    Poor Ligation Efficiencies on illumina Libraries

    I was just wondering what are the ligation efficiencies on illumina libraries for people working on it.I get poor efficiencies on QPCR, ranging from 40-5%, and even worst for no PCR libraries.I calculate my efficiency as mentioned below.
    QPCR/Bioanalyzer Conc.*100
    I did compare on Taqman as wel as sybrgreen kapa reagents, but din't really had a significant difference.
    Please let me know if anybody is having similar problems
  • gaster
    Member
    • May 2009
    • 17

    #2
    me too

    I have had the same problem. I am using a no-amp approach and with the two libraries I have tried the ligation efficiency (including recovery efficiency) is less than 1%. I have tried both NEB ligase and Enzymatics Ultra-pure ligase.

    Do you see an size shift an obvious size shift in your library post ligation indicating that adaptors have been ligated to the insert?

    Comment

    • niceday
      Member
      • Apr 2010
      • 68

      #3
      I would be more interested in the ligation efficiency before PCR.
      After PCR the amount of ligated material is much higher because PCR only amplifies up properly ligated library.

      qPCR results can be quite a bit less than bioanalyser or qubit results as they show all the DNA and qPCR shows the good library.

      Depending on your ligation efficiency and how many cycles of PCR you have done will depend on what % of your library is good.

      Does anyone else out there have a feel for their ligation efficiency? We have decided to start logging it as a QC step and I would be interested in what sorts of figures other sites are getting.

      Comment

      • Liting
        Member
        • Jul 2011
        • 12

        #4
        We have had the same problem.We can not get enough libraries after PCR .We have confirmed that there is noting wrong with the ligase by ligating marker DNA.We think the problem is because of adaptors.We have tried Illumina and NEB kit.The problem is still there.Does anyone tell us how to solve the problem? Thank you very much.

        Comment

        • Heisman
          Senior Member
          • Dec 2010
          • 534

          #5
          Originally posted by Liting View Post
          We have had the same problem.We can not get enough libraries after PCR .We have confirmed that there is noting wrong with the ligase by ligating marker DNA.We think the problem is because of adaptors.We have tried Illumina and NEB kit.The problem is still there.Does anyone tell us how to solve the problem? Thank you very much.
          Do you make sure the adapters are hybridized together prior to using them?

          Comment

          • NextGenSeq
            Senior Member
            • Apr 2009
            • 482

            #6
            Also make sure your adapters are synthesized with a phosphotioate bond between the last two 3' bases.

            Comment

            • Liting
              Member
              • Jul 2011
              • 12

              #7
              Library preparation

              Originally posted by NextGenSeq View Post
              Also make sure your adapters are synthesized with a phosphotioate bond between the last two 3' bases.
              Thank you very much.Our adapters work well and don't have any problems.We have got enough libraries by PCR.But the size of DNA were smaller.It's abnormal.I can't find how it happens.We already carried on llibraries preparation for two months.It is unsuccessful.

              Comment

              • epistatic
                Senior Member
                • Mar 2009
                • 129

                #8
                We go immediately from dA-tailing to ligation to get the best yields as the 3'A is unstable. I also limit the ligation to 15 minutes and stop with EDTA before going to bead purification to avoid ligating too long.

                Comment

                • pmiguel
                  Senior Member
                  • Aug 2008
                  • 2328

                  #9
                  Originally posted by epistatic View Post
                  We go immediately from dA-tailing to ligation to get the best yields as the 3'A is unstable. I also limit the ligation to 15 minutes and stop with EDTA before going to bead purification to avoid ligating too long.
                  Do you have a reference to support that the 3' A is "unstable". I can't think of any reason it would be any more unstable than any other 3' overhang.

                  --
                  Phillip

                  Comment

                  • pmiguel
                    Senior Member
                    • Aug 2008
                    • 2328

                    #10
                    Originally posted by Liting View Post
                    Thank you very much.Our adapters work well and don't have any problems.We have got enough libraries by PCR.But the size of DNA were smaller.It's abnormal.I can't find how it happens.We already carried on llibraries preparation for two months.It is unsuccessful.
                    How much smaller? By what assay did you determine size?

                    Single-stranded DNA migrates more slowly on several types of Bioanalyzer chips than double-stranded. So single stranded regions (eg, Y-adapters), could cause amplicons to "look" longer than they really are on a Bioanalyzer chip. After "enrichment" PCR the Y-adapters would no longer be single-stranded.

                    --
                    Phillip

                    Comment

                    • Shifty
                      Junior Member
                      • Feb 2011
                      • 1

                      #11
                      I have similar problems with the Truseq adapters on no-amp materiaal.
                      2,5 uq of DNA sheared with covaris (max peak 400 bp)

                      end-repair
                      add 'a
                      ligation (all in NEB kit) according to protocol and AMPURE xp washes between.
                      size selection with Agilent XT 400 to 500 bp

                      On bioanalyzer I get a nice peak (2 nM concentration)

                      But poor cluster results.

                      I thought that I could see good results from the test PCR to check ligation. But cluster density shows otherwise. Next week qPCR test.

                      Never had this problem before. Happend now twice.

                      Comment

                      • pmiguel
                        Senior Member
                        • Aug 2008
                        • 2328

                        #12
                        If you are not doing amplification you might be able to get an approximate assay on % of fragments with adapters on the agilent chip. Just save an aliquot of the fragmented, end polished, A-tailed DNA. Use the rest for ligation as normal. Then run pre- and post- ligation on the Agilent high sensitivity DNA chip.

                        You may be able to see a shift in the size of the fragments. The adapters comprise a 63 and a 58 bp oligo. So, you would expect at least a 120 bp shift for fragments that have been ligated at both ends to adapters. (The shift might be larger due to ssDNA drag.)

                        --
                        Phillip

                        Comment

                        • sehrrot
                          Member
                          • Jul 2010
                          • 58

                          #13
                          Dear all,

                          I hope this would be helpful
                          To increase ligation efficiency,
                          1. Purify your samples after covaris shearing. I've noticed that some samples had been kept in TE buffer and ligation wasn't worked well. So, every time after shearing, I purify samples using Illumina resuspension buffer then make dilution 20 ng/ul (for 1 ug input)
                          2. Check your concentration before PCR and then compare the concentration b/w pre-pcr and post-pcr. I've used both gel-free and gel method; PCR efficiency of gel-free method is around 10X and that of gel method is over 30.
                          3. Check the expiry date of Truseq library prep kit. This is pretty short than the previous version.

                          Comment

                          • fluo
                            Junior Member
                            • Jul 2011
                            • 2

                            #14
                            sehrrot is right.

                            check the expriy date of Truseq library prep kit. It is important to avoid expired reagent.
                            I used expired reagent which was out of expriy date about a month and got a very poor PCR efficiency(about 2X). Now, I changed new reagents and get a PCR efficiency which is over 20X.

                            Comment

                            • sehrrot
                              Member
                              • Jul 2010
                              • 58

                              #15
                              Hey fluo

                              Have you ever tried Herculase pol. instead of Phusion HF buffer? I feel it gives me more efficient amplification..

                              Comment

                              Latest Articles

                              Collapse

                              • SEQadmin2
                                Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
                                by SEQadmin2



                                Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

                                Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
                                ...
                                10-14-2026, 09:55 AM
                              • SEQadmin2
                                New Genomics Technologies Take Aim at Long-Standing Limits
                                by SEQadmin2


                                Researchers using sequencing and genomics tools often have to make trade-offs. They can choose between speed or scale, short reads or long-range information, or targeted panels or a view of the whole transcriptome. New technologies that have been released this year are built to address those tough choices.

                                We asked six companies the same four questions to learn about their latest products. The new technologies bring a lot to the table, including rethinking sequencing
                                ...
                                09-28-2026, 10:25 AM

                              ad_right_rmr

                              Collapse

                              News

                              Collapse

                              Topics Statistics Last Post
                              Started by SEQadmin2, 10-22-2026, 11:13 AM
                              0 responses
                              38 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 10-22-2026, 10:48 AM
                              0 responses
                              28 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 10-15-2026, 10:52 AM
                              0 responses
                              41 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 10-14-2026, 05:39 AM
                              0 responses
                              45 views
                              0 reactions
                              Last Post SEQadmin2  
                              Working...