    How to pool ATAC-seq libraries in equimolar concentrations?

    Hi! I have 24 ATAC libraries prepared for sequencing. For this, I want to pool them all together. I ran them on the TapeStation and I have the distinctive nucleosomal banding pattern (see attached files). I'm hesitating a bit on how to properly estimate the molarity of every library when I have multiple bands, since the Tape is not always detecting the same bands. Any suggestion about this? Should I just use Qubit quantification for pooling?
    Attached Files
  • kmcarr
    Senior Member
    • May 2008
    • 1181

    #2
    Originally posted by ECorrales View Post
    Hi! I have 24 ATAC libraries prepared for sequencing. For this, I want to pool them all together. I ran them on the TapeStation and I have the distinctive nucleosomal banding pattern (see attached files). I'm hesitating a bit on how to properly estimate the molarity of every library when I have multiple bands, since the Tape is not always detecting the same bands. Any suggestion about this? Should I just use Qubit quantification for pooling?
    Use Region Analysis on the TapeStation software instead of Peak Analysis to determine an average size for each sample. Set up a Region from ~150-1000bp on all samples and use that calculate the average size. Use this average size and Qubit measurements for each to then calculate a molar concentration for each library.

    Before you do this though, correct the Upper Marker in wells E1, A2, C2, F2 and G2 in the 10-24 file. The Upper Marker should be manually set tot he taller of the two peaks (Upper Marker peak height should alway be almost 2X the Lower Marker peak height.)

    Comment

    • ECorrales
      Junior Member
      • Nov 2019
      • 3

      #3
      Originally posted by kmcarr View Post
      Before you do this though, correct the Upper Marker in wells E1, A2, C2, F2 and G2 in the 10-24 file. The Upper Marker should be manually set tot he taller of the two peaks (Upper Marker peak height should alway be almost 2X the Lower Marker peak height.)
      Thank you for your reply. Would you say these peaks corresponds to the upper marker? I was assuming they were larger un-tagmented regions, so I was planning to do a second size selection round with Ampure beads to remove these peaks. But if these are the markers then a second size selection might be unnecessary.

      Comment

      • kmcarr
        Senior Member
        • May 2008
        • 1181

        #4
        Originally posted by ECorrales View Post
        Thank you for your reply. Would you say these peaks corresponds to the upper marker? I was assuming they were larger un-tagmented regions, so I was planning to do a second size selection round with Ampure beads to remove these peaks. But if these are the markers then a second size selection might be unnecessary.
        What I am saying is that for those 5 lanes I mentioned there are 2 very closely spaced peaks, one labeled as a peak of ~1,260-1,270bp and the peak just after that labeled as the Upper marker. The software has incorrectly identified the later migrating peak as the Upper marker. In the TapeStation analysis software you should select the peak which is labeled ~1,270 and manually set that as the upper marker. If you do not correct the upper marker then the size calculation for these lanes will be incorrect.

        Comment

        • ECorrales
          Junior Member
          • Nov 2019
          • 3

          #5
          Originally posted by kmcarr View Post
          What I am saying is that for those 5 lanes I mentioned there are 2 very closely spaced peaks, one labeled as a peak of ~1,260-1,270bp and the peak just after that labeled as the Upper marker. The software has incorrectly identified the later migrating peak as the Upper marker. In the TapeStation analysis software you should select the peak which is labeled ~1,270 and manually set that as the upper marker. If you do not correct the upper marker then the size calculation for these lanes will be incorrect.
          Oh, I see. Thank you so very much for clarifying it!

          Comment

          Latest Articles

          Collapse

          • SEQadmin2
            Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
            by SEQadmin2



            Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

            Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
            ...
            10-14-2026, 09:55 AM
          • SEQadmin2
            New Genomics Technologies Take Aim at Long-Standing Limits
            by SEQadmin2


            Researchers using sequencing and genomics tools often have to make trade-offs. They can choose between speed or scale, short reads or long-range information, or targeted panels or a view of the whole transcriptome. New technologies that have been released this year are built to address those tough choices.

            We asked six companies the same four questions to learn about their latest products. The new technologies bring a lot to the table, including rethinking sequencing
            ...
            09-28-2026, 10:25 AM

          ad_right_rmr

          Collapse

          News

          Collapse

          Topics Statistics Last Post
          Started by SEQadmin2, 10-15-2026, 10:52 AM
          0 responses
          17 views
          0 reactions
          Last Post SEQadmin2  
          Started by SEQadmin2, 10-14-2026, 05:39 AM
          0 responses
          26 views
          0 reactions
          Last Post SEQadmin2  
          Started by SEQadmin2, 10-13-2026, 09:56 AM
          0 responses
          22 views
          0 reactions
          Last Post SEQadmin2  
          Started by SEQadmin2, 10-09-2026, 11:14 AM
          0 responses
          27 views
          0 reactions
          Last Post SEQadmin2  
          Working...