  • kwalton
    Junior Member
    • Jan 2014
    • 5

    2500 vs 4000

    We are in a position to buy an Illumina. I have a lot of experience with the 2500 at my previous job, but have not run Illumina technology in ~2yrs. At my current position we are debating the 2500 vs. 4000. We predominantly run RNA-seq (with Kapa lib build), ChIP-seq, PCR-seq (low diversity), targeted seq, "weird" libraries, some exome seq and with the Illumina technology we will be running some WGS. Who has run samples on both platforms? What is your experience? How do the above mentioned libraries work on the 4000? I have read there are problems with the 4000 (AD, not tolerant of differing library sizes, does not tolerate low diversity, etc). Are these actual concerns in the field with anyone running the 4000? How many million reads per lane are people actually getting with the 4000 with the above library types, or any library that you are running? How are you spiking in phiX on the 4000? Do you see a problem with spiking in phiX? I realize this is a lot of questions and I appreciate any help you can give.
  • nucacidhunter
    Jafar Jabbari
    • Jan 2013
    • 1250

    #2
    The short answer is that with HiSeq 2500 you can run all of those libraries but not with 4000 and currently there is no long read (2x250) option either. For those limitation one gets 25% increased raw read but if library has not been prepared carefully, that 25% can be lost after processing and removing low quality reads and primer-dimers.

    Comment

    • Brian Bushnell
      Super Moderator
      • Jan 2014
      • 2709

      #3
      The 2500 is certainly a safer bet. The 4000 *may* reach the quality of the 2500 some day, but I think that's hardly a valid justification for committing >$1m in some direction. The 2500 is good *now*.

      Comment

      • luc
        Senior Member
        • Dec 2010
        • 469

        #4
        With libraries made by us, we are getting 370 million reads on average on the 4000. The quality suffers with longer insert sizes and the system is less tolerant to low complexity issues (it requires more spike in of high complexity libraries), but you can run basically anything and it is more tolerant to library concentration variations. If you are operating the sequencer you will certainly prefer the 4000.
        Last edited by luc; 02-26-2016, 07:07 PM.

        Comment

        • nucacidhunter
          Jafar Jabbari
          • Jan 2013
          • 1250

          #5
          @luc
          I wonder if you have done any bisulfite converted library either RRBS or WGBS on 4000.

          Comment

          • Brian Bushnell
            Super Moderator
            • Jan 2014
            • 2709

            #6
            Originally posted by luc View Post
            With libraries made by us, we are getting 370 million reads on average on the 4000. The quality suffers with longer insert sizes and the system is less tolerant to low complexity issues (it requires more spike in of high complexity libraries), but you can run basically anything and it is more tolerant to library concentration variations. If you are operating the sequencer you will certainly prefer the 4000.
            That's quite interesting. Have you done any analysis of the quality of output reads?

            Comment

            • GenoMax
              Senior Member
              • Feb 2008
              • 7142

              #7
              Originally posted by luc View Post
              With libraries made by us
              This is probably the most critical aspect.

              @kwalton: If you are going to make your own libraries then 4000 would work but if you expect to run libraries made by someone else then you may find 2500 a safer bet (especially if you have no control over the library quality/characteristics).

              @luc: Have you run any libraries that were not made by you?

              Comment

              • luc
                Senior Member
                • Dec 2010
                • 469

                #8
                Hi,
                we are running libraries made by customers, which often have very wide insert size ranges (e.g. the ugly Nextera libraries). For these the yields (while still better than 2500) and the quality scores tend to be lower. We have run epignome WGBS as well as RRBS libraries. For these it seems to be difficult to dose the PhiX - the amount of spike in has not a good correlation to the % of resulting PhiX reads - requiring a test lane. Likely due to some algorithmic problem the reverse reads of bisulfite converted lanes have low quality scores, but the reverse read sequences seem not to contain notably more errors (e.g. alignment rates are similar).
                Problematic are libraries that have sudden changes between very low complexity and high complexity regions.
                Last edited by luc; 02-29-2016, 04:27 PM.

                Comment

                • bilyl
                  Member
                  • Aug 2013
                  • 52

                  #9
                  Dumb question, but how is it possible that the HiSeq 4000 is not robust to low diversity? Isn't the whole point of patterned flow cells to get good registration of clusters, and then from there get reliable base calls? Low diversity was primarily a problem because of bad base calling when some channels get flooded with signal and you can't see any defined clusters.

                  Comment

                  • luc
                    Senior Member
                    • Dec 2010
                    • 469

                    #10
                    Sorry, I am as dumb. I did post this question several times to Illumina people but never got an answer.
                    Originally posted by bilyl View Post
                    Dumb question, but how is it possible that the HiSeq 4000 is not robust to low diversity? Isn't the whole point of patterned flow cells to get good registration of clusters, and then from there get reliable base calls? Low diversity was primarily a problem because of bad base calling when some channels get flooded with signal and you can't see any defined clusters.

                    Comment

                    • GenoMax
                      Senior Member
                      • Feb 2008
                      • 7142

                      #11
                      Speculating about a cause: it is possible that 4000 can't tolerate low diversity in the first 25 cycles when chastity filtering is active for each cluster. Even though the features are defined, having majority of them light up at the same time may interfere with RTA/chastity filtering.
                      Last edited by GenoMax; 03-01-2016, 05:13 PM.

                      Comment

                      • Genohub
                        Registered Vendor
                        • Mar 2013
                        • 210

                        #12
                        Highlights of current HiSeq 3000/4000 limitations: https://genohub.com/services/sequenc...seq-3000-4000/

                        Comment

                        • luc
                          Senior Member
                          • Dec 2010
                          • 469

                          #13
                          With regards to the Genohub page info:

                          As mentioned before - low complexity sequencing is possible with the HiSeq 4000 (although not recommended by Illumina).

                          Point 6 is misleading. What is this supposed to mean? - without any quantification or details? Starting in autumn there were (are?) reagent quality problems that showed up only the last 40 bases of PE150 reads (reagent quality problems tend to show up on all illumina platforms). For us these reagent problems fortunately seem to have disappeared. HiSeq 3000s and 4000s are identical with regards to parts and read qualities. How big is the install base of the HiSeq 3000 ? (perhaps 3 or 4 machines? - the vast majority are HiSeq 4000 sequencers).
                          Last edited by luc; 03-07-2016, 12:09 AM.

                          Comment

                          • huguesparri
                            Member
                            • May 2008
                            • 97

                            #14
                            As far as I know, some facility running a HS3000 is still facing reagent issues in february 2016.
                            And from what Illumina told them, they are still investigating on the source of this issue.

                            Comment

                            • bmangeat
                              Junior Member
                              • Jul 2015
                              • 2

                              #15
                              Does anyone have some information regarding how the HiSeq 3000/4000 compares with the NextSeq when sequencing low complexity libraries? Is it as bad, or intermediate between the NextSeq and HiSeq 2500?

                              Comment

                              Latest Articles

                              Collapse

                              • mylaser
                                Reply to Advanced Sequencing Platforms Tackle Neuroscience’s Toughest Genomics Problems
                                by mylaser
                                Kheloyar – Everything You Need to Know About Kheloyaar Login and Kheoyar Id
                                If you are looking for an online gaming platform that offers a user-friendly experience, Kheloyar has become a name that many users search for. Whether you're interested in creating a new account, accessing your dashboard through Kheloyaar Login, or learning how to obtain a Kheoyar Id, understanding the platform's features and account process is essential.
                                This guide explains everything you need to know about...
                                Yesterday, 01:13 AM
                              • SEQadmin2
                                Advanced Sequencing Platforms Tackle Neuroscience’s Toughest Genomics Problems
                                by SEQadmin2



                                Genomics studies in neuroscience face a special challenge due to the brain’s complexity and scarcity of samples. Mapping changes in cell type and state using conventional next-generation sequencing methods remains challenging. Advances in technologies like single-cell sequencing, spatial transcriptomics, and long-read sequencing have opened the door to deeper studies of the brain and diseases like Alzheimer’s, amyotrophic lateral sclerosis (ALS), and schizophrenia.
                                ...
                                07-09-2026, 11:10 AM
                              • SEQadmin2
                                Cancer Drug Resistance: The Lingering Barrier to Rising Survival
                                by SEQadmin2



                                Cancer survival rates have significantly increased in the last few decades in the United States, reaching a combined 70% 5-year survival rate by 2021. Behind this number, there are years of research to find new therapies, drug targets, and early detection methods. But there is one core challenge that keeps slowing down these advances, and it’s about drug resistance.

                                There is no single reason why many patients don’t respond to treatment as expected. Cancer is...
                                07-08-2026, 05:17 AM

                              ad_right_rmr

                              Collapse

                              News

                              Collapse

                              Topics Statistics Last Post
                              Started by SEQadmin2, 07-09-2026, 10:04 AM
                              0 responses
                              22 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 07-08-2026, 10:08 AM
                              0 responses
                              14 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 07-07-2026, 11:05 AM
                              0 responses
                              32 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 07-02-2026, 11:08 AM
                              0 responses
                              31 views
                              0 reactions
                              Last Post SEQadmin2  
                              Working...