    Adapter Dimers after Size Selection (?)

    Hello all!

    Recently I ran a library on our lab's new Next Seq 500 and came across an interesting problem. While processing the data from the instrument, I noticed that one of my samples had an excess of 35 bp poly(N) reads...as in 40% (yikes!). From what I can tell this generally means adapter dimer, as the bcl2fastq tool masks adapters with "N" as well as reads below the default of 35 bases. None of my other samples had this problem.

    We had previously ran the samples before pooling on an Agilent TapeStation, and noticed this particular sample had a high concentration of adapter dimer. So we ran the library on a Blue Pippin to select for a bp range of 200-500, and rechecked with the TapeStation to confirm the 120 bp peak was gone. Which it was.

    So we proceeded with the run. And then I discovered this problem.

    I then ran some of what I had left of the size selected library on both a regular agarose gel and a denaturing gel. A ~120 bp band showed up on the denaturing gel only!

    My question is, has anyone run into this before or do they know if some secondary structure could be causing it? Where could this band be hiding?
  • jdk787
    josh kinman
    • Apr 2014
    • 72

    #2
    I've seen this with libraries that had a lot of dimer going into amplification and the library was overamplified.

    I think that when the PCR primers have been exhausted, the dimer strands will anneal to library strands at the adapter regions creating a half dimer half library bubble product. When these products are run on a non-denaturing gel they will migrate slower than a 120bp dimer and appear larger, but when you use a denaturing gel the dimer strand will run as expected.

    Also, the dimer strands will cluster more efficiently during sequencing which will cause them to be over-represented in the sequencing results.
    Josh Kinman

    Comment

    • pmiguel
      Senior Member
      • Aug 2008
      • 2328

      #3
      Wow mhapp95, that is awesome that you actually ran the denaturing gel. I always suspected that what jdk787 was describing, was what was going on, but I was too lazy to do what you did.

      Can you give details on your procedure for doing the denaturing gel? Seems like if you used the denaturing gel instead of your Blue Pippin, you would be able to completely remove adapter dimers from a library.

      --
      Phillip

      Comment

      • CoastGen
        Registered Vendor
        • Sep 2012
        • 3

        #4
        I know of a group that saw this years ago when working with miRNA preps. Native gels would show a single band after size selection, but 2 bands on denaturing gels. They actually sequenced it and found that is was indeed an adaptermer. In the native gel, it formed a double-stranded species with actual library fragments, which led to a hairpin structure with aberrant mobility that allowed this structure to migrate with the normal library in a native gel.

        Comment

        • Christine
          Junior Member
          • Dec 2008
          • 3

          #5
          We've seen the typical ssDNA-bubble products/daisy chain libraries many times. I used to ignore it until we encountered a library that was hiding enormous amounts of adaptor dimers in there.

          Here's an image of the original library we received (a ~600 bp amplicon):

          Click image for larger version

Name:	Before.png
Views:	1
Size:	29.9 KB
ID:	305261

          Almost every sequence was clearly a short primer dimer. We reasoned the same as others here: when the primers are limiting in the amplification, dimers could hybridize to the library fragment ends and escape clean-up.

          To fix this we re-amplified for a few cycles with fresh PCR reagents to denture these duplexes and make everything properly double-stranded, then we could clean up the dimers.

          Here's what the library looked like after that PCR test, prior to clean-up:

          Click image for larger version

Name:	After.png
Views:	1
Size:	25.9 KB
ID:	305262

          Comment

          Latest Articles

          Collapse

          • GATTACAT
            Reply to Nine Things a Sample Prep Scientist Thinks About Before Sequencing
            by GATTACAT
            Love this - good data definitely starts from good input, and poor input can only give relatively poor data. I particularly like the mention of Nanodrop/absorbance based methods for quantification. It's such a toss up if you'll get an accurate reading or what amounts to a randomly generated number, and a lot of library/sequencing related issues can be traced back to poor quant.
            07-01-2026, 11:43 AM
          • SEQadmin2
            Nine Things a Sample Prep Scientist Thinks About Before Sequencing
            by SEQadmin2


            I’m not a sequencing expert. I’m a purification scientist who uses NGS to evaluate workflows my group develops. With this perspective, we think about the sample first and the NGS workflow second. The sequencer is an exceptionally honest reporter, but it can only report on what you give it, so whether you get clean, interpretable data from an NGS workflow is largely determined before you begin.

            Here are nine questions we think about, in roughly the order they matter, before...
            06-18-2026, 07:11 AM

          ad_right_rmr

          Collapse

          News

          Collapse

          Topics Statistics Last Post
          Started by SEQadmin2, Yesterday, 11:08 AM
          0 responses
          6 views
          0 reactions
          Last Post SEQadmin2  
          Started by SEQadmin2, 06-30-2026, 05:37 AM
          0 responses
          11 views
          0 reactions
          Last Post SEQadmin2  
          Started by SEQadmin2, 06-26-2026, 11:10 AM
          0 responses
          19 views
          0 reactions
          Last Post SEQadmin2  
          Started by SEQadmin2, 06-17-2026, 06:09 AM
          0 responses
          53 views
          0 reactions
          Last Post SEQadmin2  
          Working...