    MiSeq run_Error

    Hello,

    Can anyone give me any help on this problem: Today I started an amplicon sequencing run on the MiSeq platform and after hundred minutes a window popped up with the following error message: "Poor-quality focus model data" Aborting run!

    Thanks a lot!
    Steffi1
  • GenoMax
    Senior Member
    • Feb 2008
    • 7142

    #2
    @Steffi: Please contact Illumina tech support who can remotely look at this run (assuming your machine is accessible over the network otherwise you will need to send them some files) to diagnose the issue.

    Sounds like the machine had gone past the clustering step and must have encountered some other error. Was the cluster density number populated on the screen in MCS?

    Comment

    • Steffi1
      Junior Member
      • Apr 2016
      • 3

      #3
      I have already contact Illumina tech support.

      No, the cluster density number was not populated on the screen.

      Comment

      • Jessica_L
        Senior Member
        • Feb 2010
        • 117

        #4
        What chemistry version and run length are you running?

        Also, what version of MCS are you running? There are known issues with MCS 2.6 that can cause focusing problems on sequencing runs.

        Comment

        • Steffi1
          Junior Member
          • Apr 2016
          • 3

          #5
          @ Jessica: I used the MiSeq v3 reagents and the length of my amplicons were between 113-112bp (without adapters, indices..). Do you think, these products/amplicons are too small to detect by the camera (i have not any experience with sequencing). I know, I can also use the MiSeq v2 chemistry, but the 2x250bp output 7,5Gb, no. of reads 15M is too little (because i need to do targeted deep sequencing).

          We use MCS 2.5.0.5, do you have same problems?

          Comment

          • avo
            Member
            • Sep 2013
            • 14

            #6
            Probably worth checking a few thumbnail images to have at least an estimate if and how many clusters are present.

            I would assume either massive over- or underclustering

            Comment

            • Jessica_L
              Senior Member
              • Feb 2010
              • 117

              #7
              Originally posted by Steffi1 View Post
              @ Jessica: I used the MiSeq v3 reagents and the length of my amplicons were between 113-112bp (without adapters, indices..). Do you think, these products/amplicons are too small to detect by the camera (i have not any experience with sequencing). I know, I can also use the MiSeq v2 chemistry, but the 2x250bp output 7,5Gb, no. of reads 15M is too little (because i need to do targeted deep sequencing).

              We use MCS 2.5.0.5, do you have same problems?
              MCS 2.5.x should be fine. Illumina had issued a notice some months back about a problem with v2.6 that was causing focusing issues, so I wanted to make sure that wasn't causing your failure. The amplicon size without any adapters or indices also sounds okay to me.

              I know some users have reported issues with the 2x300 v3 chemistry, but it sounds like you're experiencing a failure at cluster generation or first base incorporation, which would be something altogether different.

              Assuming the run stopped in the first cycle and you don't have thumbnail images to share, can you at least let us know what kind of libraries these are (amplicons, but how did you prep them? Kit manufacturer, etc.)? Also, how did you quantify the finished libraries and what concentration did you load on the sequencer?

              Lastly, has Illumina tech support gotten back to you yet?

              Comment

              Latest Articles

              Collapse

              • SEQadmin2
                Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
                by SEQadmin2



                Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

                Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
                ...
                Yesterday, 09:55 AM
              • SEQadmin2
                New Genomics Technologies Take Aim at Long-Standing Limits
                by SEQadmin2


                Researchers using sequencing and genomics tools often have to make trade-offs. They can choose between speed or scale, short reads or long-range information, or targeted panels or a view of the whole transcriptome. New technologies that have been released this year are built to address those tough choices.

                We asked six companies the same four questions to learn about their latest products. The new technologies bring a lot to the table, including rethinking sequencing
                ...
                09-28-2026, 10:25 AM

              ad_right_rmr

              Collapse

              News

              Collapse

              Topics Statistics Last Post
              Started by SEQadmin2, Today, 10:52 AM
              0 responses
              8 views
              0 reactions
              Last Post SEQadmin2  
              Started by SEQadmin2, Yesterday, 05:39 AM
              0 responses
              15 views
              0 reactions
              Last Post SEQadmin2  
              Started by SEQadmin2, 10-13-2026, 09:56 AM
              0 responses
              15 views
              0 reactions
              Last Post SEQadmin2  
              Started by SEQadmin2, 10-09-2026, 11:14 AM
              0 responses
              27 views
              0 reactions
              Last Post SEQadmin2  
              Working...