Unconfigured Ad

Collapse
X
 
  • Time
  • Show
Clear All
new posts
  • SeqTroubles
    Member
    • Sep 2016
    • 20

    #1

    CTAB DNA extraction - more RNA than DNA

    Hi everyone,

    Looking to get additional perspective on an issue driving me crazy. I am currently troubleshooting some library prep issues and am doing multiple extractions of one sample.

    1 kit - MoBio ultraclean microbial DNA
    1 kit - Qiagen DNeasy blood and tissue
    1 CTAB + phenol chloroform

    After the extractions I QC with nanodrop and qubit HS.

    MoBio the results correlate nicely between the nanodrop and qubit. For the Qiagen and CTAB I get a tenfold difference in concentration.

    I have done a gel and it appears to be the presence of RNA. I have already incorporated RNase A into the protocols. I double checked the efficacy of our stocks and they are working fine.
    For the CTAB method the only thing I can think of is it might be contamination in the chloroform or the ethanol?

    Thanks
  • ATϟGC
    Member
    • Jun 2013
    • 56

    #2
    What type of samples are you extracting from? I ask this because your poor correlations could also be due to co-precipitants like polysaccharides that give you poor A260/230 values. Ideally, A260/230 for DNA should be between about 2.0-2.2. I find that the more my DNA extracts are outside this range, the poorer correlations I get between nanodrop and qubit. CTAB-chloroform sometimes gives me very poor A260-230 values from polysaccharide contamination.

    Comment

    • Olaf Blue
      Member
      • Nov 2010
      • 58

      #3
      Are you using Polyvinyl pyrollidone in any of the cleanup steps? Which plant/sample type are you extracting?

      Comment

      • SeqTroubles
        Member
        • Sep 2016
        • 20

        #4
        Hi Thanks for the replys. Im extracting DNA from a gram negative bacteria species - Cronobacter. The absorbance readings for all extraction done do far are within good QC ratios. Some of the A260/230 are >2.2.

        Comment

        • Olaf Blue
          Member
          • Nov 2010
          • 58

          #5
          OK. PVP shouldn't be necessary when extracting DNA from bacteria.

          Comment

          • ATϟGC
            Member
            • Jun 2013
            • 56

            #6
            Hi again,

            I usually see poor nanodrop-Qubit correlations when the A260/230 get below ~1.7. Here is a link to another thread discussing common contaminants and their effects on the spectra of samples in case it helps you determine what the cause is:

            Techniques and protocol discussions on sample preparation, library generation, methods and ideas

            Comment

            • nucacidhunter
              Jafar Jabbari
              • Jan 2013
              • 1250

              #7
              NanoDrop quantification is based on absorption at 260 nm. RNA, dsDNA, ssDNA and nucleotides all absorb this wavelength. Qubit quantifies only dsDNA (using dsDNA reagent) so the discrepancy is due to presence of other nucleic acids, nucleotides and some contaminants.

              If you add RNase with the buffer used for Resuspension then all the nucleotides from RNA degradation will be present resulting in higher read. The same applies to protocols that use RNase in a step that nucleotides are not removed in consequent steps. You can check following for effect of contaminates on NanoDrop readings:

              Comment

              • SeqTroubles
                Member
                • Sep 2016
                • 20

                #8
                So I did another extraction yesterday with fresh chloroform and its the same problem. I used a pH strip to check out my P/C mix and it looks to be a definite pH of 7 maybe even slightly lower. So Im going to order in fresh P/C with a tris buffer and see how it goes.
                Fingers crossed.

                Comment

                • huguesparri
                  Member
                  • May 2008
                  • 97

                  #9
                  Maybe I missed something but you said that you have "library prep issues".
                  Could you tell us what kind of protocol did you use to build your libraries?

                  Comment

                  Latest Articles

                  Collapse

                  • SEQadmin2
                    Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
                    by SEQadmin2



                    Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

                    Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
                    ...
                    10-14-2026, 09:55 AM
                  • SEQadmin2
                    New Genomics Technologies Take Aim at Long-Standing Limits
                    by SEQadmin2


                    Researchers using sequencing and genomics tools often have to make trade-offs. They can choose between speed or scale, short reads or long-range information, or targeted panels or a view of the whole transcriptome. New technologies that have been released this year are built to address those tough choices.

                    We asked six companies the same four questions to learn about their latest products. The new technologies bring a lot to the table, including rethinking sequencing
                    ...
                    09-28-2026, 10:25 AM

                  ad_right_rmr

                  Collapse

                  News

                  Collapse

                  Topics Statistics Last Post
                  Started by SEQadmin2, Yesterday, 11:13 AM
                  0 responses
                  16 views
                  0 reactions
                  Last Post SEQadmin2  
                  Started by SEQadmin2, Yesterday, 10:48 AM
                  0 responses
                  16 views
                  0 reactions
                  Last Post SEQadmin2  
                  Started by SEQadmin2, 10-15-2026, 10:52 AM
                  0 responses
                  35 views
                  0 reactions
                  Last Post SEQadmin2  
                  Started by SEQadmin2, 10-14-2026, 05:39 AM
                  0 responses
                  42 views
                  0 reactions
                  Last Post SEQadmin2  
                  Working...