    Amount of ChIP DNA required for Solexa seqeuncing

    Hi-
    Illumina recommends using between 10-50 ng of ChIP DNA for a sample prep. Does anyone have experience with using less than 10 ng? I tried using about 2 ng (it's all that I had), figuring the worst that could happen would be to get fewer reads than usual, but I ended up getting basically no reads at all. I calculated the amount of PCR-enriched DNA to load onto the flow cell based on the A280 reading. I couldn't see the PCR-enriched sample on the Bioanalyzer, but I figured that this wasn't a big deal, as another sample that gave excellent sequencing results was only barely detectable on the Bioanalyzer.
    Does anyone know if there is a general problem with using less than 10 ng starting material?

    Thanks!
  • DSB
    Junior Member
    • May 2009
    • 6

    #2
    If you are using the Nanodrop for measuring quantity of DNA, this can be out by an order of magnitude or more. I don't lnow if this generally applies to other A280 readings at low concentrations. The recommendation is to use the Picogreen assay or a RT-PCR of properly adapted sequences for all quantitative measurements .

    Comment

    • jbot
      Junior Member
      • Feb 2009
      • 3

      #3
      I agree that the Nanodrop readings can be way off. But does anyone know whether ChIP-Seq generally works ok if one uses less than 10 ng of DNA to start off with? I quantify the concentration of ChIP DNA using a PicoGreen assay and have had no trouble when using more than 10 ng to prepare the sequencing library -- but what happens if one uses less?

      Comment

      • xenia.zhang
        Member
        • Jan 2010
        • 10

        #4
        I think it's problematic if you can't detect any PCR enriched DNA on bioanalyzer,
        however, 1ng to 10ng starting material for lib. construction worked.

        Comment

        Latest Articles

        Collapse

        • SEQadmin2
          Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
          by SEQadmin2



          Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

          Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
          ...
          10-14-2026, 09:55 AM
        • SEQadmin2
          New Genomics Technologies Take Aim at Long-Standing Limits
          by SEQadmin2


          Researchers using sequencing and genomics tools often have to make trade-offs. They can choose between speed or scale, short reads or long-range information, or targeted panels or a view of the whole transcriptome. New technologies that have been released this year are built to address those tough choices.

          We asked six companies the same four questions to learn about their latest products. The new technologies bring a lot to the table, including rethinking sequencing
          ...
          09-28-2026, 10:25 AM

        ad_right_rmr

        Collapse

        News

        Collapse

        Topics Statistics Last Post
        Started by SEQadmin2, Yesterday, 11:13 AM
        0 responses
        17 views
        0 reactions
        Last Post SEQadmin2  
        Started by SEQadmin2, Yesterday, 10:48 AM
        0 responses
        18 views
        0 reactions
        Last Post SEQadmin2  
        Started by SEQadmin2, 10-15-2026, 10:52 AM
        0 responses
        36 views
        0 reactions
        Last Post SEQadmin2  
        Started by SEQadmin2, 10-14-2026, 05:39 AM
        0 responses
        42 views
        0 reactions
        Last Post SEQadmin2  
        Working...