    Hi all,

    I ran MiSeq using a 500 cycles kit.
    The value for cluster density was 310K/mm2 but if I'm going to look the thumbnail images, I think I have much more cluster than that.

    Illumina tech looked the results and says "you are overclusterized".
    But for me cluster density is OK when I've reviewed the images.

    I'm attaching here 4 images from cycle 1.

    Could you tell me your opinion?
    Thank you!!
    Soledad
    Attached Files
  • nucacidhunter
    Jafar Jabbari
    • Jan 2013
    • 1250

    #2
    To me reported cluster density is what I would expect from the images. The run has been underclustered for a library that looks high diversity although I am assuming that the base diversity of other cycles are similar to the first cycle.

    Comment

    • Brian Bushnell
      Super Moderator
      • Jan 2014
      • 2709

      #3
      What's the scale on the little yellow boxes?

      Comment

      • misterc
        Member
        • Jan 2016
        • 46

        #4
        That looks about right to me for 300-500K/mm2 raw density (and NOT overclustered). Those intensities look reasonable to me too for cycle 1 given the MiSeq ramps up exposure on the cameras at each cycle to counter the typical decreasing intensity plots on their instruments.

        Comment

        • soleulloa
          Member
          • Aug 2011
          • 34

          #5
          Ok! Thank you.
          I'm using 10 pM (from 4 nM) to get this result.
          What do you think I have to use now? 12 or 15 pM to get a good run?

          Comment

          • microgirl123
            Senior Member
            • Jun 2012
            • 199

            #6
            What's your percentage passing filter?

            Comment

            • soleulloa
              Member
              • Aug 2011
              • 34

              #7
              Hi,

              95.27% PF

              Comment

              • microgirl123
                Senior Member
                • Jun 2012
                • 199

                #8
                If 95% of them are passing filter, then you are definitely not overclustered. If you were overclustered your percentage passing filter would be low.

                Unfortunately, the relationship between loading concentration and clusters is not linear so it's always a guess. I find that getting good cluster density is the Achilles' heel of the Illumina Miseq. My guess would be to go with the 15 pM since your density was so low.

                Comment

                • pmiguel
                  Senior Member
                  • Aug 2008
                  • 2328

                  #9
                  Originally posted by Brian Bushnell View Post
                  What's the scale on the little yellow boxes?
                  I don't know, but when I looked into it in the past, it seemed that the size of a cluster was around 1um in diameter. But that may have been on a HiSeq. I'm not sure whether the scales would be different on a MiSeq.

                  Old-timers like @ECO would be more likely to know. (Alas, I don't think @ callouts actually work on SeqAnswers...)

                  --
                  Phillip

                  Comment

                  • GenoMax
                    Senior Member
                    • Feb 2008
                    • 7142

                    #10
                    Illumina has a note on optimizing cluster density available here.

                    @soledad: Have you checked what the images look like in the middle and at the end of run?
                    Last edited by GenoMax; 02-17-2017, 08:51 AM.

                    Comment

                    • soleulloa
                      Member
                      • Aug 2011
                      • 34

                      #11
                      Someone has quantified libraries before dilution? When you prepare 2 nM dilutions?
                      I've quantified before and after that final dilution now because this is the first run when I have this problem and I'd like to compare these results.
                      Always I was between 1000-1200 K/mm2 using a 300 cycles kit.
                      This time I did the same process... and the only variable is I'm using 500 cycles kit.
                      But the library concentration will be the same.. so I don't understand what happens.

                      Comment

                      • nucacidhunter
                        Jafar Jabbari
                        • Jan 2013
                        • 1250

                        #12
                        It is most likely overestimating library concentration if you have used the same type of library. qPCR is the best method for quantification, other methods that quantify dsDNA are inadequate for libraries that contain non-sequenceable dsDNA fragments.

                        Comment

                        Latest Articles

                        Collapse

                        • SEQadmin2
                          Nine Things a Sample Prep Scientist Thinks About Before Sequencing
                          by SEQadmin2


                          I’m not a sequencing expert. I’m a purification scientist who uses NGS to evaluate workflows my group develops. With this perspective, we think about the sample first and the NGS workflow second. The sequencer is an exceptionally honest reporter, but it can only report on what you give it, so whether you get clean, interpretable data from an NGS workflow is largely determined before you begin.

                          Here are nine questions we think about, in roughly the order they matter, before...
                          06-18-2026, 07:11 AM
                        • SEQadmin2
                          From Collection to Sequencing: Why Sample Preparation and Preservation Define Sequencing Data
                          by SEQadmin2


                          Data variability is still an issue in sequencing technologies despite the advances in reproducibility and accuracy of these platforms. But the problem does not originate in the sequencing itself, but in the previous steps, before the sample reaches the sequencer.


                          The first step is collection, followed by preservation and sample preparation for analysis. Most scientists overlook those steps, but not being careful might just be skewing the experiment’s results.
                          ...
                          06-02-2026, 10:05 AM

                        ad_right_rmr

                        Collapse

                        News

                        Collapse

                        Topics Statistics Last Post
                        Started by SEQadmin2, 06-26-2026, 11:10 AM
                        0 responses
                        15 views
                        0 reactions
                        Last Post SEQadmin2  
                        Started by SEQadmin2, 06-17-2026, 06:09 AM
                        0 responses
                        49 views
                        0 reactions
                        Last Post SEQadmin2  
                        Started by SEQadmin2, 06-09-2026, 11:58 AM
                        0 responses
                        107 views
                        0 reactions
                        Last Post SEQadmin2  
                        Started by SEQadmin2, 06-05-2026, 10:09 AM
                        0 responses
                        125 views
                        0 reactions
                        Last Post SEQadmin2  
                        Working...