    Increasing MiSeq v3 cycle number for >600bp reads

    Hello people!

    We have a sequencing application where we could really need, say, 50–100 more cycles from a 600-cycle MiSeq v3 kit.

    Illumina states that the cartridge contains reagents for 625 cycles [1] (to account for dual indices we don't need, so 10–15 more cycles there already).
    Then again, ECO got about 370 cycles from a 300 cycle kit [2] which is a fair bit more than the 325 stated by Illumina. Other people also mentioned that they pushed above the 25 reserve index cycles [3].

    Lastly, this post explains how MadsAlbertsen got 600 cycles from a 400 cycle kit by filling up some wells with reagents from leftover cartridges.

    So, my question: Does anybody has experience with increasing the cycle number on the 600-cycle v3 kits and can tell me what safely works? Either by experimenting how many possible cycles are in one cartrige or by filling up from an old cartridge.
  • Bukowski
    Senior Member
    • Jan 2010
    • 388

    #2
    People have been doing similar things on the HiSeq too - but the issue here isn't whether it's technically feasible to hack the kits, it's whether your application will be able to cope with the awful sequencing quality of anything over 300bp in length. If you routinely use the 600 cycle kits, you'll know how the R2 drop off is - so imagine what this is going to be like after 700 cycles...

    But see also: https://www.ncbi.nlm.nih.gov/pmc/articles/PMC3673215/

    Comment

    • JasperGeh
      Member
      • Sep 2018
      • 15

      #3
      Thanks, what they did in the paper looks really cool. And yes, I am aware of the decreasing quality, but a small amount of data (after filtering) helps us more than none, in this case...

      Comment

      • GenoMax
        Senior Member
        • Feb 2008
        • 7142

        #4
        Have you considered using nanopore sequencing if you need long(er) reads? No hacking of sequencer needed.

        Comment

        • JasperGeh
          Member
          • Sep 2018
          • 15

          #5
          Not for this, as it's an antibody library and we need a single paired read with low error rate to cover one fragment. Hacking it will probably be

          Comment

          • nucacidhunter
            Jafar Jabbari
            • Jan 2013
            • 1250

            #6
            If you do not need high number of reads, PacBio CCS will be a good option as it will cover the whole fragment length with very high accuracy.

            Comment

            Latest Articles

            Collapse

            • GATTACAT
              Reply to Nine Things a Sample Prep Scientist Thinks About Before Sequencing
              by GATTACAT
              Love this - good data definitely starts from good input, and poor input can only give relatively poor data. I particularly like the mention of Nanodrop/absorbance based methods for quantification. It's such a toss up if you'll get an accurate reading or what amounts to a randomly generated number, and a lot of library/sequencing related issues can be traced back to poor quant.
              07-01-2026, 11:43 AM
            • SEQadmin2
              Nine Things a Sample Prep Scientist Thinks About Before Sequencing
              by SEQadmin2


              I’m not a sequencing expert. I’m a purification scientist who uses NGS to evaluate workflows my group develops. With this perspective, we think about the sample first and the NGS workflow second. The sequencer is an exceptionally honest reporter, but it can only report on what you give it, so whether you get clean, interpretable data from an NGS workflow is largely determined before you begin.

              Here are nine questions we think about, in roughly the order they matter, before...
              06-18-2026, 07:11 AM

            ad_right_rmr

            Collapse

            News

            Collapse

            Topics Statistics Last Post
            Started by SEQadmin2, 07-02-2026, 11:08 AM
            0 responses
            14 views
            0 reactions
            Last Post SEQadmin2  
            Started by SEQadmin2, 06-30-2026, 05:37 AM
            0 responses
            15 views
            0 reactions
            Last Post SEQadmin2  
            Started by SEQadmin2, 06-26-2026, 11:10 AM
            0 responses
            20 views
            0 reactions
            Last Post SEQadmin2  
            Started by SEQadmin2, 06-17-2026, 06:09 AM
            0 responses
            54 views
            0 reactions
            Last Post SEQadmin2  
            Working...