample to 2 nM, and then combine equal volume of each together to make a 2 nM pool, which would then be the starting point for cluster generation. Does this sound reasonable? Am I missing something here?

    ****I should mention that samples were quantified with both picogreen and bioanalyzer HS DNA. These gave very similar results (r = 0.8) so I doubt quantification is a major issue here****
    Last edited by theduke; 02-03-2016, 11:10 AM.
  • nucacidhunter
    Jafar Jabbari
    • Jan 2013
    • 1250

    #2
    If the library sizes were similar, they have failed to do equimolar pooling. Using different volumes for pooling or equal volumetric pooling of normalised libraries should not affect the read ratios if done correctly. qPCR is the most accurate library quantification method and the ones mentioned here are not.

    Comment

    • theduke
      Member
      • Aug 2010
      • 14

      #3
      Agreed that qPCR is the way to go for accurate quantification, but that's a separate issue. Here we are talking about relative concentrations and the pooling of multiple samples. It certainly looks like equimolar pooling has failed in this instance.

      Comment

      • HESmith
        Senior Member
        • Oct 2009
        • 512

        #4
        We've consistently observed lower-than-calculated cluster numbers when working with dilute (<1ng/ul) libraries. We don't think it's an issue with adsorption b/c we use LoBind tubes, nor an issue with the quantification (using PicoGreen fluorometry and BioAnalyzer, same as OP). We suspect that it's related to library quality, particularly when prepared in parallel. If the same amount of RNA input yields substantially different amounts of library, it's reasonable to assume that something wonky happened.

        Also, you may want to recalculate your correlation. For example, sample 2 (9.6 nM) is 4X the concentration of sample 4 (2.4 nM), but produces 21X as many reads (70M / 3.3M).

        Comment

        • theduke
          Member
          • Aug 2010
          • 14

          #5
          What was the cutoff for the observed weirdness? <1ng/ul? I think we will redo the 5 or 6 samples that are odd ones out and start over.

          Does my pooling plan sound reasonable?

          Comment

          • nucacidhunter
            Jafar Jabbari
            • Jan 2013
            • 1250

            #6
            Originally posted by HESmith View Post
            We've consistently observed lower-than-calculated cluster numbers when working with dilute (<1ng/ul) libraries. We don't think it's an issue with adsorption b/c we use LoBind tubes, nor an issue with the quantification (using PicoGreen fluorometry and BioAnalyzer, same as OP). We suspect that it's related to library quality, particularly when prepared in parallel. If the same amount of RNA input yields substantially different amounts of library, it's reasonable to assume that something wonky happened.

            Also, you may want to recalculate your correlation. For example, sample 2 (9.6 nM) is 4X the concentration of sample 4 (2.4 nM), but produces 21X as many reads (70M / 3.3M).
            Lower than expected yield indicates some issues with library prep but it could be unrelated to quality, for instance loss during library clean-up of a good library. If there was an issue with library quality, then the cluster density and yield should be lower than expected as the bad library would not contribute to sequencing output. Because of formatting issue I can't figure out the stats given on output here. Pooling equimolar quantity of libraries quantified with qPCR with below 1 ng/ul and up to 10 ng/ul I have not observed this.

            Comment

            • theduke
              Member
              • Aug 2010
              • 14

              #7
              Originally posted by nucacidhunter View Post
              Because of formatting issue I can't figure out the stats given on output here.
              Apologies for that. In the first table, there are five columns in the following order:

              Sample name
              Number of reads
              % of the lane
              %PF clusters
              Quality score

              There doesn't appear to be any quality issues per se.

              Comment

              • nucacidhunter
                Jafar Jabbari
                • Jan 2013
                • 1250

                #8
                Sequencing reads is over 230M which is around average for V4 chemistry on HiSeq 2500 if that is the platform used for sequencing and read quality and PF is up to specs.

                Comment

                Latest Articles

                Collapse

                • SEQadmin2
                  Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
                  by SEQadmin2



                  Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

                  Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
                  ...
                  10-14-2026, 09:55 AM
                • SEQadmin2
                  New Genomics Technologies Take Aim at Long-Standing Limits
                  by SEQadmin2


                  Researchers using sequencing and genomics tools often have to make trade-offs. They can choose between speed or scale, short reads or long-range information, or targeted panels or a view of the whole transcriptome. New technologies that have been released this year are built to address those tough choices.

                  We asked six companies the same four questions to learn about their latest products. The new technologies bring a lot to the table, including rethinking sequencing
                  ...
                  09-28-2026, 10:25 AM

                ad_right_rmr

                Collapse

                News

                Collapse

                Topics Statistics Last Post
                Started by SEQadmin2, 10-22-2026, 11:13 AM
                0 responses
                47 views
                0 reactions
                Last Post SEQadmin2  
                Started by SEQadmin2, 10-22-2026, 10:48 AM
                0 responses
                35 views
                0 reactions
                Last Post SEQadmin2  
                Started by SEQadmin2, 10-15-2026, 10:52 AM
                0 responses
                46 views
                0 reactions
                Last Post SEQadmin2  
                Started by SEQadmin2, 10-14-2026, 05:39 AM
                0 responses
                46 views
                0 reactions
                Last Post SEQadmin2  
                Working...