    Agilent Bioanalyzer: Unexpected signal in 5S region

    Hi guys!

    I keep having this problem with zebrafish RNA isolated with Tri Reagent. On nanodrop everything seems to be OK (260/280 ratio ~2 and 260/230 ratio >2).

    Do you have any idea what can cause this unexpected signal and how to get rid of it? Can I expect this to do harm later on with RT-qPCR?

    Electropherogram seems quite nice but RIN values cannot be calculated.. (Hopefully managed to upload a picture)
    Attached Files
  • juelraja
    Junior Member
    • Aug 2013
    • 3

    #2
    Hi,

    I got rid of the unexpected signal in most samples by denaturating the samples in 70C for 2min (as suggested in protocol)...

    Comment

    • pmiguel
      Senior Member
      • Aug 2008
      • 2328

      #3
      Sometimes with Trizol samples you see enormous small RNA peaks that freak out the Agilent software. Nothing wrong with the samples though. They just produced a high yield of small RNA. In your case I don't see anything wrong with your chromatograms. That means the software produced an error that it should not have.

      What we do is find the parameter setting (in the pane on the right side of the the 2100 Expert software window) that deals with the threshold for giving an "unexpected signal in the 5S region" and increase it until it removes the error.

      Comment

      • juelraja
        Junior Member
        • Aug 2013
        • 3

        #4
        Thanks a lot for your answer pmiguel!

        Do you know if this observation is reported in literature? Is this related to zebrafish RNA or do you find similar problems with other samples also?

        -Juho

        Comment

        • pmiguel
          Senior Member
          • Aug 2008
          • 2328

          #5
          Originally posted by juelraja View Post
          Thanks a lot for your answer pmiguel!

          Do you know if this observation is reported in literature? Is this related to zebrafish RNA or do you find similar problems with other samples also?

          -Juho
          I don't recall any mention of it in the literature. It is a goofy interaction between Trizol (acid phenol) RNA preps and the Bioanalyzer 2100 expert software.

          Actually problems of this sort, Agilent's tech support is pretty good with. They can tell you which parameter you need to alter to get the error to go away.

          The other samples I saw this with, did happen to be fish, although not zebrafish. But I really think it is just a matter of the efficiency of the clean-up methodology retaining small RNAs. If it is good, you get giant small RNA peaks.

          That said, your small RNA peaks did not look large to me. So, I don't know what the deal was there.

          --
          Phillip

          Comment

          • mchekula
            Junior Member
            • Oct 2018
            • 7

            #6
            Hi, I have a large peak in 5S region, larger than rRNA peaks, but RIN calculated based on 28S/18S ratio looks OK (8-8.5). RNA isolated TRIfast reagent. Any idea what it could be? Thanks!
            Attached Files

            Comment

            • pmiguel
              Senior Member
              • Aug 2008
              • 2328

              #7
              Just looking at this ancient thread that mcheckula revived. The answer I gave in 2013 misses the actual issue here. The "signal in the fast region" is the actual 25 nt standard peak. The software has incorrectly identified peaks inside the 5S region as being the 25 nt standard. So the best fix is to set the real 25 nt peak as the lower size marker.
              --
              Phillip

              Comment

              • mchekula
                Junior Member
                • Oct 2018
                • 7

                #8
                In my case it's not 25nt peak though. Any opinions on the attached plot?

                Comment

                • pmiguel
                  Senior Member
                  • Aug 2008
                  • 2328

                  #9
                  Originally posted by mchekula View Post
                  Hi, I have a large peak in 5S region, larger than rRNA peaks, but RIN calculated based on 28S/18S ratio looks OK (8-8.5). RNA isolated TRIfast reagent. Any idea what it could be? Thanks!
                  Most likely it is a mixture of small RNAs -- 5S, tRNA, etc. "TRIfast" I have not heard of, but I presume it is another version of "Trizol", the commercial acid phenol reagents. For whatever reason, acid phenol methods seem to be extremely efficient at retaining low molecular weight RNAs that other methods mostly do not retain.

                  That said, I don't know that I've ever seen the 5S, etc. peak that large. So there might be something else going on.

                  --
                  Phillip

                  Comment

                  • mchekula
                    Junior Member
                    • Oct 2018
                    • 7

                    #10
                    Trifast is an analog of Trizol indeed. Yes, the height of the 5S/short RNA peak in comparision with rRNAs troubles me too - cant imagine any combination of naturally occuring short RNAs make such a large peak. But can it be degradation?

                    Comment

                    • cement_head
                      Senior Member
                      • Mar 2012
                      • 264

                      #11
                      Looks pretty good. Likely won't interfere with anything, including NGS. If you are concerned, you could try a bead clean-up (Promega ProNex), but I personally wouldn't.

                      Comment

                      • mchekula
                        Junior Member
                        • Oct 2018
                        • 7

                        #12
                        Originally posted by cement_head View Post
                        Looks pretty good. Likely won't interfere with anything, including NGS. If you are concerned, you could try a bead clean-up (Promega ProNex), but I personally wouldn't.
                        so you would fully rely on RIN? What do think is the source of the 5S peak, if not degradation?

                        Comment

                        • pmiguel
                          Senior Member
                          • Aug 2008
                          • 2328

                          #13
                          You would probably need to sequence it to find out what it is. But my guess would be lots of tRNA, or some other small RNA. Or something unexpected.

                          Since the rRNA peaks appear to be intact, it seems unlikely your RNA is degraded.

                          --
                          Phillip

                          Comment

                          • cement_head
                            Senior Member
                            • Mar 2012
                            • 264

                            #14
                            Originally posted by mchekula View Post
                            so you would fully rely on RIN? What do think is the source of the 5S peak, if not degradation?
                            No, I do not rely on RIN, but I analyse a lot of non-model critters. I usually go by 28S vs 18S peaks and the gel.

                            Comment

                            • mchekula
                              Junior Member
                              • Oct 2018
                              • 7

                              #15
                              Originally posted by pmiguel View Post
                              You would probably need to sequence it to find out what it is.
                              Phillip
                              good idea, thank you

                              Comment

                              Latest Articles

                              Collapse

                              • SEQadmin2
                                From Collection to Sequencing: Why Sample Preparation and Preservation Define Sequencing Data
                                by SEQadmin2


                                Data variability is still an issue in sequencing technologies despite the advances in reproducibility and accuracy of these platforms. But the problem does not originate in the sequencing itself, but in the previous steps, before the sample reaches the sequencer.


                                The first step is collection, followed by preservation and sample preparation for analysis. Most scientists overlook those steps, but not being careful might just be skewing the experiment’s results.
                                ...
                                06-02-2026, 10:05 AM
                              • SEQadmin2
                                Single-Cell Sequencing at an Inflection Point: Early Impacts of New Platforms and Emerging Trends
                                by SEQadmin2


                                With the launch of new single-cell sequencing platforms in 2026, the field stands at an exciting inflection point. This article surveys the most impactful advances in the field and discusses how they’re reshaping research in cancer, immunology, and beyond.


                                Introduction

                                Single-cell sequencing technologies have undergone remarkable advances over the past decade, transitioning from low-throughput experimental approaches to highly scalable platforms capable of...
                                05-22-2026, 06:42 AM
                              • SEQadmin2
                                Environmental Genomics in the Age of NGS: From Microbes to Conservation Strategies
                                by SEQadmin2

                                Studying ecosystems means dealing with complex, multi-species communities that are hard to observe at scale. This complexity, however, hides many important questions to be answered, from how biogeochemical cycles work and how climate change can affect species distribution to how conservation strategies can work best.


                                Genomics, particularly since the expansion of NGS, has transformed ecosystem ecology. By sequencing environmental DNA, we can now assess biodiversity without direct...
                                05-06-2026, 09:04 AM

                              ad_right_rmr

                              Collapse

                              News

                              Collapse

                              Topics Statistics Last Post
                              Started by SEQadmin2, Yesterday, 08:59 AM
                              0 responses
                              13 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 06-02-2026, 12:03 PM
                              0 responses
                              21 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 06-02-2026, 11:40 AM
                              0 responses
                              18 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 05-28-2026, 11:40 AM
                              0 responses
                              31 views
                              0 reactions
                              Last Post SEQadmin2  
                              Working...