    Low QC scores?

    I have just set up my very first sequencing experiment on the Next-Seq 500. It is now 45 cycles into PE sequencing, however it seems that my QC scores are a bit low. I only have a % >= Q30%: 25.2%

    I have around:
    280 million reads in the QC 10 bin.
    25 million in the QC 20 bin
    8 million in the QC 25 bin
    65 million in QC 30 bin
    44 million in QC 35 bin.

    Is this a bad result? Apologies for my stupidness, just was wondering whether data could be salvaged from this?
  • ox_zoo
    Junior Member
    • Nov 2016
    • 5

    #2
    Yes, now it has finished and the QC is only 15.2.

    The machine also has a warning message on two of the lanes:

    camera disabled:failed to detect clusters


    Does this mean that the cameras are not working properly?

    Thanks

    Comment

    • GenoMax
      Senior Member
      • Feb 2008
      • 7142

      #3
      Probably. You should contact Illumina tech support. They can take a look at your sequencer remotely to diagnose the problem.

      If your instrument is not directly accessible over the network then you will need to send them some diagnostic files.

      Comment

      • ox_zoo
        Junior Member
        • Nov 2016
        • 5

        #4
        Thanks for your reply.

        Yes I will do that - unfortunately since its a weekend they cannot reply.

        I was thinking - if clusters are not formed in those two particular areas, would the cameras be disabled? My pooled library concentration was sufficient, so I did not anticipate any problems. Also, if the cluster density is low then I am not sure how QC would be affected..

        Many thanks
        Last edited by ox_zoo; 11-19-2016, 07:39 AM.

        Comment

        • GenoMax
          Senior Member
          • Feb 2008
          • 7142

          #5
          What was the cluster density for this run?

          If the run was borderline overclustered to begin with, it is possible that sequencer lost the ability to distinguish the clusters over time (they become fatter). There is also the possibility of this being some sort of a hardware issue.

          Comment

          • ox_zoo
            Junior Member
            • Nov 2016
            • 5

            #6
            The cluster density for the four lanes range from 0-250k/mm2.

            The lanes:
            Lane 1: 235 ±41
            Lane 2: 237 ±38
            Lane 3: 212 ±56
            Lane 4: 230 ±35

            I hope this helps.

            Comment

            • GenoMax
              Senior Member
              • Feb 2008
              • 7142

              #7
              This run is borderline overclustered (as I recollect for NextSeq optimal cluster density is 170-220 K cluster/mm^2).

              Since there is nothing to lose at this time contact tech support and see what they say. You may have to eat the cost on this one and re-run.

              Comment

              • bini
                Junior Member
                • Jul 2018
                • 2

                #8
                Hi, How was this error fixed?

                Comment

                Latest Articles

                Collapse

                • SEQadmin2
                  Advanced Sequencing Platforms Tackle Neuroscience’s Toughest Genomics Problems
                  by SEQadmin2



                  Genomics studies in neuroscience face a special challenge due to the brain’s complexity and scarcity of samples. Mapping changes in cell type and state using conventional next-generation sequencing methods remains challenging. Advances in technologies like single-cell sequencing, spatial transcriptomics, and long-read sequencing have opened the door to deeper studies of the brain and diseases like Alzheimer’s, amyotrophic lateral sclerosis (ALS), and schizophrenia.
                  ...
                  Yesterday, 11:10 AM
                • SEQadmin2
                  Cancer Drug Resistance: The Lingering Barrier to Rising Survival
                  by SEQadmin2



                  Cancer survival rates have significantly increased in the last few decades in the United States, reaching a combined 70% 5-year survival rate by 2021. Behind this number, there are years of research to find new therapies, drug targets, and early detection methods. But there is one core challenge that keeps slowing down these advances, and it’s about drug resistance.

                  There is no single reason why many patients don’t respond to treatment as expected. Cancer is...
                  07-08-2026, 05:17 AM
                • GATTACAT
                  Reply to Nine Things a Sample Prep Scientist Thinks About Before Sequencing
                  by GATTACAT
                  Love this - good data definitely starts from good input, and poor input can only give relatively poor data. I particularly like the mention of Nanodrop/absorbance based methods for quantification. It's such a toss up if you'll get an accurate reading or what amounts to a randomly generated number, and a lot of library/sequencing related issues can be traced back to poor quant.
                  07-01-2026, 11:43 AM

                ad_right_rmr

                Collapse

                News

                Collapse

                Topics Statistics Last Post
                Started by SEQadmin2, Yesterday, 10:04 AM
                0 responses
                10 views
                0 reactions
                Last Post SEQadmin2  
                Started by SEQadmin2, 07-08-2026, 10:08 AM
                0 responses
                7 views
                0 reactions
                Last Post SEQadmin2  
                Started by SEQadmin2, 07-07-2026, 11:05 AM
                0 responses
                14 views
                0 reactions
                Last Post SEQadmin2  
                Started by SEQadmin2, 07-02-2026, 11:08 AM
                0 responses
                31 views
                0 reactions
                Last Post SEQadmin2  
                Working...