    Good index sequences for unique dual indexing?

    After the index-hopping scandal (see Illumina's white paper) it sounds like the field is moving in the direction of unique dual indexing, i.e. dual indexing where every index combination is unique, so that no two libraries have the same P7 index and no two libraries have thes same P5 index. The means that instead of 12 versions of the P7 adapter and 8 versions of the P5 adapter, you'd need 96 versions of P7 and 96 versions of P5. This is an expensive investment for homebrew library protocols.

    Does anyone know of experimentally validated index sequences for 96-plexity?
    It would be nice to keep the 2 x 8 nt index-read lengths so we don't have to steal cycles from the actual target sequence. I see Bioo has a set of 96 barcodes that are 8 nt long, so I guess you could just reuse those same 96 on each end. But 68 out of 96 have an A or C in the first position, and a whopping 90 out of 96 have a A/C in the last position. That seems like it would be a serious problem since A/C are read by the same-color laser and Illumina says you must have color balance in your barcode combinations, even on the older machines that do four imaging cycles.
  • kmcarr
    Senior Member
    • May 2008
    • 1181

    #2
    I have heard unconfirmed reports that both Kapa and Bioo are planning on offering sets of 96 dual-unique indexes.

    Comment

    • MU Core
      Member
      • Apr 2008
      • 60

      #3
      I've been told that Illumina will soon (~1-2 months) be releasing a kit with 96 dual-unique indexes.

      Comment

      • jlei_face
        Member
        • Oct 2014
        • 10

        #4
        As you have probably already seen, Illumina will be working with IDT to create these UDI's. I know that IDT already has 96 unique indices, 8 nt long, as I have used them in the past. Instead of waiting for Illumina to release their kit it might be worth contacting IDT directly...

        Comment

        • apredeus
          Senior Member
          • Jul 2012
          • 151

          #5
          Along the same lines, and rather than multiplying threads - are there any comprehensive reviews of all the indexing strategies available for Illumina?

          And what would be a most reliable indexing strategy for single-end sequencing on Hiseq3000/Hiseq4000? Some long custom indexes, maybe?

          I would appreciate any input or pointers to discussions on other sites or here.

          Comment

          • nucacidhunter
            Jafar Jabbari
            • Jan 2013
            • 1250

            #6
            I have not seen one. Generally it includes:
            1- Indexes included in adapter (i5 and i7) which is Illumina’s standard method
            2- Inline barcode where one can include a barcode sequence at the start of one or both of reads. Barcode can be fixed or variable length for different samples to increase sequence diversity if required.
            3- Combinatorial indexing which is using inline barcode and standard index for samples

            To minimise read miss-assignment due to index hopping one can dual index libraries with unique i5 and i7 sequences. Longer index will be inefficient in prevention or identification of index hopping.

            Comment

            • kmcarr
              Senior Member
              • May 2008
              • 1181

              #7
              Originally posted by apredeus View Post
              Along the same lines, and rather than multiplying threads - are there any comprehensive reviews of all the indexing strategies available for Illumina?

              And what would be a most reliable indexing strategy for single-end sequencing on Hiseq3000/Hiseq4000? Some long custom indexes, maybe?

              I would appreciate any input or pointers to discussions on other sites or here.
              You can use dual indexes (or dual unique indexes) even with single end reads. Here is a link to Illumina's index sequencing overview which describes all possibilities of indexing, sequencing and instruments.

              Comment

              • apredeus
                Senior Member
                • Jul 2012
                • 151

                #8
                Originally posted by kmcarr View Post
                You can use dual indexes (or dual unique indexes) even with single end reads. Here is a link to Illumina's index sequencing overview which describes all possibilities of indexing, sequencing and instruments.
                Thank you very much! This is extremely useful.

                Comment

                • Innovelty
                  Member
                  • Sep 2012
                  • 14

                  #9
                  Still a little confused

                  I'm a bit confused about how to accomplish the dual-indexing using the same index on each end. My workflow ligates an adapter, and then uses indexing primers in PCR to get flow-cell adapters and indices onto the molecules. When I'm single indexing, this involves a "universal forward primer" and a unique reverse primer with the index. I can't just use the reverse primer without the universal forward primer, can I? I won't get indices on both ends, will I?

                  This seems awfully elementary, I know, but it's hard to keep track of exactly what protocol people are talking about doing. I 100% understand using two unique index primers for your library, out of, say, Illumina's TruSeq dual-indexing kit. But I don't understand how people are suggesting using the same index on both ends. You've got to be using custom oligos, right, with a forward and reverse primer that contain the same index region?

                  Are some sequencing cores now stocking indexing primers like this that customers can purchase? Because I sure as heck cannot afford 96 pairs of custom Illumina primers.

                  Comment

                  • luc
                    Senior Member
                    • Dec 2010
                    • 469

                    #10
                    You will require barcoded primers for both the P5 and and the P7 end of the library molecules. The "universal forward primer" needs to be substituted with a P5 index primer.
                    Yep, you will require custom oligos.

                    Originally posted by Innovelty View Post
                    I'm a bit confused about how to accomplish the dual-indexing using the same index on each end. My workflow ligates an adapter, and then uses indexing primers in PCR to get flow-cell adapters and indices onto the molecules. When I'm single indexing, this involves a "universal forward primer" and a unique reverse primer with the index. I can't just use the reverse primer without the universal forward primer, can I? I won't get indices on both ends, will I?

                    This seems awfully elementary, I know, but it's hard to keep track of exactly what protocol people are talking about doing. I 100% understand using two unique index primers for your library, out of, say, Illumina's TruSeq dual-indexing kit. But I don't understand how people are suggesting using the same index on both ends. You've got to be using custom oligos, right, with a forward and reverse primer that contain the same index region?

                    Are some sequencing cores now stocking indexing primers like this that customers can purchase? Because I sure as heck cannot afford 96 pairs of custom Illumina primers.

                    Comment

                    • systembio
                      Junior Member
                      • Jan 2018
                      • 2

                      #11
                      Has anyone used IDT unique dual indexes on various Illumina sequencing platforms? Do UDIs behave differently on Miseq (v3 chemistry) vs HiseqX? Does a Miseq chemistry has a role to play in index representation?

                      Comment

                      • Innovelty
                        Member
                        • Sep 2012
                        • 14

                        #12
                        Thanks for bumping this thread up again, systembio.

                        From what I know about the MiSeq chemistry (and it was the first sequencing platform I used) the unique dual indices shouldn't be a *problem* for MiSeq. Libraries made with UDI will sequence just fine on a MiSeq. They're unecessary, because MiSeq does not have the index-hopping problem that the HiSeq does, because the MiSeq's cluster-formation process is different. But definitely doing a little MiSeq nano run of your libraries to make sure they're what you wanted would be fine, prior to a HiSeq run. (If I'm wrong about this, though, I am sure someone will correct me!)

                        To answer my own question from a few months ago, the barcodes must be unique at both ends, but they don't have to be "the same index" at each end. Just each p5 index can only be used once, and each p7 index can only be used once, so each library gets a unique PAIR of indices. (Since I was struggling with it awhile back, I think I might as well clarify it for anyone else.) We ended up sequencing at Berkeley's genomics core, and they sell plates of UDI. Very pleased with them.
                        Last edited by Innovelty; 01-17-2018, 11:49 AM. Reason: Needed to add clarification

                        Comment

                        • nucacidhunter
                          Jafar Jabbari
                          • Jan 2013
                          • 1250

                          #13
                          Originally posted by systembio View Post
                          Has anyone used IDT unique dual indexes on various Illumina sequencing platforms? Do UDIs behave differently on Miseq (v3 chemistry) vs HiseqX? Does a Miseq chemistry has a role to play in index representation?
                          They work fine on NovaSeq and HS4000 and should be fine on MiSeq too as Innovelty has mentioned.

                          I wonder if you have used adapters directly from IDT or Illumina branded plates from Illumina. It will be interesting and useful if you could describe the issue that you have seen.

                          Comment

                          • systembio
                            Junior Member
                            • Jan 2018
                            • 2

                            #14
                            Thanks Innovelty and Nucacidhunter for your replies!
                            The problem i see here is that the libraries containing UDIs when an aliquot is run on Miseq show different index representation than the one run on HiseqX!!
                            i.e UDI pair poorly performing on Miseq is actually the best one on HiseqX! Why so?
                            Last edited by systembio; 01-17-2018, 01:46 PM.

                            Comment

                            • Innovelty
                              Member
                              • Sep 2012
                              • 14

                              #15
                              I think it would be helpful to us to know where you got your indices from, as nucacidhunter mentioned. I don't know if it's possible, but perhaps the indices from Illumina are selected because they work very well on the HiSeq. Since the way that they cluster is different, I *could* see it being possible that certain sequences on the HiSeq perform better than on the MiSeq. But I really can't imagine why that would be so, unless certain parts of the sequence upstream of the flow-cell adapter are better for the recombinase enzyme that the HiSeq uses?

                              It could also be a function of the libraries themselves, and NOT the indices. Were your libraries all of uniform size distribution? I've observed that the MiSeq can be very sensitive to the presence of long (>800bp) inserts -- those molecules simply won't cluster as well as the shorter inserts. If the poor-on-MiSeq library had larger inserts than the other libraries, that could explain why it didn't form as many clusters on the MiSeq flowcell, but was able to form clusters well on the HiSeq.

                              I feel your frustration. Clustering is some kind of mystical voodoo to those of us who don't work at Genomics cores and don't have tons of experience with the way different libraries cluster on different platforms. I'm dealing with this myself right now -- wide spread of representation over 50+ libraries, only *partly* explained by an aggressive Pippin prep.

                              Comment

                              Latest Articles

                              Collapse

                              • SEQadmin2
                                From Collection to Sequencing: Why Sample Preparation and Preservation Define Sequencing Data
                                by SEQadmin2


                                Data variability is still an issue in sequencing technologies despite the advances in reproducibility and accuracy of these platforms. But the problem does not originate in the sequencing itself, but in the previous steps, before the sample reaches the sequencer.


                                The first step is collection, followed by preservation and sample preparation for analysis. Most scientists overlook those steps, but not being careful might just be skewing the experiment’s results.
                                ...
                                06-02-2026, 10:05 AM
                              • SEQadmin2
                                Single-Cell Sequencing at an Inflection Point: Early Impacts of New Platforms and Emerging Trends
                                by SEQadmin2


                                With the launch of new single-cell sequencing platforms in 2026, the field stands at an exciting inflection point. This article surveys the most impactful advances in the field and discusses how they’re reshaping research in cancer, immunology, and beyond.


                                Introduction

                                Single-cell sequencing technologies have undergone remarkable advances over the past decade, transitioning from low-throughput experimental approaches to highly scalable platforms capable of...
                                05-22-2026, 06:42 AM
                              • SEQadmin2
                                Environmental Genomics in the Age of NGS: From Microbes to Conservation Strategies
                                by SEQadmin2

                                Studying ecosystems means dealing with complex, multi-species communities that are hard to observe at scale. This complexity, however, hides many important questions to be answered, from how biogeochemical cycles work and how climate change can affect species distribution to how conservation strategies can work best.


                                Genomics, particularly since the expansion of NGS, has transformed ecosystem ecology. By sequencing environmental DNA, we can now assess biodiversity without direct...
                                05-06-2026, 09:04 AM

                              ad_right_rmr

                              Collapse

                              News

                              Collapse

                              Topics Statistics Last Post
                              Started by SEQadmin2, Yesterday, 08:59 AM
                              0 responses
                              13 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 06-02-2026, 12:03 PM
                              0 responses
                              22 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 06-02-2026, 11:40 AM
                              0 responses
                              19 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 05-28-2026, 11:40 AM
                              0 responses
                              31 views
                              0 reactions
                              Last Post SEQadmin2  
                              Working...