    Illumina NextSeq non homogenous cluster density

    Hello. I am performing whole genome sequencing of bacterial strains and I got very variable cluster density (and hence signal intensity) over the four lanes of the flow cell. Has anyone ever seen this and can help me troubleshooting? Thanks for any suggestions you might want to give me.
    Attached Files
  • JoeKutch
    Member
    • Oct 2016
    • 17

    #2
    Intensity =/= cluster density, though they are related. There can be variations in intensity from imperfections in the flow cell, smudges on the surface, and focusing. Even reagent age/batch.

    The first image I've attached is from a low density (150K/mm^2) and the second is from a high density (280K/mm^2). Notice that the lower density run has a much higher intensity (see the scale on the right). Despite the variations in intensity, all four lanes from the first run were within 5% (<3million reads) of each other in terms of reads PF. A similar percentage variability was seen in the high density, lower intensity run.

    As long as there are no dramatic quality drops associated with the areas of low intensity, I wouldn't be concerned.
    Attached Files

    Comment

    • thiNGS
      Member
      • Sep 2014
      • 24

      #3
      Many thanks for your reply. I see this pattern with all my runs despite reagents and flow cell batches though. I wonder if my library concentration might be too high and so if I am "overloading" the flow cell. The problem is in lanes like 3 and 4 I lose quite a lot of reads because at least 15% won't pass the filter. Do you think overloading the flow cell might be a possible reason?

      Comment

      • JoeKutch
        Member
        • Oct 2016
        • 17

        #4
        Without knowing the cluster densities reported by your instrument, I couldn't say.

        What are the average densities for each of the 4 lanes, and the number of raw and PF reads for each lane?

        Our NextSeq generally maintains >90% PF with cluster densities up to 240K/mm^2, but the recommended spec by Illumina is 180-220K/mm^2. If you're outside of that range, I would recommend reducing your loading concentration proportionally based on how far over you are loading (i.e. if you're loading at what you believe to be 1.8pM, and getting densities of 270K/mm^2, try loading instead at 1.8 * (200/270) = 1.3 pM to target that 200K/mm^2 density.


        One other possibility - Is your phasing/prephasing higher in lanes 3 and 4? The fact that both lanes seem to trail off in intensity (and presumably density) indicates a possible delivery issue. If the phasing and prephasing for those two lanes is higher than L1, that would support the idea of a reagent delivery/fluidics problem. If this is the case, only a service engineer call can guarantee restored functionality, but you could try doing a few full washes with warm MilliQ water to flush any residue/clogs from the lines.

        Comment

        • misterc
          Member
          • Jan 2016
          • 46

          #5
          NextSeq fluidics lane pairs are odd vs. even (i.e. lanes 1 & 3 get SBS pumping while lanes 2 & 4 are imaged).

          Comment

          • JoeKutch
            Member
            • Oct 2016
            • 17

            #6
            Originally posted by misterc View Post
            NextSeq fluidics lane pairs are odd vs. even (i.e. lanes 1 & 3 get SBS pumping while lanes 2 & 4 are imaged).
            That's true, yes. But are lanes 1/3 and 2/4 treated as a single continuous lane with SBS flowing through 1 and then into 3, or does the line split and feed both of them? I always thought it was the latter, which doesn't preclude the individual lanes clogging. I could definitely be wrong about that though, I've only peeked inside our NextSeq during service calls, I haven't dug in it myself.

            Comment

            • misterc
              Member
              • Jan 2016
              • 46

              #7
              Good point. Yes, it's the latter (separate syringes pull the fluid through each of the two lanes simultaneously). It's entirely possible thiNGS has a fluidics issue on lane 1 there, but it seems like the fluidics on NextSeq are much more robust against clogs than it's HiSeq predecessor.

              Comment

              Latest Articles

              Collapse

              • SEQadmin2
                Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
                by SEQadmin2



                Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

                Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
                ...
                10-14-2026, 09:55 AM
              • SEQadmin2
                New Genomics Technologies Take Aim at Long-Standing Limits
                by SEQadmin2


                Researchers using sequencing and genomics tools often have to make trade-offs. They can choose between speed or scale, short reads or long-range information, or targeted panels or a view of the whole transcriptome. New technologies that have been released this year are built to address those tough choices.

                We asked six companies the same four questions to learn about their latest products. The new technologies bring a lot to the table, including rethinking sequencing
                ...
                09-28-2026, 10:25 AM

              ad_right_rmr

              Collapse

              News

              Collapse

              Topics Statistics Last Post
              Started by SEQadmin2, Yesterday, 11:13 AM
              0 responses
              17 views
              0 reactions
              Last Post SEQadmin2  
              Started by SEQadmin2, Yesterday, 10:48 AM
              0 responses
              18 views
              0 reactions
              Last Post SEQadmin2  
              Started by SEQadmin2, 10-15-2026, 10:52 AM
              0 responses
              36 views
              0 reactions
              Last Post SEQadmin2  
              Started by SEQadmin2, 10-14-2026, 05:39 AM
              0 responses
              42 views
              0 reactions
              Last Post SEQadmin2  
              Working...