    cutting cost for library prep

    We do a small number of nextera libraries. Not enough to justify really delving into the kit and doing lots of experimentation to reduce costs. But I'm always wanting to cut the easy costs. So was thinking of just doing 1/5 reactions (follow their protocol but 1/5 the volumes for everything).

    I'd guess that this is fairly common way to cut costs. So am posting to request any tips or cautions from people doing this before we start?
    Microbial ecologist, running a sequencing core. I have lots of strong opinions on how to survey communities, pretty sure some are even correct.
  • nucacidhunter
    Jafar Jabbari
    • Jan 2013
    • 1250

    #2
    There are few papers on this subject. Generally reduce reaction volumes and input in proportion, strip transposase by adding SDS directly followed by PCR with mix of Illumina index primer and P5, P7 primers (flow cell binding motives). PCR reactions need to be done with a third paty polymerase at 50 ul for good yield.

    Comment

    • Friesac
      Junior Member
      • Sep 2009
      • 1

      #3
      I've been looking at this paper to do the exact same thing... http://journals.plos.org/plosone/art...l.pone.0128036. My reasoning is small genomes/high throughput. I'd be interested to hear how it goes for you thermophile.

      Comment

      • thermophile
        Senior Member
        • Apr 2015
        • 243

        #4
        We don't do enough to justify rewriting the protocol that much. Our last run with 1/2 reactions looked identical to full reactions and used the same procedural steps so I think I'll stick to that unless we suddenly are getting full plates of genomes to prep.

        thanks for the replies
        Microbial ecologist, running a sequencing core. I have lots of strong opinions on how to survey communities, pretty sure some are even correct.

        Comment

        • Simone78
          Senior Member
          • Oct 2010
          • 208

          #5
          Originally posted by thermophile View Post
          We don't do enough to justify rewriting the protocol that much. Our last run with 1/2 reactions looked identical to full reactions and used the same procedural steps so I think I'll stick to that unless we suddenly are getting full plates of genomes to prep.

          thanks for the replies
          I cut reaction volumes by a factor of 10 using a liquid handling robot and the Nextera kit (exactly the same protocol, no changes). Tagmentation is done in 2 ul, I add 0.5 ul NT (or 0.2% SDS) followed by 1 ul i5/i7 (1:5 dilution to avoid excess adaptors and save money) and 1.5 ul NPM. Input is 50-150 pg/rxn.
          Even without a robot this is perfectly doable with a pipette.
          I this way a 3000 USD kit will last for about 1000 samples instead of 96!

          /Simone

          Comment

          • thermophile
            Senior Member
            • Apr 2015
            • 243

            #6
            Thanks for that! Our liquid handler doesn't go under 1ul, but good to know you were able to just reduce vol.
            Microbial ecologist, running a sequencing core. I have lots of strong opinions on how to survey communities, pretty sure some are even correct.

            Comment

            • docsascha
              Junior Member
              • Dec 2013
              • 5

              #7
              Originally posted by Simone78 View Post
              I cut reaction volumes by a factor of 10 using a liquid handling robot and the Nextera kit ....

              /Simone
              Dear Simone

              Thanks for this info, which robot are you using, and how do you deal with magnetic beats separation and shaking, we have seen here some problems on our way for cutting cost

              Does anybody have some tips on how to use a more time saving method than Qbit for DNA measurement? We would like to use a plate reader or similar to reduce hands-on time

              Alexander

              Comment

              • gabrieltw
                Member
                • Mar 2013
                • 24

                #8
                Quant-iT PicoGreen dsDNA Assay Kit can do the plate format

                Comment

                • thermophile
                  Senior Member
                  • Apr 2015
                  • 243

                  #9
                  Further reduce costs for picogreen (which is already much cheaper than qubit). You can do 50 ul reactions of picogreen (25 dilute sample, 25 standard diluted reagent) in 384 well plates, rather than 200ul reactions that manufacturer suggests.
                  Microbial ecologist, running a sequencing core. I have lots of strong opinions on how to survey communities, pretty sure some are even correct.

                  Comment

                  • SAV
                    Junior Member
                    • Nov 2012
                    • 3

                    #10
                    Please see this publication showing nextera XT being prepared in 2, 4 and 8 uL and results for single cell RNAseq being compared and showing no loss of complexity.


                    And this one having samples prepared using same method but not mentioned in detail in methods.

                    libraries were normalized and prepared in 4 uL total volume.

                    Comment

                    • Simone78
                      Senior Member
                      • Oct 2010
                      • 208

                      #11
                      Originally posted by docsascha View Post
                      Dear Simone

                      Thanks for this info, which robot are you using, and how do you deal with magnetic beats separation and shaking, we have seen here some problems on our way for cutting cost

                      Does anybody have some tips on how to use a more time saving method than Qbit for DNA measurement? We would like to use a plate reader or similar to reduce hands-on time

                      Alexander
                      I use a Tecan Freedom Evo 200 with a 384 head. To save time and avoid (decrease) cDNA losses I don not perform any ethanol wash after binding. I just remove the supernatant and elute the cDNA from the beads. Since I am resuspending the cDNA in 15 ul (and I dilute it even more for the library prep) the leftover salt from the RT/preampl does not affect the efficiency of the downstream reactions.

                      Comment

                      • Arthur45
                        Junior Member
                        • May 2017
                        • 7

                        #12
                        Originally posted by Simone78 View Post
                        I cut reaction volumes by a factor of 10 using a liquid handling robot and the Nextera kit (exactly the same protocol, no changes). Tagmentation is done in 2 ul, I add 0.5 ul NT (or 0.2% SDS) followed by 1 ul i5/i7 (1:5 dilution to avoid excess adaptors and save money) and 1.5 ul NPM. Input is 50-150 pg/rxn.
                        Even without a robot this is perfectly doable with a pipette.
                        I this way a 3000 USD kit will last for about 1000 samples instead of 96!

                        /Simone
                        Dear Simone and other Nextera users,

                        I was wondering how to go about the Tagmented DNA Clean-up in the Nextera protocol. It uses a Zymo spin column, which I guess will not be possible for a smaller volume (e.g. 2-10 ul). What do you do at this stage? Do you clean with beads? Skip the whole clean-up? Dilute the Tagmented DNA in buffer to make up for the original 50ul volume needed?

                        Thanks,

                        Arthur


                        Would it b

                        Comment

                        • Simone78
                          Senior Member
                          • Oct 2010
                          • 208

                          #13
                          Originally posted by Arthur45 View Post
                          Dear Simone and other Nextera users,

                          I was wondering how to go about the Tagmented DNA Clean-up in the Nextera protocol. It uses a Zymo spin column, which I guess will not be possible for a smaller volume (e.g. 2-10 ul). What do you do at this stage? Do you clean with beads? Skip the whole clean-up? Dilute the Tagmented DNA in buffer to make up for the original 50ul volume needed?

                          Thanks,

                          Arthur


                          Would it b

                          Hi Arthur,
                          no need to clean up the tagmented DNA before the final PCR. I do the tagmentation reaction in 2 ul, add 0.5 ul of NT buffer (or 0.2% SDS) and simply go on with the PCR mix (2.5 ul). 0.2% SDS will remove the Tn5 from the DNA and, after adding the PCR mix, will be diluted down and won´t interfere with the final PCR.
                          Only after PCR you will do a clean up and, if you want to multiplex tens or hundreds of cells, I would suggest to pool all of them and then do the cleanup, not the other way around.
                          Best,
                          Simone

                          Comment

                          • Arthur45
                            Junior Member
                            • May 2017
                            • 7

                            #14
                            Originally posted by Simone78 View Post
                            Hi Arthur,
                            no need to clean up the tagmented DNA before the final PCR. I do the tagmentation reaction in 2 ul, add 0.5 ul of NT buffer (or 0.2% SDS) and simply go on with the PCR mix (2.5 ul). 0.2% SDS will remove the Tn5 from the DNA and, after adding the PCR mix, will be diluted down and won´t interfere with the final PCR.
                            Only after PCR you will do a clean up and, if you want to multiplex tens or hundreds of cells, I would suggest to pool all of them and then do the cleanup, not the other way around.
                            Best,
                            Simone
                            Hi Simone,

                            thank you for the message - this is so much easier than expected!

                            You mention after the PCR I should do a bead clean-up. The original protocol mentions to add twice the beads to the PCR product :

                            4 Add 30 μl AMPure XP beads to NAP2.
                            5 Add 30 μl AMPure XP beads to NAP2.

                            To me, this ratio of 20:30+30 looks like a mistake. Say I have 11ul of PCR products, would I add twice 16.5ul?

                            I hope someone can help me out there.

                            Cheers,
                            Arthur

                            Comment

                            • Arthur45
                              Junior Member
                              • May 2017
                              • 7

                              #15
                              Hi Simone,

                              thank you for your answer. I experimented with scaling-down the Nextera protocol, with mixed results.

                              My main worry is that the transposome activity is not behaving as expected. See the TapeStation traces, before and after tagmentation. The 1kb peak has disappeared during tagmentation, which is good news (we are aiming for a 300bp library). But I fail to be convinced that there is any DNA left after tagmentation, looking at the TapeStation electropherogram.

                              What do you think about it?

                              Cheers,

                              arthur
                              Attached Files

                              Comment

                              Latest Articles

                              Collapse

                              • SEQadmin2
                                From Collection to Sequencing: Why Sample Preparation and Preservation Define Sequencing Data
                                by SEQadmin2


                                Data variability is still an issue in sequencing technologies despite the advances in reproducibility and accuracy of these platforms. But the problem does not originate in the sequencing itself, but in the previous steps, before the sample reaches the sequencer.


                                The first step is collection, followed by preservation and sample preparation for analysis. Most scientists overlook those steps, but not being careful might just be skewing the experiment’s results.
                                ...
                                06-02-2026, 10:05 AM
                              • SEQadmin2
                                Single-Cell Sequencing at an Inflection Point: Early Impacts of New Platforms and Emerging Trends
                                by SEQadmin2


                                With the launch of new single-cell sequencing platforms in 2026, the field stands at an exciting inflection point. This article surveys the most impactful advances in the field and discusses how they’re reshaping research in cancer, immunology, and beyond.


                                Introduction

                                Single-cell sequencing technologies have undergone remarkable advances over the past decade, transitioning from low-throughput experimental approaches to highly scalable platforms capable of...
                                05-22-2026, 06:42 AM
                              • SEQadmin2
                                Environmental Genomics in the Age of NGS: From Microbes to Conservation Strategies
                                by SEQadmin2

                                Studying ecosystems means dealing with complex, multi-species communities that are hard to observe at scale. This complexity, however, hides many important questions to be answered, from how biogeochemical cycles work and how climate change can affect species distribution to how conservation strategies can work best.


                                Genomics, particularly since the expansion of NGS, has transformed ecosystem ecology. By sequencing environmental DNA, we can now assess biodiversity without direct...
                                05-06-2026, 09:04 AM

                              ad_right_rmr

                              Collapse

                              News

                              Collapse

                              Topics Statistics Last Post
                              Started by SEQadmin2, Today, 08:59 AM
                              0 responses
                              4 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 06-02-2026, 12:03 PM
                              0 responses
                              21 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 06-02-2026, 11:40 AM
                              0 responses
                              14 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 05-28-2026, 11:40 AM
                              0 responses
                              29 views
                              0 reactions
                              Last Post SEQadmin2  
                              Working...