    myseq error "waiting for FPGA command CAMDBLSNAP"

    Hi,

    I would appreciate any help in regard to the following MySeq error:

    Timeout of 70.000 seconds exceeded while waiting for FPGA command "CAMDBLSNAP 390 390 188 188 34390 150" to complete.

    see image:



    The error emerges any time after the run is started and ruins the run. It first appeared a year ago and now the fifth time since.

    The only advice we are continually getting from Illumina support is that we should reboot our myseq prior to the run (we are doing that since the first time the error occurred) and that the error is "very unlikely to repeat".
    Unfortunately, the error is repeating and keeps ruining runs.

    The MySeq's warranty has expired.

    If anybody had experienced the problem himself or has any information regarding the error (what does the code mean?), we would be very happy to hear from you.

    Thank you,
    Lovro
  • GenoMax
    Senior Member
    • Feb 2008
    • 7142

    #2
    If your MiSeq is not under a maintenance contract then you are living dangerously (or you just don't use it enough to justify the contract).

    This error may be due to some sort of hardware issue that can only be fixed by a visit from an Illumina engineer. You may have to bite the bullet and pay for that visit.

    Comment

    • Jessica_L
      Senior Member
      • Feb 2010
      • 117

      #3
      What Genomax said.

      I'd add that if you don't have your MiSeq under a maintenance agreement, you may want to consider getting one first, as the service visit and repair can easily exceed the cost of a contract. There are usually some hoops to jump through, but that's probably the cheapest way to get your instrument fixed.

      An initial visit with an FSE might be a good idea, at least, and if you have a decent relationship with them, they could give you a ballpark idea of how much repairs might cost and how that would compare to the cost of a service agreement.

      Comment

      • Lovro
        Member
        • Mar 2015
        • 19

        #4
        Hi, thanks for the answers.
        We have considered the latter.
        However, I was more interested in finding out the meaning of the error code.

        We had a hardware problem with a moveable mirror part (bouncing while snapping into position) while the machine was still under warranty. It was replaced by illumina and might still be under warranty.
        "CAMDBLSNAP" sounds a lot like some part of the camera snapping into position, so I want to investigate the issue.

        The error codes descriptions might be impossible to get a hold of though...?

        tnx

        Comment

        • Jessica_L
          Senior Member
          • Feb 2010
          • 117

          #5
          Your best bet on that score is probably your field support from Illumina, for a MiSeq probably your FAS. The error codes aren't closely guarded secrets, but I'm not aware of a collection of codes with descriptions that's accessible in any kind of searchable format.

          Comment

          • misterc
            Member
            • Jan 2016
            • 46

            #6
            CAMDBLSNAP is an FPGA or cabling issue on one of the 2 USB cables to the cameras internally. You can try re-seating the USB cables to the cameras to prevent/reduce the frequency of these errors. The ones you want to re-seat are on the inside of the MiSeq, facing the front of the instrument, up on the top right (behind the monitor), readily accessible when you take the top skin/panel off.

            Comment

            • Lovro
              Member
              • Mar 2015
              • 19

              #7
              Thank you very much. I will mention this to their representative.

              Comment

              Latest Articles

              Collapse

              • SEQadmin2
                Advanced Sequencing Platforms Tackle Neuroscience’s Toughest Genomics Problems
                by SEQadmin2



                Genomics studies in neuroscience face a special challenge due to the brain’s complexity and scarcity of samples. Mapping changes in cell type and state using conventional next-generation sequencing methods remains challenging. Advances in technologies like single-cell sequencing, spatial transcriptomics, and long-read sequencing have opened the door to deeper studies of the brain and diseases like Alzheimer’s, amyotrophic lateral sclerosis (ALS), and schizophrenia.
                ...
                07-09-2026, 11:10 AM
              • SEQadmin2
                Cancer Drug Resistance: The Lingering Barrier to Rising Survival
                by SEQadmin2



                Cancer survival rates have significantly increased in the last few decades in the United States, reaching a combined 70% 5-year survival rate by 2021. Behind this number, there are years of research to find new therapies, drug targets, and early detection methods. But there is one core challenge that keeps slowing down these advances, and it’s about drug resistance.

                There is no single reason why many patients don’t respond to treatment as expected. Cancer is...
                07-08-2026, 05:17 AM
              • GATTACAT
                Reply to Nine Things a Sample Prep Scientist Thinks About Before Sequencing
                by GATTACAT
                Love this - good data definitely starts from good input, and poor input can only give relatively poor data. I particularly like the mention of Nanodrop/absorbance based methods for quantification. It's such a toss up if you'll get an accurate reading or what amounts to a randomly generated number, and a lot of library/sequencing related issues can be traced back to poor quant.
                07-01-2026, 11:43 AM

              ad_right_rmr

              Collapse

              News

              Collapse

              Topics Statistics Last Post
              Started by SEQadmin2, 07-13-2026, 10:26 AM
              0 responses
              20 views
              0 reactions
              Last Post SEQadmin2  
              Started by SEQadmin2, 07-09-2026, 10:04 AM
              0 responses
              31 views
              0 reactions
              Last Post SEQadmin2  
              Started by SEQadmin2, 07-08-2026, 10:08 AM
              0 responses
              20 views
              0 reactions
              Last Post SEQadmin2  
              Started by SEQadmin2, 07-07-2026, 11:05 AM
              0 responses
              34 views
              0 reactions
              Last Post SEQadmin2  
              Working...