  • MisUser
    Member
    • Jan 2017
    • 13

    #1

    Miseq read 2 bad quality

    Hi all,

    We did a Miseq run (read 1-48 cycles, read 2-50 cycles) on a custom made library. The read quality for read 1 is OK but the read 2 quality is very low. Since it is an amplicon library, the client added some stuffer nucleotides for color balancing.
    I attached some SAV screen captures since I don't have the FastQC data yet. Should we suspect an issue with the library construction or could it be faulty reagents?
    Thank you very much for your help.
    Attached Files
  • GenoMax
    Senior Member
    • Feb 2008
    • 7142

    #2
    I doubt it would be reagent related. Can you post cluster density and PF?

    In any case having Illumina tech support do a remote login to look at the run should rule out any hardware/software/reagent related issues in advance.

    Comment

    • MisUser
      Member
      • Jan 2017
      • 13

      #3
      Thank you for your quick reply,

      We had 1200k/mm2 but only 30% of them PF. After talking to the client, it may be possible that there is no A in the first 5 cycles of the read 2 but the library is well color balanced though. Could it be the reason why the quality drops dramatically during the 2nd read squencing?
      Last edited by MisUser; 01-20-2017, 08:46 AM.

      Comment

      • nucacidhunter
        Jafar Jabbari
        • Jan 2013
        • 1250

        #4
        One reason could be high cluster density. Actual density could have been higher than reported 1200 which is at high end. From SAV analysis Tab, %Base under Data By Cycle pane would show if the colour balance is an issue. Looking at density tab under flowcell Chart would indicate overclustering if cluster density does not decrease from bottom to flowcell top. Also looking at images will indicate overclustering issue.

        Comment

        • nucacidhunter
          Jafar Jabbari
          • Jan 2013
          • 1250

          #5
          Your library description also points to a custom prep. If library was not prepared by ligating commercial adapters then some adapters may not have full complement of sequencing primers and this also will contribute to low %PF and low Q scores.

          Comment

          • MisUser
            Member
            • Jan 2017
            • 13

            #6
            Originally posted by nucacidhunter View Post
            Your library description also points to a custom prep. If library was not prepared by ligating commercial adapters then some adapters may not have full complement of sequencing primers and this also will contribute to low %PF and low Q scores.
            Hi nucacidhunter,

            Thank you for your input. The client did not use a ligation process to prepare his libraries (instead he built them by PCR) but he used illumina's adapter sequences. We quantified them with qPCR (commercial kit with Illumina's primers) as usual. Since we got signal, I guess the adapter sequences are present.
            I joined the SAV screencapt (sorry for the bad quality). Indeed there seems to be a big issue with the color balancing...
            Attached Files

            Comment

            • GenoMax
              Senior Member
              • Feb 2008
              • 7142

              #7
              Was phiX (or some other genomic DNA) spiked-in to help normalize the nucleotide distribution somewhat? If not that would be one thing to try along with under-loading of the library. You can probably read the amplicon sequence from the posted plot

              Comment

              • nucacidhunter
                Jafar Jabbari
                • Jan 2013
                • 1250

                #8
                I think there are at least two contributing factors:

                1- PCR based addition of adapter is done in one or two step PCR protocols. The quality of oligos will affect both %PF and Q scores because any substitution or indel in oligos will affect sequencing primer binding and phasing/prephasing.

                2- The client staggering techniques does not seem very successful. As GenoMax has mentioned this type of library need to be clustered under 800 K/mm2 and %15 or more PhiX to be added to obtain high Q scores. For longer reads cluster density need to be reduced further.

                Comment

                • thermophile
                  Senior Member
                  • Apr 2015
                  • 243

                  #9
                  something strange is happening to read1 too. I sequence mostly low diversity libraries and agree with the lower density and phiX spike in. How long are these amplicons?
                  Microbial ecologist, running a sequencing core. I have lots of strong opinions on how to survey communities, pretty sure some are even correct.

                  Comment

                  Latest Articles

                  Collapse

                  • SEQadmin2
                    Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
                    by SEQadmin2



                    Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

                    Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
                    ...
                    10-14-2026, 09:55 AM
                  • SEQadmin2
                    New Genomics Technologies Take Aim at Long-Standing Limits
                    by SEQadmin2


                    Researchers using sequencing and genomics tools often have to make trade-offs. They can choose between speed or scale, short reads or long-range information, or targeted panels or a view of the whole transcriptome. New technologies that have been released this year are built to address those tough choices.

                    We asked six companies the same four questions to learn about their latest products. The new technologies bring a lot to the table, including rethinking sequencing
                    ...
                    09-28-2026, 10:25 AM

                  ad_right_rmr

                  Collapse

                  News

                  Collapse

                  Topics Statistics Last Post
                  Started by SEQadmin2, 10-15-2026, 10:52 AM
                  0 responses
                  17 views
                  0 reactions
                  Last Post SEQadmin2  
                  Started by SEQadmin2, 10-14-2026, 05:39 AM
                  0 responses
                  22 views
                  0 reactions
                  Last Post SEQadmin2  
                  Started by SEQadmin2, 10-13-2026, 09:56 AM
                  0 responses
                  22 views
                  0 reactions
                  Last Post SEQadmin2  
                  Started by SEQadmin2, 10-09-2026, 11:14 AM
                  0 responses
                  27 views
                  0 reactions
                  Last Post SEQadmin2  
                  Working...