  • sinnafoch
    Junior Member
    • Sep 2013
    • 9

    #1

    Pooling and sequencing NexteraXT and TrueSeq PCR free libraries on a single lane

    Hi everyone,

    I have a question regarding the pooling of NexteraXT and TrueSeq PCR free libraries onto a single Illumina HighSeq lane. I am sharing a lane with a colleague and we have different needs: I have enough DNA so I would like to use PCR free libraries. My colleague on the other hand has very little DNA and wants to use NexteraXT preps. Each of use would like to use 1/2 of a lane for sequencing.

    The person at our sequencing facility told us that pooling PCRed and PCR-free libraries could be problematic because it is hard to balance molarities in such a way that the different library types each occupy 1/2 of a lane.

    Now we are wondering about the possible reasons behind this? Does anyone have an idea why this could happen. I have never seen this mentioned in a paper.

    To which extent is this really a problem? When aiming for a 50:50 lane split how far off could the output be? 60:40? 80:20?

    thank you all for your help!
  • GenoMax
    Senior Member
    • Feb 2008
    • 7142

    #2
    What are the expected insert sizes in both cases? Keep in mind that in general short insert libraries will preferentially cluster (you may need to adjust loading conc going in, taking that into account) and may compete out larger fragments.

    Comment

    • Qibao
      Junior Member
      • Jul 2010
      • 1

      #3
      It also depends on what instrument you are using. We tried it on HiSeq4000 before and it worked well. It uses 2-3nM for PCR based libraries and 1-2nM for PCR free libraries. We mixed the samples to 3:2 ratio and it gave us expected sequencing percentage.

      Comment

      • Genetic Librarian
        Member
        • May 2017
        • 31

        #4
        well, mixing two libraries is a tricky task in general, no matter if PCR-free or not.
        It works well with libraries of the same fragment size. If the two libraries are of different sizes, there is a certain amount of educated guessing to counteract the bias for smaller fragments.

        For your particular issue, you need to make sure that both libraries are quantified using the same methods. For PCR free libraries, qPCR is needed, whereas for Nextera XT you are fine with just Qubit. So in this case, both should be measured with qPCR.

        Comment

        • sinnafoch
          Junior Member
          • Sep 2013
          • 9

          #5
          Thanks everyone! That the libraries probably need to be quantified with the same method is a good hint, we have also thought that this may minimize problems in our case.
          We have not thought about insert size however, I don't really understand why this may be an issue.

          Anyways, thank you again for your replies, gives me something to think about

          Comment

          • Genetic Librarian
            Member
            • May 2017
            • 31

            #6
            Insert size is a problem because bias works in favour of smaller fragments.
            So if you mix a 300 bp library and a 500 bp library 50/50, you will end up with a ~70/30 read distribution.
            One reason is the normal PCR bias during bridge amplification.
            Another reason is the ExAmp clustering reaction specific to the HiSeq 3000/4000. In the ExAmp reaction, the reaction mix is extremely viscous. Apparently, this should limit DNA motility so that the chance that multiple fragments attach inside the same nanowell is reduced. Basically, the ExAmp reaction relies on the fact that library binding is much slower than library amplification. So, you could think of the reaction mix like of an agarose gel, where smaller fragments move faster. Thus, smaller libraries will populate and block a proportional higher number of nanowells on the patterned flowcell.

            Comment

            • sinnafoch
              Junior Member
              • Sep 2013
              • 9

              #7
              Ok thank you for the explanation. This makes it more clear.

              Comment

              • finswimmer
                Member
                • Oct 2016
                • 60

                #8
                Originally posted by sinnafoch View Post
                We have not thought about insert size however, I don't really understand why this may be an issue.
                Beside the mentioned reason: If you have two librarys with the same concentration but with different fragment size, this also means, that the total number of fragments in the library with the lower fragment size is higher. And every fragment is apotential read.

                fin swimmer

                Comment

                • Genetic Librarian
                  Member
                  • May 2017
                  • 31

                  #9
                  Originally posted by finswimmer View Post
                  Beside the mentioned reason: If you have two librarys with the same concentration but with different fragment size, this also means, that the total number of fragments in the library with the lower fragment size is higher. And every fragment is apotential read.
                  That should not be the case, because you pool equimolar.

                  Comment

                  • finswimmer
                    Member
                    • Oct 2016
                    • 60

                    #10
                    Originally posted by Genetic Librarian View Post
                    That should not be the case, because you pool equimolar.
                    You're right. I didn't have a coffee at the time of writing

                    fin swimmer

                    Comment

                    Latest Articles

                    Collapse

                    • SEQadmin2
                      Proteomic Platforms: How to Choose the Right Analytical Strategy to Improve Detection and Clinical Applications
                      by SEQadmin2


                      Proteomics platforms are evolving rapidly, with advances in mass spectrometry and affinity-based approaches expanding what researchers can detect and at what scale. As the field moves toward deeper proteome coverage and clinical applications, scientists face an increasingly complex landscape of tools. This article will explore how researchers are navigating these choices to find the right platform for their work.

                      The systematic characterization of the human proteome has
                      ...
                      07-20-2026, 11:48 AM
                    • SEQadmin2
                      Advanced Sequencing Platforms Tackle Neuroscience’s Toughest Genomics Problems
                      by SEQadmin2



                      Genomics studies in neuroscience face a special challenge due to the brain’s complexity and scarcity of samples. Mapping changes in cell type and state using conventional next-generation sequencing methods remains challenging. Advances in technologies like single-cell sequencing, spatial transcriptomics, and long-read sequencing have opened the door to deeper studies of the brain and diseases like Alzheimer’s, amyotrophic lateral sclerosis (ALS), and schizophrenia.
                      ...
                      07-09-2026, 11:10 AM
                    • SEQadmin2
                      Cancer Drug Resistance: The Lingering Barrier to Rising Survival
                      by SEQadmin2



                      Cancer survival rates have significantly increased in the last few decades in the United States, reaching a combined 70% 5-year survival rate by 2021. Behind this number, there are years of research to find new therapies, drug targets, and early detection methods. But there is one core challenge that keeps slowing down these advances, and it’s about drug resistance.

                      There is no single reason why many patients don’t respond to treatment as expected. Cancer is...
                      07-08-2026, 05:17 AM

                    ad_right_rmr

                    Collapse

                    News

                    Collapse

                    Topics Statistics Last Post
                    Started by SEQadmin2, 07-24-2026, 12:17 PM
                    0 responses
                    10 views
                    0 reactions
                    Last Post SEQadmin2  
                    Started by SEQadmin2, 07-23-2026, 11:41 AM
                    0 responses
                    11 views
                    0 reactions
                    Last Post SEQadmin2  
                    Started by SEQadmin2, 07-20-2026, 11:10 AM
                    0 responses
                    23 views
                    0 reactions
                    Last Post SEQadmin2  
                    Started by SEQadmin2, 07-13-2026, 10:26 AM
                    0 responses
                    37 views
                    0 reactions
                    Last Post SEQadmin2  
                    Working...