I used 200 ng total RNA input and the KAPA HyperPrep Kit to generate cDNA libraries. Usually, our libraries look like Sample 1 in the attached pdf, but this time we saw two peaks and a trailing 'belly'. Does anyone have an idea what this could be? Of note, for this High-Sensitivity Bioanalyzer run, I left the cDNA samples at room temperature for 30 min, together with the Bioanalyzer reagents; I usually leave the samples on ice, hence I mention it; maybe this could be a reason? Otherwise, I am thinking of overamplification, i.e. too many cycles for the library amplification step. It would be great to get some feedback and advice on this.
Unconfigured Ad
Collapse
X
-
-
Hi patkrat,
Yes, the apparently longer products are likely "bubble products". It is hypothesized that bubble products form during PCR when the concentration of denatured products is high enough to allow strands from different library molecules to anneal at their adapters, leaving the two insert strands unannealed because they are not related. The floppy insert strands apparently retard the progress of these bubble product through the sieving matrix of the chip.
You can ignore the issue if you like. The bubble products will denature along with the normal double stranded products during denaturation and cluster normally. But if you need an average size for your library to calculate its concentration (eg, for qPCR), then you want to use the left peak to determine that.
--
Phillip
-
-
Thanks for your response, Phillip. In fact, I ran another HighSens chip with 1:10 dilution of the cDNA library and found that this completely resolved the issue. While you may be right about the bubble product, I think that the main issue in this case actually was simple overloading of the chip (which has a 10 ng max capacity); I sometimes do a quick photospec to get the library concentration before running it on the Bioanalyzer, but this time the concentration was way off. I checked with the Qubit and found that I actually had up to 30 ng/ul libraries.Originally posted by pmiguel View PostHi patkrat,
Yes, the apparently longer products are likely "bubble products". It is hypothesized that bubble products form during PCR when the concentration of denatured products is high enough to allow strands from different library molecules to anneal at their adapters, leaving the two insert strands unannealed because they are not related. The floppy insert strands apparently retard the progress of these bubble product through the sieving matrix of the chip.
You can ignore the issue if you like. The bubble products will denature along with the normal double stranded products during denaturation and cluster normally. But if you need an average size for your library to calculate its concentration (eg, for qPCR), then you want to use the left peak to determine that.
--
Phillip
Conclusion: Overloading of the chip causes a split peak, but 30 ng/ul also indicates too many library amp cycles, which I will reduce from 13 to 12 (or maybe even 11) next time.
Thanks again for your answer.
Comment
-
Latest Articles
Collapse
-
by SEQadmin2
Genomics studies in neuroscience face a special challenge due to the brain’s complexity and scarcity of samples. Mapping changes in cell type and state using conventional next-generation sequencing methods remains challenging. Advances in technologies like single-cell sequencing, spatial transcriptomics, and long-read sequencing have opened the door to deeper studies of the brain and diseases like Alzheimer’s, amyotrophic lateral sclerosis (ALS), and schizophrenia.
...-
Channel: Articles
07-09-2026, 11:10 AM -
-
by SEQadmin2
Cancer survival rates have significantly increased in the last few decades in the United States, reaching a combined 70% 5-year survival rate by 2021. Behind this number, there are years of research to find new therapies, drug targets, and early detection methods. But there is one core challenge that keeps slowing down these advances, and it’s about drug resistance.
There is no single reason why many patients don’t respond to treatment as expected. Cancer is...-
Channel: Articles
07-08-2026, 05:17 AM -
-
by GATTACATLove this - good data definitely starts from good input, and poor input can only give relatively poor data. I particularly like the mention of Nanodrop/absorbance based methods for quantification. It's such a toss up if you'll get an accurate reading or what amounts to a randomly generated number, and a lot of library/sequencing related issues can be traced back to poor quant.
-
Channel: Articles
07-01-2026, 11:43 AM -
ad_right_rmr
Collapse
News
Collapse
| Topics | Statistics | Last Post | ||
|---|---|---|---|---|
|
Started by SEQadmin2, Today, 10:26 AM
|
0 responses
9 views
0 reactions
|
Last Post
by SEQadmin2
Today, 10:26 AM
|
||
|
Started by SEQadmin2, 07-09-2026, 10:04 AM
|
0 responses
24 views
0 reactions
|
Last Post
by SEQadmin2
07-09-2026, 10:04 AM
|
||
|
Started by SEQadmin2, 07-08-2026, 10:08 AM
|
0 responses
16 views
0 reactions
|
Last Post
by SEQadmin2
07-08-2026, 10:08 AM
|
||
|
Started by SEQadmin2, 07-07-2026, 11:05 AM
|
0 responses
33 views
0 reactions
|
Last Post
by SEQadmin2
07-07-2026, 11:05 AM
|
Comment