    MiSeq v3 600

    Hi everyone,

    we have troubles with our MiSeq Runs. At about cycle 80 we have a massive decrease in Q30scores.In general the quality of the runs is not good. Q30 scores of 50 or lower. In addition, during my last two runs- the intensity does not increase . I am very new to this technique and have therefore attached some Plots from basespace.
    Maybe someone had the same problem? Or has an idea about what is going wrong- what are the possible causes? I went through the troubleshooting guides, but could find yet any answer.

    thanks :-)
    Attached Files
  • GenoMax
    Senior Member
    • Feb 2008
    • 7142

    #2
    What kind of libraries are these? Are you spiking in phiX and if so at what %? Do you expect to have low nucleotide diversity after cycle 80 in sequencing? This run is also borderline overclustered and that could be one the reasons for this.

    Comment

    • HESmith
      Senior Member
      • Oct 2009
      • 512

      #3
      The run is significantly over-clustered, as indicated by the low PF (only 68%; should be mid-80s to mid-90s) in combination with the high error rate (4.4%; should be <1%). Quantify your libraries by qPCR and load at the recommended concentration (actually, a bit lower for PE-300bp runs).

      Comment

      • thermophile
        Senior Member
        • Apr 2015
        • 243

        #4
        Are you sure your template is more than 80bp?
        Microbial ecologist, running a sequencing core. I have lots of strong opinions on how to survey communities, pretty sure some are even correct.

        Comment

        • MU Core
          Member
          • Apr 2008
          • 60

          #5
          Is there a possibility of adapter dimers? May need to perform an additional library clean-up.

          Comment

          • HESmith
            Senior Member
            • Oct 2009
            • 512

            #6
            Short templates/adapter dimers produce a drop-off in signal intensity, which @ATGCT doesn't observe (first thumbnail). And adapter dimers produce a diagnostic spiky signal, also not observed.

            Comment

            • nucacidhunter
              Jafar Jabbari
              • Jan 2013
              • 1250

              #7
              I wonder if you can post %Base from Data by cycle pan.

              Comment

              • ATGCT
                Junior Member
                • Oct 2017
                • 4

                #8
                Hi, thanks for your answers !!!
                We are doing Rad-seq libraries. Yes, I am quite sure, that my templates are bigger than 80bp- but we are checking this now using the bioanalyzer kit.
                We already contacted Illumina, and they said, that we should increase PhiX concentration and lower our library concetration (from 8pM)- I think because of low diversity. I attached now % Base- at first it looks like problem of low intensities- then I dont know what happens
                Attached Files

                Comment

                • ATGCT
                  Junior Member
                  • Oct 2017
                  • 4

                  #9
                  Ah yes @ GenoMax - we spike with 5% PhiX. I thin we should increase to 10%

                  Comment

                  • nucacidhunter
                    Jafar Jabbari
                    • Jan 2013
                    • 1250

                    #10
                    Two issues:
                    1- low sequence diversity at the start which has resulted in low PF%. This should not happen if following standard RAD-seq protocol

                    2- short inserts and adapter-dimers that start reading into adapter in cycles 80-120

                    Both can be fixed by more careful library prep.

                    Comment

                    • GenoMax
                      Senior Member
                      • Feb 2008
                      • 7142

                      #11
                      @ATGCT: You should take this data and use BBMap to estimate the insert size. You can do it in different ways. See this link for examples. You likely have very short inserts as @nucacidhunter has said. This will just confirm that observation.

                      Comment

                      • ATGCT
                        Junior Member
                        • Oct 2017
                        • 4

                        #12
                        Hi, thanks :-) I will have a look at it. But the insert size should be around 500 bp - I cut them out of a Agarose gel - so I have seen the size- unless something happens after gel extraction. We will have our Bioanalyzer results soon. Then I can also have a look at that. But I will try also BBMap.
                        Thanks for your suggestions :-) !!!

                        Comment

                        • GenoMax
                          Senior Member
                          • Feb 2008
                          • 7142

                          #13
                          Don't be surprised. Keep in mind that small inserts preferentially cluster so if they are in there they are attaching to flowcell first.

                          Comment

                          • thermophile
                            Senior Member
                            • Apr 2015
                            • 243

                            #14
                            your %base also suggest short fragments and reading into the flowcell.
                            Microbial ecologist, running a sequencing core. I have lots of strong opinions on how to survey communities, pretty sure some are even correct.

                            Comment

                            Latest Articles

                            Collapse

                            • SEQadmin2
                              Cancer Drug Resistance: The Lingering Barrier to Rising Survival
                              by SEQadmin2



                              Cancer survival rates have significantly increased in the last few decades in the United States, reaching a combined 70% 5-year survival rate by 2021. Behind this number, there are years of research to find new therapies, drug targets, and early detection methods. But there is one core challenge that keeps slowing down these advances, and it’s about drug resistance.

                              There is no single reason why many patients don’t respond to treatment as expected. Cancer is...
                              Today, 05:17 AM
                            • GATTACAT
                              Reply to Nine Things a Sample Prep Scientist Thinks About Before Sequencing
                              by GATTACAT
                              Love this - good data definitely starts from good input, and poor input can only give relatively poor data. I particularly like the mention of Nanodrop/absorbance based methods for quantification. It's such a toss up if you'll get an accurate reading or what amounts to a randomly generated number, and a lot of library/sequencing related issues can be traced back to poor quant.
                              07-01-2026, 11:43 AM
                            • SEQadmin2
                              Nine Things a Sample Prep Scientist Thinks About Before Sequencing
                              by SEQadmin2


                              I’m not a sequencing expert. I’m a purification scientist who uses NGS to evaluate workflows my group develops. With this perspective, we think about the sample first and the NGS workflow second. The sequencer is an exceptionally honest reporter, but it can only report on what you give it, so whether you get clean, interpretable data from an NGS workflow is largely determined before you begin.

                              Here are nine questions we think about, in roughly the order they matter, before...
                              06-18-2026, 07:11 AM

                            ad_right_rmr

                            Collapse

                            News

                            Collapse

                            Topics Statistics Last Post
                            Started by SEQadmin2, Today, 10:08 AM
                            0 responses
                            6 views
                            0 reactions
                            Last Post SEQadmin2  
                            Started by SEQadmin2, Yesterday, 11:05 AM
                            0 responses
                            8 views
                            0 reactions
                            Last Post SEQadmin2  
                            Started by SEQadmin2, 07-02-2026, 11:08 AM
                            0 responses
                            31 views
                            0 reactions
                            Last Post SEQadmin2  
                            Started by SEQadmin2, 06-30-2026, 05:37 AM
                            0 responses
                            28 views
                            0 reactions
                            Last Post SEQadmin2  
                            Working...