    Miseq RTA crashed

    Hello all,

    We are running a 300v2 run on our miseq. Sadly, rta crashed during the first cycle. We tried restarting rta, but the process crashed over and over. Meanwhile, miseq continued to sequence and pile up raw data untill the D drive was completely full (200+ GB of raw data). Obviously the run stopped (at cycle 145) untill more space is cleared from d drive.

    As the machine stopped sequencing, awaiting space on d drive, the internal computer has become more responsive and re launching rta at this point does not seem to crash it. However rta does not process the data that is waiting in the queue, instead it pops out error messages.

    Does anyone know how we can continue the run? The first priority should be processing the data that is in the queueu now. As rta has crahed during cycle 1, no cluster densities have been reported ...

    Any help would be appreciated !
  • sklages
    Senior Member
    • May 2008
    • 628

    #2
    The first priority should be contacting Illumina's Tech support IMHO.

    Comment

    • kcchan
      Senior Member
      • Jul 2012
      • 186

      #3
      How are you restarting RTA? Are you just clicking on the .bat file in the run folder or are you making a new .bat file?

      Comment

      • kodbeeck
        Junior Member
        • Jan 2013
        • 4

        #4
        @ sklages: illlumina's tech support is not available during the weekend

        @ kcchan: just clicking the bat file indeed.

        this is my home account, from which i'm posting now
        Last edited by kodbeeck; 04-27-2013, 10:28 PM.

        Comment

        • kcchan
          Senior Member
          • Jul 2012
          • 186

          #5
          Can you double check the Windows task manager to make sure that RTA isn't open multiple times? If so you should close them all, then re-start a new instance.

          Edit: also, if you get RTA back up and running, pause the MiSeq for at least 15 minutes to give it a chance to process some the images. When the drive frees up a bit you can resume normal cycling.

          Comment

          • kodbeeck
            Junior Member
            • Jan 2013
            • 4

            #6
            There is just one RTA running.

            I've tried the 15 minute tip, but rta remains idle.

            Thanks anyway. Will contact Illumina tech support tomorrow.

            Comment

            • MadsAlbertsen
              Member
              • Aug 2010
              • 26

              #7
              Seems like we just got the same problem. Huge amounts of raw data filling up the drives without being processed.

              Did it work to just restart the RTA?

              rgds
              Mads Albertsen

              Comment

              • kodbeeck
                Junior Member
                • Jan 2013
                • 4

                #8
                We were not able to recover the run.

                On monday, following the weekend where we had the problem, I contacted tech support, after I tried restarting RTA for countless times which -for us- did not work.

                They took a look at the run, but in the end, miseq software terminated the run and we were directed to a post run wash.
                Illumina compensated all reagents for this run.


                Maybe you can try restarting RTA, it might help. In our case, the problem was that another RTA process from a previous run was still running in the background, corrupting the current run. This process was invisible for windows, so we could not shut it down. The simple fix was to restart miseq between each run.

                Good luck.

                Comment

                • MadsAlbertsen
                  Member
                  • Aug 2010
                  • 26

                  #9
                  OK. Thanks for the fast answer - we'll give it a go.

                  rgds
                  Mads

                  Comment

                  • MadsAlbertsen
                    Member
                    • Aug 2010
                    • 26

                    #10
                    It seem to have been a tile failing that caused the processing of the images to stop. The Illumina tech support was able to disable the specific tile and after a restart of RTA we seem to be up and running again.

                    rgds
                    Mads Albertsen

                    Comment

                    • kerryp
                      Member
                      • Apr 2015
                      • 17

                      #11
                      my RTA failed less than 50 cycles into a 600 cycle MiSeq run. I've been in contact with tech support and they've agreed to replace reagents. BUT because it stopped so early on, could I re-use the reagent cartridge and/or flow cell using the same pooled library? The flow cell will be clustered - if I bypass the barcode recognition, is there any way of skipping ahead to avoid upsetting the software, or could I just start from the start and see what happens? I've got nothing to lose but breaking the instrument (I don't want to do that).

                      Comment

                      Latest Articles

                      Collapse

                      • SEQadmin2
                        Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
                        by SEQadmin2



                        Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

                        Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
                        ...
                        10-14-2026, 09:55 AM
                      • SEQadmin2
                        New Genomics Technologies Take Aim at Long-Standing Limits
                        by SEQadmin2


                        Researchers using sequencing and genomics tools often have to make trade-offs. They can choose between speed or scale, short reads or long-range information, or targeted panels or a view of the whole transcriptome. New technologies that have been released this year are built to address those tough choices.

                        We asked six companies the same four questions to learn about their latest products. The new technologies bring a lot to the table, including rethinking sequencing
                        ...
                        09-28-2026, 10:25 AM

                      ad_right_rmr

                      Collapse

                      News

                      Collapse

                      Topics Statistics Last Post
                      Started by SEQadmin2, 10-22-2026, 11:13 AM
                      0 responses
                      26 views
                      0 reactions
                      Last Post SEQadmin2  
                      Started by SEQadmin2, 10-22-2026, 10:48 AM
                      0 responses
                      20 views
                      0 reactions
                      Last Post SEQadmin2  
                      Started by SEQadmin2, 10-15-2026, 10:52 AM
                      0 responses
                      36 views
                      0 reactions
                      Last Post SEQadmin2  
                      Started by SEQadmin2, 10-14-2026, 05:39 AM
                      0 responses
                      43 views
                      0 reactions
                      Last Post SEQadmin2  
                      Working...