    3500 genetic analyzer low quality value

    Dear professor
    we have a new 3500 genetic analyzer that sanger/dye terminator from applied biosystem. All amplicon have e low to medium quality value the same site 75-85 bp including the control PGEM, however the seq standard is ok. please check attached files
    Any comments
    regards
    Ghad
    Attached Files
  • MU Core
    Member
    • Apr 2008
    • 60

    #2
    Ghada, you're observing the presence of "dye blobs" which is the result of poor removal of unincorporated dye terminators from the post-sequencing reaction. A common issue in dye terminator sequencing. You'll want to revisit your reaction clean-up method to see if changes can be made to optimize the method. Or, you may want to consider a different approach to reaction clean-up. Performing a google search with "removing dye blobs" will provide you with information on the topic.

    Comment

    • Ghada
      Junior Member
      • Nov 2014
      • 5

      #3
      thank you MU core for your help.
      Regards
      Ghada

      Comment

      Latest Articles

      Collapse

      ad_right_rmr

      Collapse

      News

      Collapse

      Topics Statistics Last Post
      Started by SEQadmin2, 06-09-2026, 11:58 AM
      0 responses
      14 views
      0 reactions
      Last Post SEQadmin2  
      Started by SEQadmin2, 06-05-2026, 10:09 AM
      0 responses
      26 views
      0 reactions
      Last Post SEQadmin2  
      Started by SEQadmin2, 06-04-2026, 08:59 AM
      0 responses
      36 views
      0 reactions
      Last Post SEQadmin2  
      Started by SEQadmin2, 06-02-2026, 12:03 PM
      0 responses
      60 views
      0 reactions
      Last Post SEQadmin2  
      Working...