    Low intensity & quality for Index Read 2

    We've encountered a problem with the last couple of paired end, dual indexed runs on our HiSeq 2500 and I wanted to check if any others have had similar experience. The problem is that the signal intensity and thus the read quality of Index read #2 is very poor. All other aspects of the run, meaning Read 1, Index 1 and Read 2 work perfectly fine, good signal and quality. The poor quality of Index #2 results in problems with demultiplexing.

    We have run a couple of single read, dual index runs and in those cases index read 2 is fine, of course the manner in which index read 2 is done on single read vs paired end read flow cells differs.

    We contacted Illumina and they have no explanation. Nothing stood out in the log files to suggest a cause. The engineer did some back flushing and checking of the fluidics but we have not tested it since that was done.

    Does this sound familiar to anyone? Any ideas would be appreciated as well.
  • LaoR
    Junior Member
    • Jan 2012
    • 3

    #2
    Can you look the %base? See if you can read off the base pairs from it.

    I've encountered something similar for 3 rapid runs.If you can read off the base pairs for your second index, search to see if it's from part of the I5 adapter sequence. If so, it may be a "transient" fluidics problem. The machine probably didn't deliver reagents to strip off the previous index sequencing primer.

    Comment

    • HeinKey
      Member
      • May 2009
      • 21

      #3
      We have the same issue. Low quality index2 in paired-end runs, and normal quality index2 in Single read runs.
      Illumina finally recognized the problem, (we did some validation runs for them).
      We now have a work around where we receive a different dual index primer and run a custom receipe on our HiSeq2500 (V4). This way the index 2 primer is not the grafted oligo, but the separately delivered index 2 primer which probably is covering the dark cycle nucleotides. We did not get the details of this primer sequence....
      We will test this in the upcoming weeks.

      Comment

      • kmcarr
        Senior Member
        • May 2008
        • 1181

        #4
        Originally posted by HeinKey View Post
        We have the same issue. Low quality index2 in paired-end runs, and normal quality index2 in Single read runs.
        Illumina finally recognized the problem, (we did some validation runs for them).
        We now have a work around where we receive a different dual index primer and run a custom receipe on our HiSeq2500 (V4). This way the index 2 primer is not the grafted oligo, but the separately delivered index 2 primer which probably is covering the dark cycle nucleotides. We did not get the details of this primer sequence....
        We will test this in the upcoming weeks.
        HeinKey,

        Yes, it sounds exactly the same as the issues we have had. At least now I know it isn't something totally unique to our instrument. The problem was so specific to paired end, dual indexed runs that I assumed it had to have something to do with priming index 2 from the grafted oligo.

        We are starting another PE, dual index run today and we'll see how it goes. I'll report back and please do keep me posted on your results (you can PM me if your wish).

        Thanks, Kevin

        Comment

        • LaoR
          Junior Member
          • Jan 2012
          • 3

          #5
          If that's the case, wouldn't it be reproducible by using reagents from the same lot to run the libraries again using similar settings?

          In the beginning, Illumina wants to blame on the library prep. We used same lot numbers to rerun the "failed index 2" run on the instrument. Rerun was totally fine.

          Comment

          • Ryjinko
            Junior Member
            • Mar 2016
            • 1

            #6
            Our sequencer is NextSeq 500, and we encountered the same problem in paired-end dual-index sequencing of library constructed from PCR amplicon.

            And now we have solved this question. During the library construction procedure, we use cutomized P5 and P5 adapters synthesized by commercial provider, by sequence analyis, we found that the adapter missed one "A" base when we designed, while this "A" base was added by dA-tailing in the standard Illumina library construction peocudure.

            This is just the case for our NextSeq 500 sequencer using customized P5 and P7 adapters during paired-end dual-index sequencing.

            Hope that will be of little help.

            Comment

            Latest Articles

            Collapse

            • SEQadmin2
              Beyond CRISPR/Cas9: Understand, Choose, and Use the Right Genome Editing Tool
              by SEQadmin2



              CRISPR/Cas9 sparked the gene editing revolution for both research and therapeutics.1 But this system still showed severe issues that limited its applications. The most prominent were the heavy reliance on PAM sequences, delivery limitations, double-stranded breaks that prompt unintended edits and cell death, and editing inefficiency (both in targeting and in knock-in reliability).

              Despite this, “CRISPR helped turn genome editing from a specialized technique into
              ...
              07-31-2026, 11:01 AM
            • SEQadmin2
              Proteomic Platforms: How to Choose the Right Analytical Strategy to Improve Detection and Clinical Applications
              by SEQadmin2


              Proteomics platforms are evolving rapidly, with advances in mass spectrometry and affinity-based approaches expanding what researchers can detect and at what scale. As the field moves toward deeper proteome coverage and clinical applications, scientists face an increasingly complex landscape of tools. This article will explore how researchers are navigating these choices to find the right platform for their work.

              The systematic characterization of the human proteome has
              ...
              07-20-2026, 11:48 AM
            • SEQadmin2
              Advanced Sequencing Platforms Tackle Neuroscience’s Toughest Genomics Problems
              by SEQadmin2



              Genomics studies in neuroscience face a special challenge due to the brain’s complexity and scarcity of samples. Mapping changes in cell type and state using conventional next-generation sequencing methods remains challenging. Advances in technologies like single-cell sequencing, spatial transcriptomics, and long-read sequencing have opened the door to deeper studies of the brain and diseases like Alzheimer’s, amyotrophic lateral sclerosis (ALS), and schizophrenia.
              ...
              07-09-2026, 11:10 AM

            ad_right_rmr

            Collapse

            News

            Collapse

            Topics Statistics Last Post
            Started by SEQadmin2, 08-06-2026, 07:41 AM
            0 responses
            13 views
            0 reactions
            Last Post SEQadmin2  
            Started by SEQadmin2, 08-03-2026, 10:13 AM
            0 responses
            31 views
            0 reactions
            Last Post SEQadmin2  
            Started by SEQadmin2, 07-31-2026, 02:55 AM
            0 responses
            40 views
            0 reactions
            Last Post SEQadmin2  
            Started by SEQadmin2, 07-24-2026, 12:17 PM
            0 responses
            26 views
            0 reactions
            Last Post SEQadmin2  
            Working...