    Junior Member
    • Oct 2017
    • 3

    Illumina MiSeq High cluster density, but low %PF problem

    Hello everyone.

    Please, I need a help with my MiSeq genome sequencing run. I am sequencing M. tuberculosis genomes using Nextera XT and a 500 cycles v2 cartridge/flow cell. I always obtained aproximately 950-1050 K/mm2 and a passing filter more than 85%; However, in my last run I obtained 1063k/mm2 (i think it is ok), BUT JUST 50,3% %PF... sincerely, I dont have any idea what happened. I know with overclustering there will be a lower %PF, but in think is not my case. I s the first time i have this issue.

    Thank in advance for your answers.
  • finswimmer
    Member
    • Oct 2016
    • 60

    #2
    Hello,

    take a look at the cluster pictures. If you overcluster to much, the MiSeq cannot detect all clusters correctly and reports a much smaller number.

    fin swimmer

    Comment

    • GenoMax
      Senior Member
      • Feb 2008
      • 7142

      #3
      1000K/mm2 is outside limit per sepc for V2 chemistry. While it is tempting to push cluster density (trying to get additional data) the fall in data yield/quality is precipitous when you go over a certain density (like you discovered). You would want to carefully look at the data you have to make sure there are no artifacts.

      Comment

      • provirus
        Junior Member
        • Oct 2017
        • 7

        #4
        Have a look at your intensities of the first 25 cycles (e.g. in the sequence analysis viewer) (which might not only be a problem for amplicon sequencing). Below a quote from a Illumina white paper:

        The %PF calculations involve the application of a chastity filter to
        each cluster. Chastity is defined as the ratio of the brightest base
        intensity divided by the sum of the brightest and second brightest
        base intensities. Clusters “pass filter” if no more than 1 base call has
        a chastity value below 0.6 in the first 25 cycles.

        Comment

        • setramis
          Junior Member
          • Oct 2017
          • 3

          #5
          Thanks for your answers ... I read all of it., I got surprised because at two previous runs the last year I got a cluster denstiy of 1200 and 1100k / mm2 aproximately (definetly an overclustering) but the passing filter was approx 80%. Then I recalculated my steps and I corrected that. But now, I think I did not get it too much (just 1069K/mm2), and the passing filter fell too much.

          Now I am suspecting for the machine, MISeq. Because other people used to have the same, even worse, results with passing filter fell until 15%.

          Now I taked a look for my intensities at the first 25 cycles (like "provirus" recomendation) and I notice some extrange: first, definetly i got overclustering; second, the "T" base shows a marked overclustering (see picture), and thats the same result for all the run. could there be a particular reason for such behavior?



          For now, given the picture intensities Will it decrease my input DNA from 20 ul to 18 ul?

          Thanks for your answers and time.

          Comment

          • setramis
            Junior Member
            • Oct 2017
            • 3

            #6
            whats i hapennig... all my replies never come out

            Comment

            • GenoMax
              Senior Member
              • Feb 2008
              • 7142

              #7
              Originally posted by setramis View Post
              whats i hapennig... all my replies never come out
              It looks like your replies were caught in the moderation queue. I just approved them.

              It appears that you did not attach any images though (if you thought you had). You can edit your post and attach them.

              Comment

              • V_Ang
                Junior Member
                • Mar 2025
                • 1

                #8
                Hello everyone

                We are currently experiencing a sequencing issue with our Miseq libraries (made with DNA prep Illumina kit). For several runs now, we've been getting very low % PF clusters (<to 5%) with a high cluster density limit (1000-1100 k/mm2). We think it's overclustering, by rerunning more diluted, we get better results but we don't understand where this overclustering comes from. We had 3 out of our 4 last runs which were overclustered. Our bioanalyzer profiles are ok, and our concentrations dosed to both Qubit and Tapestation are in the usual ranges. We've checked all our dilutions, which are OK too.

                I came across an old post by Setramis (quoted above) that talked about this same problem and looking at the images, I have the same observation: an overclustering of T and a %base in the SAV that indeed shows a very strong overrepresentation of T.
                Does anyone have any idea about the cause of this problem?

                Thank you very much for your help.



                Originally posted by setramis View Post
                Thanks for your answers ... I read all of it., I got surprised because at two previous runs the last year I got a cluster denstiy of 1200 and 1100k / mm2 aproximately (definetly an overclustering) but the passing filter was approx 80%. Then I recalculated my steps and I corrected that. But now, I think I did not get it too much (just 1069K/mm2), and the passing filter fell too much.

                Now I am suspecting for the machine, MISeq. Because other people used to have the same, even worse, results with passing filter fell until 15%.

                Now I taked a look for my intensities at the first 25 cycles (like "provirus" recomendation) and I notice some extrange: first, definetly i got overclustering; second, the "T" base shows a marked overclustering (see picture), and thats the same result for all the run. could there be a particular reason for such behavior?



                For now, given the picture intensities Will it decrease my input DNA from 20 ul to 18 ul?

                Thanks for your answers and time.

                Comment

                Latest Articles

                Collapse

                • SEQadmin2
                  Nine Things a Sample Prep Scientist Thinks About Before Sequencing
                  by SEQadmin2


                  I’m not a sequencing expert. I’m a purification scientist who uses NGS to evaluate workflows my group develops. With this perspective, we think about the sample first and the NGS workflow second. The sequencer is an exceptionally honest reporter, but it can only report on what you give it, so whether you get clean, interpretable data from an NGS workflow is largely determined before you begin.

                  Here are nine questions we think about, in roughly the order they matter, before...
                  06-18-2026, 07:11 AM
                • SEQadmin2
                  From Collection to Sequencing: Why Sample Preparation and Preservation Define Sequencing Data
                  by SEQadmin2


                  Data variability is still an issue in sequencing technologies despite the advances in reproducibility and accuracy of these platforms. But the problem does not originate in the sequencing itself, but in the previous steps, before the sample reaches the sequencer.


                  The first step is collection, followed by preservation and sample preparation for analysis. Most scientists overlook those steps, but not being careful might just be skewing the experiment’s results.
                  ...
                  06-02-2026, 10:05 AM

                ad_right_rmr

                Collapse

                News

                Collapse

                Topics Statistics Last Post
                Started by SEQadmin2, Yesterday, 05:37 AM
                0 responses
                5 views
                0 reactions
                Last Post SEQadmin2  
                Started by SEQadmin2, 06-26-2026, 11:10 AM
                0 responses
                16 views
                0 reactions
                Last Post SEQadmin2  
                Started by SEQadmin2, 06-17-2026, 06:09 AM
                0 responses
                50 views
                0 reactions
                Last Post SEQadmin2  
                Started by SEQadmin2, 06-09-2026, 11:58 AM
                0 responses
                110 views
                0 reactions
                Last Post SEQadmin2  
                Working...