    #1

    Miseq v3 600 cycle kit - poor read 1 quality

    Hi all,

    I recently completed a run using the Miseq v3 600 cycle kit.
    The read configurations were as follows:
    R1 = 300 cycles, R2 = 55 cycles, Index1 = 6 cycles.

    The library is an amplicon style construct of avg length ~550 bases and has a constant sequence at the beginning of Read 1 (~60-65 bases).

    The library was run with 20% PhiX spike-in. Custom primers were used, but these have been successfully used with v2 miseq kits as well as Nextseq kits without issue.

    Essentially, I need at least 250 bases on Read 1, whereas Read 2 need only be a max of 50-55 (no need for overlap on Read 1 and 2).

    I tried the v3 kit but the FastQC files show that the run was not particularly great for read 1. See attached slide for one of the libraries multiplexed in this run. Reads drop below Q30 pretty much within 100 bases.

    Does anyone have any insight here? After perusing through previous posts from a while back, I was thinking of maybe switching to the Miseq v2 300 cycle kit or the 500 kit. Do you think this is a good idea (e.g. going with long read for read 1 with 300 cycle kit)? Previous threads suggest that 500 kits have also been affected although their quality is still better than the 600 cycle kit.

    Appreciate your feedback!
    Attached Files
  • thermophile
    Senior Member
    • Apr 2015
    • 243

    #2
    Why pay for a 600 cycle kit but not use all of it?

    The qscores drop faster for v3 kits than v2. Regardless which kit, run the full thing so you have overlap to error correct the poor quality bases
    Microbial ecologist, running a sequencing core. I have lots of strong opinions on how to survey communities, pretty sure some are even correct.

    Comment

    • incognitos
      Junior Member
      • Sep 2015
      • 9

      #3
      Yes I see the waste of the reagents, but the 3'end of the library has a long stretch of Ts due to the polyA signal from mRNA. So going beyond 50 bases in the 3'end will result in unusable sequence anyways.

      I was wondering if anyone has any experience doing longer than 150 bases for read 1 with the miseq v2 300 cycle kit? I was considering an asymmetric run of 250 and 50 for R1 and R2, respectively.

      Comment

      • microgirl123
        Senior Member
        • Jun 2012
        • 199

        #4
        There's no reason you can't do 250x50 with a 300-cycle kit. You should end up with similar Q scores to Read 1 on a 500-cycle kit, which are okay at the end (not great, but okay).

        Comment

        Latest Articles

        Collapse

        • SEQadmin2
          Beyond CRISPR/Cas9: Understand, Choose, and Use the Right Genome Editing Tool
          by SEQadmin2



          CRISPR/Cas9 sparked the gene editing revolution for both research and therapeutics.1 But this system still showed severe issues that limited its applications. The most prominent were the heavy reliance on PAM sequences, delivery limitations, double-stranded breaks that prompt unintended edits and cell death, and editing inefficiency (both in targeting and in knock-in reliability).

          Despite this, “CRISPR helped turn genome editing from a specialized technique into
          ...
          07-31-2026, 11:01 AM
        • SEQadmin2
          Proteomic Platforms: How to Choose the Right Analytical Strategy to Improve Detection and Clinical Applications
          by SEQadmin2


          Proteomics platforms are evolving rapidly, with advances in mass spectrometry and affinity-based approaches expanding what researchers can detect and at what scale. As the field moves toward deeper proteome coverage and clinical applications, scientists face an increasingly complex landscape of tools. This article will explore how researchers are navigating these choices to find the right platform for their work.

          The systematic characterization of the human proteome has
          ...
          07-20-2026, 11:48 AM

        ad_right_rmr

        Collapse

        News

        Collapse

        Topics Statistics Last Post
        Started by SEQadmin2, 08-06-2026, 07:41 AM
        0 responses
        23 views
        0 reactions
        Last Post SEQadmin2  
        Started by SEQadmin2, 08-03-2026, 10:13 AM
        0 responses
        39 views
        0 reactions
        Last Post SEQadmin2  
        Started by SEQadmin2, 07-31-2026, 02:55 AM
        0 responses
        44 views
        0 reactions
        Last Post SEQadmin2  
        Started by SEQadmin2, 07-24-2026, 12:17 PM
        0 responses
        27 views
        0 reactions
        Last Post SEQadmin2  
        Working...