Unconfigured Ad

Collapse
X
 
  • Filter
  • Time
  • Show
Clear All
new posts
  • seb42
    Junior Member
    • Mar 2017
    • 2

    Degenerate sequencing primers

    Hello,

    We are using custom sequencing primers containing multiple degenerate bases for MiSeq sequencing of the V4 SSU region. We have included these degenerate bases to hit more taxa during the PCR steps. Since the degenerate bases are in the V4 primer region they are pretty close to the start of the sequenced region so might have a disproportional affect on sequencing.

    We have been getting poor quality scores for multiple libraries (40-60% reads pass) and only getting between 3-5 million reads back. We were wondering if anyone else has had any trouble with this and if so how they overcame the problem.

    We were thinking that due to the mix of primers the poor results could be due to low sequencing primer density. Has anyone seen results similar to ours due to low primer densities or mixing of multiple sequencing primers?

    Thank you for any advice you can give.
  • barrmur
    Member
    • Nov 2009
    • 26

    #2
    Originally posted by seb42 View Post
    Hello,

    We are using custom sequencing primers containing multiple degenerate bases for MiSeq sequencing of the V4 SSU region. We have included these degenerate bases to hit more taxa during the PCR steps. Since the degenerate bases are in the V4 primer region they are pretty close to the start of the sequenced region so might have a disproportional affect on sequencing.

    We have been getting poor quality scores for multiple libraries (40-60% reads pass) and only getting between 3-5 million reads back. We were wondering if anyone else has had any trouble with this and if so how they overcame the problem.

    We were thinking that due to the mix of primers the poor results could be due to low sequencing primer density. Has anyone seen results similar to ours due to low primer densities or mixing of multiple sequencing primers?

    Thank you for any advice you can give.
    You should make sure the sequencing run is been carried out with about 20% PhiX in order to increase the base diversity during the run. Sounds like the low diversity is leading to a number of the clusters being removed during filtering.

    Comment

    • nucacidhunter
      Jafar Jabbari
      • Jan 2013
      • 1250

      #3
      You have not mentioned what protocol you have used to prepare the libraries. But poor quality and low outputs are characteristics of libraries prepared based on Earth Microbiome Project protocols even though they include primer pads to increase Tm.

      In my experience two step PCR following Illumina's protocol gives better results and all library fragments are sequenced using standard Illumina primers that eliminates sequencing bias caused by degenerate sequencing primers.

      Edit: 20-30% PhiX spike in increases quality but output still would be lower than two step PCR method.
      Last edited by nucacidhunter; 03-19-2017, 11:44 PM.

      Comment

      • seb42
        Junior Member
        • Mar 2017
        • 2

        #4
        Thank you for your answers. We do have a phiX spike in our runs and we are using a modified version of the Earth Microbiome Project. Our primers are a bit different than the ones mentioned in that protocol however.

        Comment

        • nucacidhunter
          Jafar Jabbari
          • Jan 2013
          • 1250

          #5
          This is the limitation of using fusion primers. Using gel purified primers for both PCR steps and sequencing will improve performance by minimizing miss-synthesized oligos containing indels and mismatches.

          Comment

          • thermophile
            Senior Member
            • Apr 2015
            • 243

            #6
            check the tm of your primers. The reverse emp primer is too cool (62-65, should be min 65). If your degeneracies are even cooler you will likely not get good annealing. I've had decent luck with locking nucleotide sequencing primers from http://www.exiqon.com/lna-technology to raise the annealing temp of the custom sequencing primers.
            Microbial ecologist, running a sequencing core. I have lots of strong opinions on how to survey communities, pretty sure some are even correct.

            Comment

            • nucacidhunter
              Jafar Jabbari
              • Jan 2013
              • 1250

              #7
              I think in addition to Tm, sequencing primers has to be full length because any truncation from 3’ end will affect phasing/prephasing and therefore quality and PF%.

              Comment

              Latest Articles

              Collapse

              • GATTACAT
                Reply to Nine Things a Sample Prep Scientist Thinks About Before Sequencing
                by GATTACAT
                Love this - good data definitely starts from good input, and poor input can only give relatively poor data. I particularly like the mention of Nanodrop/absorbance based methods for quantification. It's such a toss up if you'll get an accurate reading or what amounts to a randomly generated number, and a lot of library/sequencing related issues can be traced back to poor quant.
                07-01-2026, 11:43 AM
              • SEQadmin2
                Nine Things a Sample Prep Scientist Thinks About Before Sequencing
                by SEQadmin2


                I’m not a sequencing expert. I’m a purification scientist who uses NGS to evaluate workflows my group develops. With this perspective, we think about the sample first and the NGS workflow second. The sequencer is an exceptionally honest reporter, but it can only report on what you give it, so whether you get clean, interpretable data from an NGS workflow is largely determined before you begin.

                Here are nine questions we think about, in roughly the order they matter, before...
                06-18-2026, 07:11 AM

              ad_right_rmr

              Collapse

              News

              Collapse

              Topics Statistics Last Post
              Started by SEQadmin2, 07-02-2026, 11:08 AM
              0 responses
              23 views
              0 reactions
              Last Post SEQadmin2  
              Started by SEQadmin2, 06-30-2026, 05:37 AM
              0 responses
              23 views
              0 reactions
              Last Post SEQadmin2  
              Started by SEQadmin2, 06-26-2026, 11:10 AM
              0 responses
              23 views
              0 reactions
              Last Post SEQadmin2  
              Started by SEQadmin2, 06-17-2026, 06:09 AM
              0 responses
              55 views
              0 reactions
              Last Post SEQadmin2  
              Working...