    Ion Sphere ID summary, polyclonal %

    For our first runs we have "Filtered: Polyclonal" in the 25-35% range. Is this due to more beads put into the emulsions than necessary, non-accurate library titration or possibly >1 bead in a well? Is this percentage in the expected range?
    Last edited by The_Dark_Base; 12-20-2011, 04:40 PM.
  • ionman
    Junior Member
    • Nov 2011
    • 2

    #2
    Hi There,
    the polyclonal filter filters out beads that are estimated to have more than one template per ISP. Due to the nature of the protocol there always be some beads with more than one template on them. However, for a good quality run the "polyclonal" figure should be below 10%. Generally, high polyclonal figures isuggest ePCR titration issue.
    See also "ISP Identification Summary" on the Ion Community:

    Comment

    • IonTorrent
      Member
      • Jan 2010
      • 64

      #3
      I concur, ionman. If you've upgraded to software v1.5.1 (and you should! lots of great improvements in there), here is an updated description of the ISP summary table, including the changes made to improve estimation of polyclonality:

      Comment


      • #4
        Most of the 1.5 reports seem to have higher polyclonal numbers than previous versions. I think they have changed the way it filters a couple times, so you can't compare to older reports.

        Comment

        Latest Articles

        Collapse

        • SEQadmin2
          Nine Things a Sample Prep Scientist Thinks About Before Sequencing
          by SEQadmin2


          I’m not a sequencing expert. I’m a purification scientist who uses NGS to evaluate workflows my group develops. With this perspective, we think about the sample first and the NGS workflow second. The sequencer is an exceptionally honest reporter, but it can only report on what you give it, so whether you get clean, interpretable data from an NGS workflow is largely determined before you begin.

          Here are nine questions we think about, in roughly the order they matter, before...
          06-18-2026, 07:11 AM
        • SEQadmin2
          From Collection to Sequencing: Why Sample Preparation and Preservation Define Sequencing Data
          by SEQadmin2


          Data variability is still an issue in sequencing technologies despite the advances in reproducibility and accuracy of these platforms. But the problem does not originate in the sequencing itself, but in the previous steps, before the sample reaches the sequencer.


          The first step is collection, followed by preservation and sample preparation for analysis. Most scientists overlook those steps, but not being careful might just be skewing the experiment’s results.
          ...
          06-02-2026, 10:05 AM

        ad_right_rmr

        Collapse

        News

        Collapse

        Topics Statistics Last Post
        Started by SEQadmin2, 06-26-2026, 11:10 AM
        0 responses
        15 views
        0 reactions
        Last Post SEQadmin2  
        Started by SEQadmin2, 06-17-2026, 06:09 AM
        0 responses
        49 views
        0 reactions
        Last Post SEQadmin2  
        Started by SEQadmin2, 06-09-2026, 11:58 AM
        0 responses
        107 views
        0 reactions
        Last Post SEQadmin2  
        Started by SEQadmin2, 06-05-2026, 10:09 AM
        0 responses
        125 views
        0 reactions
        Last Post SEQadmin2  
        Working...