  • Crawford505
    Junior Member
    • Jun 2009
    • 4

    Whole-genome amplification prior to Illumina sequencing?

    Hello,

    Does anyone have any experience sequencing samples on the Illumina platform that have been whole-genome amplified? Has anyone explicitly tested whether it introduces allelic/amplification bias that could be a problem for SNP calling?

    We've heard rumors and impressions that say it's a problem and should be avoided. But we have valuable samples that we would like to sequence and WGA seems to be the only way.

    Thanks in advance.
  • henry.wood
    Member
    • Apr 2010
    • 63

    #2
    How tiny are your samples. If you are gentle you can make a perfectly good library with a lot less than 1ug.

    Comment

    • Crawford505
      Junior Member
      • Jun 2009
      • 4

      #3
      Many of them are in the 400-500ng region. Have you had success with samples in this range. Do you modify the standard protocol to accommodate the lower grammage?
      Thanks.

      Comment

      • henry.wood
        Member
        • Apr 2010
        • 63

        #4
        We don't bother tweaking anything until we get to less than 100ng, so hopefully 400-500ng should be fine using normal protocols. If you are feeling twitchy you could practice using 400ng of a non-precious sample. You can tell if a library has worked by looking at the agilent/picogreen results without having to sequence it.
        Once we get to low concentration we start to fiddle around with adaptor concentrations and PCR cycles.

        Comment

        • Crawford505
          Junior Member
          • Jun 2009
          • 4

          #5
          Thanks very much that is helpful.

          However, we are considering outsourcing our sequencing (the price is right) and the facility that would do it for us is requiring what i'm assuming they are considering 'safe' amounts of starting material.

          So whole-genome amplification is still something we may want to look into. Anyone have any experience with this?

          Comment

          • jasonwood
            Member
            • May 2010
            • 10

            #6
            I used to use Sigma WGA to amplify ChIP samples for microarrays. Haven't needed to do this since I routinely make illumina libraries from 20ng or so starting material, and have been getting good sequencing results. One thing to keep in mind is that many WGA systems are PCR based, so if you then make libraries from them, you'll just end up sequencing the WGA adapter oligo instead of the sample. Maybe another WGA technique like MDA might work, but I don't have any experience with that.
            But since the illumina library prep already has an amplification built in, i don't see any need to pre-amplify. you can use very small starting amounts. Find a place that will sequence pre-made libraries if necessary, and make them yourself.

            Comment

            • csquared
              Member
              • May 2008
              • 67

              #7
              We haven't had any trouble making libraries down to 10ng input in our service lab and lower in the R&D lab. You need to appropriately alter the ratios in the library prep but it's nothing difficult.
              HudsonAlpha Institute for Biotechnology
              http://www.hudsonalpha.org/gsl

              Comment

              • krobison
                Senior Member
                • Nov 2007
                • 734

                #8
                A number of vendors are introducing kits for smaller amounts of DNA; my notes say that NuGEN's requests 200ng and I know somewhere another company claims to have kits with a 100ng requirement (but those notes aren't making themselves found).

                I also have data from WGA & at a crude level it looks fine. One thing to be careful of is any procedure using PCR -- you do risk having the PCR primer(s) occupying the ends of many of your reads.

                Comment

                • Crawford505
                  Junior Member
                  • Jun 2009
                  • 4

                  #9
                  Just so we're clear, we are interested in library preps of genomic DNA for DNA-seq runs. Are the numbers referred to here applicable to gDNA preps?

                  Comment

                  • jasonwood
                    Member
                    • May 2010
                    • 10

                    #10
                    Yes, you should be fine with just a few ng of fragmented gDNA. Most of my samples are chip-seq (which is also fragmented genomic), and i start with 20ng if available, but routinely go down below 10ng without many problems if the IP is low efficiency. Certainly anything in the 20-100 range is quite robust (with diluted adapter primers).

                    Comment

                    • kainsteven
                      Member
                      • Jul 2009
                      • 16

                      #11
                      The Ovation WGA FFPE System from NuGEN can be used for DNA from FFPE or Fresh Frozen (FF) material with input amounts as low as 100 ng.

                      Steve

                      Comment

                      • rthornton4
                        Junior Member
                        • Sep 2010
                        • 6

                        #12
                        Hi,
                        We just attempted to sequence a WGA ChIP-seq library on our GA-II and it failed miserably. The adapter that Sigma uses in their WGA kit confused the RTA software and what the sequencer did call was junk. From talking with our FAS, the problem is that the new software needs to identify all four bases to call any single cluster and this adapter does not accommodate that. Currently we are being told that the only work-around is to design a sequencing primer that will hybridize to the adapter sequence or run these samples on the older software.

                        Comment

                        • frozenlyse
                          Senior Member
                          • Sep 2008
                          • 135

                          #13
                          Alternatively you could concatamerise your WGA'd material and refragment, that way the WGA primer sequences will not be on the end of your fragments - of course your downstream analysis will be made that much harder... You are probably better off optimising Illumina library prep for small amounts of DNA (which many labs have done)

                          Comment

                          • rthornton4
                            Junior Member
                            • Sep 2010
                            • 6

                            #14
                            Thanks frozenlyse for the advice! Unfortunately I didn't prepare the WGA sample. It was prepped by an outside user and I made the library. I do agree with optimizing library preps with smaller amounts of input but I don't have all of the information on why the user WGA'd their material. Another user had done this to compare WGA material vs. non-WGA material to see what, if any, the differences were in the sequencing. They had found the WGA material was highly comparable to the non-WGA material. I don't know if that was this users intent. All I do know is that they are not happy with Illumina currently.

                            Comment

                            • elaney_k
                              Member
                              • Mar 2008
                              • 55

                              #15
                              Originally posted by rthornton4 View Post
                              Thanks frozenlyse for the advice! Unfortunately I didn't prepare the WGA sample. It was prepped by an outside user and I made the library. I do agree with optimizing library preps with smaller amounts of input but I don't have all of the information on why the user WGA'd their material. Another user had done this to compare WGA material vs. non-WGA material to see what, if any, the differences were in the sequencing. They had found the WGA material was highly comparable to the non-WGA material. I don't know if that was this users intent. All I do know is that they are not happy with Illumina currently.
                              Hi,

                              I've managed to sequence samples for a collaborator on the Illumina GA that were target-enriched using the nimblegen/454 prep. The input DNA had the 454 sequencing adapters on either end of the DNA (22 bases long on the 5' end). I thought about concatamerising the DNA before doing an Illumina prep but figured the downstream bioinformatics might be a bit of a nightmare so prepped them without it. I made sure not to overload the flowcell and recalled all the data from the saved images. I got great usable data from it. I have to say in your case Illumina can't really be blamed, if you understand how the cluster calling algorithm/chemistry works then it's much easier to predict these kinds of problems.

                              Elaine

                              Comment

                              Latest Articles

                              Collapse

                              • SEQadmin2
                                Proteomic Platforms: How to Choose the Right Analytical Strategy to Improve Detection and Clinical Applications
                                by SEQadmin2


                                Proteomics platforms are evolving rapidly, with advances in mass spectrometry and affinity-based approaches expanding what researchers can detect and at what scale. As the field moves toward deeper proteome coverage and clinical applications, scientists face an increasingly complex landscape of tools. This article will explore how researchers are navigating these choices to find the right platform for their work.

                                The systematic characterization of the human proteome has
                                ...
                                Yesterday, 11:48 AM
                              • SEQadmin2
                                Advanced Sequencing Platforms Tackle Neuroscience’s Toughest Genomics Problems
                                by SEQadmin2



                                Genomics studies in neuroscience face a special challenge due to the brain’s complexity and scarcity of samples. Mapping changes in cell type and state using conventional next-generation sequencing methods remains challenging. Advances in technologies like single-cell sequencing, spatial transcriptomics, and long-read sequencing have opened the door to deeper studies of the brain and diseases like Alzheimer’s, amyotrophic lateral sclerosis (ALS), and schizophrenia.
                                ...
                                07-09-2026, 11:10 AM
                              • SEQadmin2
                                Cancer Drug Resistance: The Lingering Barrier to Rising Survival
                                by SEQadmin2



                                Cancer survival rates have significantly increased in the last few decades in the United States, reaching a combined 70% 5-year survival rate by 2021. Behind this number, there are years of research to find new therapies, drug targets, and early detection methods. But there is one core challenge that keeps slowing down these advances, and it’s about drug resistance.

                                There is no single reason why many patients don’t respond to treatment as expected. Cancer is...
                                07-08-2026, 05:17 AM

                              ad_right_rmr

                              Collapse

                              News

                              Collapse

                              Topics Statistics Last Post
                              Started by SEQadmin2, Yesterday, 11:10 AM
                              0 responses
                              9 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 07-13-2026, 10:26 AM
                              0 responses
                              30 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 07-09-2026, 10:04 AM
                              0 responses
                              40 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 07-08-2026, 10:08 AM
                              0 responses
                              25 views
                              0 reactions
                              Last Post SEQadmin2  
                              Working...