    Trouble frequency of HiSeq2500

    How frequently have you been with trouble?
    We have five HiSeq2500 installed in 2012. Since last year to today we have run trouble once in every 10 runs. Troubles were caused by various reasons, hard disc error, syringe troubles, temperature control error, communication error etc. Non of our five HiSeq is trouble free.

    Does anyone share your trouble rate of HiSeq2500 sequence run?
  • MU Core
    Member
    • Apr 2008
    • 60

    #2
    Our group operates a single HiSeq 2500. We measure down time as the instrument being unavailable due to a hardware or software issue. Over the past few years it has been <5%. The past 10 months the down time has been 3.1% or 19 days based on a 365 day schedule. I think this is a typical rate and equate to 2-3 service calls per year.

    Comment

    • nkondo
      Junior Member
      • Mar 2016
      • 4

      #3
      Our average service calls are 3 - 5 per month, not a year. Even we have five instruments, these numbers seems very high.

      Comment

      • luc
        Senior Member
        • Dec 2010
        • 469

        #4
        The experience and the IQ of your Illumina engineer will make a big difference.
        We once a had a HS2500 which failed often due to multiple intermittent problems, some of which never got resolved until Illumina finally substituted the machine. The replacement system worked reliably for a year then. It is my impression that the HS2500 parts are not built to last.
        Last edited by luc; 04-24-2016, 09:33 AM.

        Comment

        • GenoMax
          Senior Member
          • Feb 2008
          • 7142

          #5
          Originally posted by nkondo View Post
          Our average service calls are 3 - 5 per month, not a year. Even we have five instruments, these numbers seems very high.
          That is the reason one maintain service contracts But if the service calls are resulting in significant down times then that may not be good for your facility.

          Your FAS has a way of plotting instrument up-times/issues found based on the backend service database at Illumina. If you have not had a meeting with him/her you should schedule one to discuss the performance of your instruments. Any trends that emerge (e.g. particular machine) can be used for targeted remediation. Money spent of consumables is always a good incentive for getting these issues fixed.

          Comment

          Latest Articles

          Collapse

          • SEQadmin2
            Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
            by SEQadmin2



            Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

            Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
            ...
            10-14-2026, 09:55 AM
          • SEQadmin2
            New Genomics Technologies Take Aim at Long-Standing Limits
            by SEQadmin2


            Researchers using sequencing and genomics tools often have to make trade-offs. They can choose between speed or scale, short reads or long-range information, or targeted panels or a view of the whole transcriptome. New technologies that have been released this year are built to address those tough choices.

            We asked six companies the same four questions to learn about their latest products. The new technologies bring a lot to the table, including rethinking sequencing
            ...
            09-28-2026, 10:25 AM

          ad_right_rmr

          Collapse

          News

          Collapse

          Topics Statistics Last Post
          Started by SEQadmin2, 10-22-2026, 11:13 AM
          0 responses
          28 views
          0 reactions
          Last Post SEQadmin2  
          Started by SEQadmin2, 10-22-2026, 10:48 AM
          0 responses
          21 views
          0 reactions
          Last Post SEQadmin2  
          Started by SEQadmin2, 10-15-2026, 10:52 AM
          0 responses
          38 views
          0 reactions
          Last Post SEQadmin2  
          Started by SEQadmin2, 10-14-2026, 05:39 AM
          0 responses
          44 views
          0 reactions
          Last Post SEQadmin2  
          Working...