    Reason for POOR per tile sequence quality

    We recently performed a sequencing run on the MiSeq sequencer using a V3 (600 cycle kit).

    With regards to the per tile quality plot (obtained when using the FastQC analysis software) for one of our samples, it seems as though there is a large area that is flagged in red for R2, indicating that those tiles had worse qualities.

    I was just wondering on possible reasons why this would happen and how it could be prevented in future runs?

    Thanks!
    Attached Files
  • Markiyan
    Senior Member
    • Sep 2010
    • 126

    #2
    Looks like there is a problem with imaging one of the flowcell surfaces...

    It looks like there is a problem with imaging one of the flowcell surfaces...

    Either focusing issue (Z axis lens mover in a need of some cleaning/re greasing/calibration) or lots of very small bubbles (if it is an upper surface) (but then they would affect the other surface with such width.

    Also check tile thumbnail images, and contact illumina if this instrument is on service contract.

    You can also try making movie from thumbnail jpg images using image magic and some scripting...

    Comment

    • GW_OK
      Senior Member
      • Sep 2009
      • 411

      #3
      2xxx tiles are on the bottom surface. I concur with the possibilities raised by Markiyan. Looks like it started somewhere around cycle 35. Definitely a call to tech support is in order.

      Comment

      Latest Articles

      Collapse

      • SEQadmin2
        Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
        by SEQadmin2



        Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

        Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
        ...
        10-14-2026, 09:55 AM
      • SEQadmin2
        New Genomics Technologies Take Aim at Long-Standing Limits
        by SEQadmin2


        Researchers using sequencing and genomics tools often have to make trade-offs. They can choose between speed or scale, short reads or long-range information, or targeted panels or a view of the whole transcriptome. New technologies that have been released this year are built to address those tough choices.

        We asked six companies the same four questions to learn about their latest products. The new technologies bring a lot to the table, including rethinking sequencing
        ...
        09-28-2026, 10:25 AM

      ad_right_rmr

      Collapse

      News

      Collapse

      Topics Statistics Last Post
      Started by SEQadmin2, Yesterday, 10:52 AM
      0 responses
      10 views
      0 reactions
      Last Post SEQadmin2  
      Started by SEQadmin2, 10-14-2026, 05:39 AM
      0 responses
      19 views
      0 reactions
      Last Post SEQadmin2  
      Started by SEQadmin2, 10-13-2026, 09:56 AM
      0 responses
      17 views
      0 reactions
      Last Post SEQadmin2  
      Started by SEQadmin2, 10-09-2026, 11:14 AM
      0 responses
      27 views
      0 reactions
      Last Post SEQadmin2  
      Working...