Unconfigured Ad

Collapse
X
 
  • Time
  • Show
Clear All
new posts
  • Brian Bushnell
    Super Moderator
    • Jan 2014
    • 2709

    #91
    I'm not sure; I think it's probably fine for quantification unless there's some bias issue, which I have not looked into. I wouldn't want to use it for variant-calling, particularly because a lot of the errors seem like systematic errors that cannot be overcome simply by sequencing deeper. We do use it for multiplexed single cells, because the NextSeq platform has shown lower rates of cross-talk than HiSeq or MiSeq and single-cell sequencing is greatly affected by even low levels of cross-talk. Also, I understand NextSeq is cheaper per base. But certainly, I would avoid the NextSeq (and HiSeq 3000/4000 which I suspect are similar) when possible, if you have access to Illumina's high quality platforms (HiSeq 2000/2500 or MiSeq).

    Comment

    • GenoMax
      Senior Member
      • Feb 2008
      • 7142

      #92
      Perhaps what you are observing is differences in bcl2fastq v.1.8.4 and 2.18.x?

      bcl2fastq v.2.x is required for processing data from NextSeq and HiSeq 3000/4000. It can be used to process data from all current Illumina sequencers. It does binned quality for reads as I recall.

      Is your data processed with the same version of bcl2fastq in all cases or was 2500 data processed using bcl2fastq v.1.8.4?

      Comment

      • Brian Bushnell
        Super Moderator
        • Jan 2014
        • 2709

        #93
        I'm not really sure. The HiSeq quality scores are not binned, though. I'm going to talk to the person who manages the Illumina software versions after gathering some more evidence, because we probably will want to roll back to an earlier version, once it's clear which earlier version was better.

        Also, does have experience with 3rd-party Illumina base-callers?

        Edit: We are using 2.16 for NextSeq and 1.8.4 for everything else.
        Last edited by Brian Bushnell; 11-17-2016, 01:20 PM.

        Comment

        • GenoMax
          Senior Member
          • Feb 2008
          • 7142

          #94
          Originally posted by Brian Bushnell View Post
          I'm not really sure. The HiSeq quality scores are not binned, though.
          That probably means they are using the older bcl2fastq (or CASAVA) v.1.8.4.

          I'm going to talk to the person who manages the Illumina software versions after gathering some more evidence, because we probably will want to roll back to an earlier version, once it's clear which earlier version was better.

          Also, does have experience with 3rd-party Illumina base-callers?
          That is NOT an option for NextSeq and HiSeq 3000/4000 which require bcl2fastq v.2.1x for conversion. Perhaps you can ask the person in charge to reprocess HiSeq 2500 data using bcl2fastq v.2.18.

          I don't know if there are any 3rd party callers for new data.

          Comment

          • AllSeq
            Registered Vendor
            • Oct 2013
            • 138

            #95
            Originally posted by Brian Bushnell View Post
            But certainly, I would avoid the NextSeq (and HiSeq 3000/4000 which I suspect are similar) when possible, if you have access to Illumina's high quality platforms (HiSeq 2000/2500 or MiSeq).
            Why would the NextSeq and HiSeq 3000/4000 be similar? The use different chemistries and different flow cells. Wouldn't the 3000/4000 be most similar to the HiSeq X? (Or did you just mean they're similar in that they're both bad platforms, but for different reasons?)
            AllSeq - The Sequencing Marketplace
            [email protected]
            www.AllSeq.com

            Comment

            • GenoMax
              Senior Member
              • Feb 2008
              • 7142

              #96
              Originally posted by AllSeq View Post
              Why would the NextSeq and HiSeq 3000/4000 be similar? The use different chemistries and different flow cells. Wouldn't the 3000/4000 be most similar to the HiSeq X? (Or did you just mean they're similar in that they're both bad platforms, but for different reasons?)
              I think it comes down to the bcl2fastq version used for data processing (binned q-scores) for NextSeq and HiSeq 4000.

              Hopefully @Brian will have some clarification once he has chased that down that information.

              Comment

              • Brian Bushnell
                Super Moderator
                • Jan 2014
                • 2709

                #97
                Originally posted by AllSeq View Post
                Why would the NextSeq and HiSeq 3000/4000 be similar? The use different chemistries and different flow cells. Wouldn't the 3000/4000 be most similar to the HiSeq X? (Or did you just mean they're similar in that they're both bad platforms, but for different reasons?)
                They use 2-color chemistry (IIRC). I don't know if the problem is the chemistry, the optics, or the software; but if it's the software, I'd expect the 3000/4000 to be more similar to the NextSeq than the 2500. Also, I've only looked at a single sample of HiSeq 4000 data, but the quality was low; similar to the NextSeq. Since I've seen both good and bad data from the same NextSeq machine, it's obviously possible to produce good data with 2-color chemistry and NextSeq optics. It would be nice if this was all a software issue.

                Comment

                • AllSeq
                  Registered Vendor
                  • Oct 2013
                  • 138

                  #98
                  The NextSeq uses 2 color chemistry, but the 3000/4000 uses the 'standard' 4 color chemistry (with patterned flow cells, just like the X).
                  AllSeq - The Sequencing Marketplace
                  [email protected]
                  www.AllSeq.com

                  Comment

                  • Brian Bushnell
                    Super Moderator
                    • Jan 2014
                    • 2709

                    #99
                    Originally posted by AllSeq View Post
                    The NextSeq uses 2 color chemistry, but the 3000/4000 uses the 'standard' 4 color chemistry (with patterned flow cells, just like the X).
                    Ah, my mistake.

                    Comment

                    • Michal2213
                      Junior Member
                      • Apr 2015
                      • 4

                      #100
                      NextSeq suitable for allele-specific analysis?

                      Do you think that NextSeq would be suitable for allele-specific analysis? I am using mouse cells with hybrid genome and sort the reads belonging to different alleles based on SNP content. So far I was using HiSeq2000 which worked well. With NextSeq I would get the data several times faster but having read this whole thread I am not sure whether the NextSeq data quality will be good enough.

                      Comment

                      • Brian Bushnell
                        Super Moderator
                        • Jan 2014
                        • 2709

                        #101
                        If the cost is reasonably similar, I'd go with HiSeq 2000. But perhaps you can get a sample of NextSeq data for your project, on a library you already ran on the HiSeq, to compare without committing yourself?

                        Comment

                        • Michal2213
                          Junior Member
                          • Apr 2015
                          • 4

                          #102
                          NextSeq suitable for allele-specific analysis?

                          Originally posted by Brian Bushnell View Post
                          If the cost is reasonably similar, I'd go with HiSeq 2000. But perhaps you can get a sample of NextSeq data for your project, on a library you already ran on the HiSeq, to compare without committing yourself?
                          Thank you for reply! I would do the same but with NextSeq we are getting data within a week as we have direct access and with HiSeq2000 in the sequencing core one need to wait in many cases more than a month. I guess we will need to go for sequencing the same library on both libraries as you suggested as well.

                          Comment

                          • dsobral
                            Member
                            • Jan 2012
                            • 23

                            #103
                            Just to add our 2 cents.

                            We have a MiSeq running since 2013, and after some hickups we're now stable with it and reasonably happy.

                            We just recently installed a NextSeq500 and our first tests are not great. Q30 is >80%, but there are many low quality bases (constantly Q=14 "/"), and the worst part is that instead of being towards the end, they seem a bit randomly distributed. When comparing PhiX in a 2x150bp NextSeq with a 2x250bp MiSeq, after alignment I see a 0.2-0.3% error rate with MiSeq and 0.9-1% error rate with NextSeq (1M sampled reads). In the "randomly" distributed Q=14 bases I seem to notice more A to T transitions, but I didn't have time to gather more systematic statistics... If I do quality trim on the MiSeq I can easily get higher quality data, with the NextSeq since its randomly distributed is harder...

                            We've complained to the Illumina people, let's see what they say...

                            Comment

                            • Brian Bushnell
                              Super Moderator
                              • Jan 2014
                              • 2709

                              #104
                              Originally posted by dsobral View Post
                              We've complained to the Illumina people, let's see what they say...
                              I don't believe that Illumina officially admits that NextSeq is lower quality than HiSeq 2500 or MiSeq. Error rate is not part of the platform specification, just quality scores.

                              Comment

                              • alexhaj
                                Junior Member
                                • Mar 2015
                                • 6

                                #105
                                Good work Brian! I am skeptical about your claim that NextSeq has less crossover than HiSeq or Miseq however. Would you please provide some data to back this up. And if that is the case maybe it's simply because demultiplexing is being done by CASAVA v2 on NextSeq and CASAVA v1 on HiSeq and Miseq. What if you did demultiplexing yourself, taking into account Quality scores (which I assume is not typically done).

                                Comment

                                Latest Articles

                                Collapse

                                • SEQadmin2
                                  Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
                                  by SEQadmin2



                                  Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

                                  Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
                                  ...
                                  10-14-2026, 09:55 AM
                                • SEQadmin2
                                  New Genomics Technologies Take Aim at Long-Standing Limits
                                  by SEQadmin2


                                  Researchers using sequencing and genomics tools often have to make trade-offs. They can choose between speed or scale, short reads or long-range information, or targeted panels or a view of the whole transcriptome. New technologies that have been released this year are built to address those tough choices.

                                  We asked six companies the same four questions to learn about their latest products. The new technologies bring a lot to the table, including rethinking sequencing
                                  ...
                                  09-28-2026, 10:25 AM

                                ad_right_rmr

                                Collapse

                                News

                                Collapse

                                Topics Statistics Last Post
                                Started by SEQadmin2, 10-15-2026, 10:52 AM
                                0 responses
                                25 views
                                0 reactions
                                Last Post SEQadmin2  
                                Started by SEQadmin2, 10-14-2026, 05:39 AM
                                0 responses
                                31 views
                                0 reactions
                                Last Post SEQadmin2  
                                Started by SEQadmin2, 10-13-2026, 09:56 AM
                                0 responses
                                27 views
                                0 reactions
                                Last Post SEQadmin2  
                                Started by SEQadmin2, 10-09-2026, 11:14 AM
                                0 responses
                                28 views
                                0 reactions
                                Last Post SEQadmin2  
                                Working...