    Hello all,

    I was prepping samples for WGS Using the Kapa HyperPrep workflow. After the final cleanup with 0.8x AMPure SPRI beads, my gel confirmation showed that most of my samples were lost.

    I'm trying to mentally troubleshoot what could have gone wrong, but I don't have extensive experience with library prepping to know where to look for the problem.

    I washed the samples+beads in 80% EtOH twice for 30s and eluted in 10mM TE+0.05% Tween20 buffer warmed to 55°C. I air dried the beads for 2 minutes between the wash and elution. I incubated in buffer for 3 minutes before removing the beads and used a plate shaker to assist with DNA re-suspension. The cleanup was done with a bead extractor to minimize sample loss.

    I redid the cleanup using pre-warmed 10mM Tris-HCl + 0.05% Tween20 instead of TE as the buffer, but the outcome was the same.

    Does anyone have suggestions on what could have happened to effect the affinity of the DNA and beads?

    Any ideas are welcome.
    Last edited by AvantS; 05-23-2019, 10:42 AM. Reason: Clarification
    ƧΛVΛПƬ
  • seqsuave
    Junior Member
    • May 2018
    • 9

    #2
    How fresh was the 80% Ethanol, if ethanol evaporated off you could have had elution during the washes. It is recommended that 80% Ethanol is not more than a week old.

    Comment

    • AvantS
      Junior Member
      • May 2019
      • 9

      #3
      @SeqSuave. It was made on the same day as the bead cleanup.
      ƧΛVΛПƬ

      Comment

      • itstrieu
        Member
        • Nov 2018
        • 27

        #4
        I am assuming you are doing the PCR-free workflow? What was the input concentration and the concentration after final clean-up. You might have to increase your input concentration or do a PCR step to amply your library.

        Also during the drying step if the beads are over-dried and you see hairline cracks, I would incubate the beads for 5 - 6 minutes to help increase yield.

        Comment

        • AvantS
          Junior Member
          • May 2019
          • 9

          #5
          @itstrieu The workflow includes 3 PCRs: End repair & A-tailing PCR, Adapter Ligation, and an Indexing. Input concentration was 100ng/uL per sample. Quantification after cleanup was not done, since the gel-confirmation showed few successful sample bands.
          ƧΛVΛПƬ

          Comment

          • itstrieu
            Member
            • Nov 2018
            • 27

            #6
            Is the workflow

            ER,AT, Ligation --> 0.8X bead cleanup --> Elute in 25 uL follow by Library Amplification --> Post-amplification Cleanup using 1x bead? With the concentration being 100ng/uL, how much total DNA is going into library prep?

            Was their a QC check after fragmentation? One common mistake I see is using elution buffer/water during the wash step instead of ethanol which could wash away your DNA. Not on purpose but accidental since both on them are clear liquids.
            Last edited by itstrieu; 05-24-2019, 10:54 AM.

            Comment

            • AvantS
              Junior Member
              • May 2019
              • 9

              #7
              Originally posted by itstrieu View Post
              Is the workflow

              ER,AT, Ligation --> 0.8X bead cleanup --> Elute in 25 uL follow by Library Amplification --> Post-amplification Cleanup using 1x bead? With the concentration being 100ng/uL, how much total DNA is going into library prep?

              Was their a QC check after fragmentation? One common mistake I see is using elution buffer/water during the wash step instead of ethanol which could wash away your DNA. Not on purpose but accidental since both on them are clear liquids.
              Our post-amp cleanup used 0.8x bead. The samples will be diluted to 20nM for sequencing.

              It's not possible that the elution buffer was used instead of EtOH because each U-bottom plate used for cleanup was labeled with the reagent it contains (Sample, EtOH W1, EtOH W2, Etc)and I added the warmed elution buffer to its plate while air drying the beads.
              ƧΛVΛПƬ

              Comment

              • itstrieu
                Member
                • Nov 2018
                • 27

                #8
                The only two things I can think of is DNA in buffers containing high concentrations of EDTA, other chelating agents or salts may inhibit the
                end repair and A-tailing reaction. The other is something going wrong with library amplification (maybe your index primers).

                If you can, check the concentration or run a gel/tape of the adapter-ligated library to see if the yield is reasonable going into library amplification.

                Comment

                • AvantS
                  Junior Member
                  • May 2019
                  • 9

                  #9
                  Originally posted by itstrieu View Post
                  The only two things I can think of is DNA in buffers containing high concentrations of EDTA, other chelating agents or salts may inhibit the
                  end repair and A-tailing reaction. The other is something going wrong with library amplification (maybe your index primers).

                  If you can, check the concentration or run a gel/tape of the adapter-ligated library to see if the yield is reasonable going into library amplification.
                  Thank you for the great advice. I'll run a gel confirmation to see if my adapter ligated libraries still have sample.
                  ƧΛVΛПƬ

                  Comment

                  • beancat
                    Junior Member
                    • Feb 2016
                    • 4

                    #10
                    I've used Kapa Hyperplus 3-4 times a week for well over a year and what causes me to decrease yield or lose samples have almost always been the the frag enzyme activity or sometimes beads.

                    Fragmentation: Under or over fragmentation will definitely mess up your yield. Make sure your kit is not expired, enzymes are always stored on ice and never vortexed, and buffers have no precipitation remaining. Kapa is also super temperature sensitive so maybe there was an issue there with frag enzyme activity? If i had to bet, I would say this is probably what happened to you.

                    Beads: Make sure the beads are well suspended and at the correct temperature. I personally elute in 10 mM Tris, ph 8.0 and think tween just gets in the way. I think its a good idea to never use TE for library preparation.

                    Comment

                    • AvantS
                      Junior Member
                      • May 2019
                      • 9

                      #11
                      Originally posted by beancat View Post
                      I've used Kapa Hyperplus 3-4 times a week for well over a year and what causes me to decrease yield or lose samples have almost always been the the frag enzyme activity or sometimes beads.

                      Fragmentation: Under or over fragmentation will definitely mess up your yield. Make sure your kit is not expired, enzymes are always stored on ice and never vortexed, and buffers have no precipitation remaining. Kapa is also super temperature sensitive so maybe there was an issue there with frag enzyme activity? If i had to bet, I would say this is probably what happened to you.

                      Beads: Make sure the beads are well suspended and at the correct temperature. I personally elute in 10 mM Tris, ph 8.0 and think tween just gets in the way. I think its a good idea to never use TE for library preparation.
                      That sounds very likely to me as well. My team and I will redo the pipeline keeping those suggestions in mind. Thank you.
                      ƧΛVΛПƬ

                      Comment

                      • jmc887
                        Junior Member
                        • Mar 2019
                        • 5

                        #12
                        Originally posted by beancat View Post
                        I've used Kapa Hyperplus 3-4 times a week for well over a year and what causes me to decrease yield or lose samples have almost always been the the frag enzyme activity or sometimes beads.

                        Fragmentation: Under or over fragmentation will definitely mess up your yield. Make sure your kit is not expired, enzymes are always stored on ice and never vortexed, and buffers have no precipitation remaining. Kapa is also super temperature sensitive so maybe there was an issue there with frag enzyme activity? If i had to bet, I would say this is probably what happened to you.

                        Beads: Make sure the beads are well suspended and at the correct temperature. I personally elute in 10 mM Tris, ph 8.0 and think tween just gets in the way. I think its a good idea to never use TE for library preparation.
                        Hi,

                        Interestingly I had a similar problem with the HyperPlus, my post can be found here:
                        Techniques and protocol discussions on sample preparation, library generation, methods and ideas


                        beancat, since you are an experienced user of KAPA Hyperplus I would like to ask you what was your average pre-capture yield? I got an average yield of 500 ng, however the protocol says we should get more than 1000 ng. I was also thinking I was loosing sample during the SPRI clean ups, I mean using the same 100ng input with other kits I get higher pre-capture yields.

                        In anycase I will be following this thread for any updates.
                        Thanks!

                        Comment

                        • beancat
                          Junior Member
                          • Feb 2016
                          • 4

                          #13
                          Originally posted by jmc887 View Post
                          Hi,

                          Interestingly I had a similar problem with the HyperPlus, my post can be found here:
                          Techniques and protocol discussions on sample preparation, library generation, methods and ideas


                          beancat, since you are an experienced user of KAPA Hyperplus I would like to ask you what was your average pre-capture yield? I got an average yield of 500 ng, however the protocol says we should get more than 1000 ng. I was also thinking I was loosing sample during the SPRI clean ups, I mean using the same 100ng input with other kits I get higher pre-capture yields.

                          In anycase I will be following this thread for any updates.
                          Thanks!
                          So I cut my reaction volume considerably with a lot of other deviations from the normal protocol (including PCR-free) so I don't think a direct comparison is useful. The only thing I can say is that it performs a lot better (nearly 25% better) than Qiagen's QIAseq FX kit with the same template input.
                          Last edited by beancat; 06-25-2019, 05:52 AM.

                          Comment

                          Latest Articles

                          Collapse

                          • SEQadmin2
                            Cancer Drug Resistance: The Lingering Barrier to Rising Survival
                            by SEQadmin2



                            Cancer survival rates have significantly increased in the last few decades in the United States, reaching a combined 70% 5-year survival rate by 2021. Behind this number, there are years of research to find new therapies, drug targets, and early detection methods. But there is one core challenge that keeps slowing down these advances, and it’s about drug resistance.

                            There is no single reason why many patients don’t respond to treatment as expected. Cancer is...
                            Today, 05:17 AM
                          • GATTACAT
                            Reply to Nine Things a Sample Prep Scientist Thinks About Before Sequencing
                            by GATTACAT
                            Love this - good data definitely starts from good input, and poor input can only give relatively poor data. I particularly like the mention of Nanodrop/absorbance based methods for quantification. It's such a toss up if you'll get an accurate reading or what amounts to a randomly generated number, and a lot of library/sequencing related issues can be traced back to poor quant.
                            07-01-2026, 11:43 AM
                          • SEQadmin2
                            Nine Things a Sample Prep Scientist Thinks About Before Sequencing
                            by SEQadmin2


                            I’m not a sequencing expert. I’m a purification scientist who uses NGS to evaluate workflows my group develops. With this perspective, we think about the sample first and the NGS workflow second. The sequencer is an exceptionally honest reporter, but it can only report on what you give it, so whether you get clean, interpretable data from an NGS workflow is largely determined before you begin.

                            Here are nine questions we think about, in roughly the order they matter, before...
                            06-18-2026, 07:11 AM

                          ad_right_rmr

                          Collapse

                          News

                          Collapse

                          Topics Statistics Last Post
                          Started by SEQadmin2, Yesterday, 11:05 AM
                          0 responses
                          7 views
                          0 reactions
                          Last Post SEQadmin2  
                          Started by SEQadmin2, 07-02-2026, 11:08 AM
                          0 responses
                          28 views
                          0 reactions
                          Last Post SEQadmin2  
                          Started by SEQadmin2, 06-30-2026, 05:37 AM
                          0 responses
                          28 views
                          0 reactions
                          Last Post SEQadmin2  
                          Started by SEQadmin2, 06-26-2026, 11:10 AM
                          0 responses
                          27 views
                          0 reactions
                          Last Post SEQadmin2  
                          Working...