I have created cDNA libraries from Drosophila melanogaster mRNA using the KAPA mRNA HyperPrep kit. The size distribution on the Bioanalyzer looks good, as I was expecting a mean of ~320 bp. I attached a screenshot of an example trace for one of my samples: My question concerns the two peaks very close to the upper marker (left-hand side): Are these adapter dimers (probably not, given their small size), or left-over primers? If not, what could they be? And, would this pose a problem for sequencing, given how much of them is present compared to the cDNA fragments? Thanks for any help!
Unconfigured Ad
Collapse
X
-
Role/Importance of adapter contamination in cDNA libraries
Last edited by patkrat; 08-12-2017, 08:03 AM. -
-
-
I did re-run the chip and contacted KAPA tech support. The latter suggested making sure after the library amplification to be very stringent with the 1X clean-up; i.e. remove all supernatant, and esp. all EtOH, etc. Now, I did a whole new library prep on the same samples yesterday, and made absolutely sure to follow all those steps, but still see this extra peak at ~38-40 bp. The 50 bp peak seems to be primer-dimer bleed-through that tends to happen with this protocol apparently (even visible in KAPA's tech manual..), but I don't know what these 38-40 bp peaks are. Could they be a chip artefact? I have read that the Bioanalyzer can be off by up to 5 bp at times, but even if that was the case, this wouldn't explain my situation, since the concentration of the 35 bp peak (upper marker) is 125.00 pg/ul. Or maybe I have a false understanding of how this concentration is measured..Originally posted by nucacidhunter View Post1x bead clean up should have removed any fragment below 100 nt. I would suggest to rerun the Bioanalyzer to rule out possibility of issues with Chip run and contact KAPA tech support for their explanation.
Comment
-
-
If you are running these on a non-patterned-flowcell Illumina sequencer, ignore the primer peaks. If you are running them on a patterned-flowcell Illumina sequencer (HiSeq 3000/4000/X and NovaSeq) then do an extra ampure (or whatever small fragment clean-up you have available to you.)
--
Phillip
Comment
-
-
Hello Phillip,Originally posted by pmiguel View PostIf you are running these on a non-patterned-flowcell Illumina sequencer, ignore the primer peaks. If you are running them on a patterned-flowcell Illumina sequencer (HiSeq 3000/4000/X and NovaSeq) then do an extra ampure (or whatever small fragment clean-up you have available to you.)
--
Phillip
Can you explain your reasoning behind the difference between patterned and non-patterned flowcells?
I can see how primers that contain an index sequence might lead to more index hopping on patterned flowcells, but what would be the issue with these primers that don’t contain an index sequence?
I have noticed these primer peaks with Kapa kits before since they use a high concentration of PCR primers to prevent exhaustion, but I have never sequenced the libraries on a patterned flowcell. Now I’m curious to see if this is going to be an issue on the NovaSeq.
Thank You
Comment
-
-
Hi dylanfofylan,
Have a look at this video: https://www.youtube.com/watch?v=pfZp5Vgsbw0
Please note how the ends of the library molecules are double-stranded when settling onto the flowcell. Any PCR primers will compete with the oligos present in the reagents - and your PCR primers will not be blocked from extension.
Comment
-
-
First of all, how do you know KAPA primers don't contain the index sequence? They didn't specify that in the TDS I purused..Originally posted by dylanfofylan View PostHello Phillip,
Can you explain your reasoning behind the difference between patterned and non-patterned flowcells?
I can see how primers that contain an index sequence might lead to more index hopping on patterned flowcells, but what would be the issue with these primers that don’t contain an index sequence?
I have noticed these primer peaks with Kapa kits before since they use a high concentration of PCR primers to prevent exhaustion, but I have never sequenced the libraries on a patterned flowcell. Now I’m curious to see if this is going to be an issue on the NovaSeq.
Thank You
This is a good point. This article explains the issue well: http://www.molecularecologist.com/20...as-hiseq-4000/
The main problem, I gather, is that left-over primers are not washed away in patterned flow cells, and can then 'hijack' the DNA polymerase by functioning as a template for the extension of library fragments. However, I have three questions:
1. Is the main problem arising from this situation that primers containing the index can generate new library fragments with wrong indices, thereby resulting in misassignment of gene expression to samples when demultiplexing?
2. If so, primers that do not contain the index should not be a problem, correct?
3. If one uses libraries made with primers not containing the index (like KAPA primers), and one has primer carry-over as shown in my Bioanalyzer trace above, is there any disadvantage of using a patterned vs. a non-patterned flow cell? I can imaging that, even if index switching is not an issue in this case, because left-over primers are not washed away from the flow cell before clustering, they may sit around in the wells an reduce coverage?
Thanks,
PatrickLast edited by patkrat; 12-22-2017, 04:55 AM.
Comment
-
Latest Articles
Collapse
-
by GATTACATLove this - good data definitely starts from good input, and poor input can only give relatively poor data. I particularly like the mention of Nanodrop/absorbance based methods for quantification. It's such a toss up if you'll get an accurate reading or what amounts to a randomly generated number, and a lot of library/sequencing related issues can be traced back to poor quant.
-
Channel: Articles
07-01-2026, 11:43 AM -
-
by SEQadmin2
I’m not a sequencing expert. I’m a purification scientist who uses NGS to evaluate workflows my group develops. With this perspective, we think about the sample first and the NGS workflow second. The sequencer is an exceptionally honest reporter, but it can only report on what you give it, so whether you get clean, interpretable data from an NGS workflow is largely determined before you begin.
Here are nine questions we think about, in roughly the order they matter, before...-
Channel: Articles
-
ad_right_rmr
Collapse
News
Collapse
| Topics | Statistics | Last Post | ||
|---|---|---|---|---|
|
Started by SEQadmin2, Yesterday, 11:08 AM
|
0 responses
7 views
0 reactions
|
Last Post
by SEQadmin2
Yesterday, 11:08 AM
|
||
|
Started by SEQadmin2, 06-30-2026, 05:37 AM
|
0 responses
11 views
0 reactions
|
Last Post
by SEQadmin2
06-30-2026, 05:37 AM
|
||
|
Started by SEQadmin2, 06-26-2026, 11:10 AM
|
0 responses
19 views
0 reactions
|
Last Post
by SEQadmin2
06-26-2026, 11:10 AM
|
||
|
Whole-Genome Sequencing Traces Faroe Islands Ancestry to a North Atlantic Founder Population
by SEQadmin2
Started by SEQadmin2, 06-17-2026, 06:09 AM
|
0 responses
53 views
0 reactions
|
Last Post
by SEQadmin2
06-17-2026, 06:09 AM
|
Comment