    Calculation of % Bead enrichment (454 FLX Titanium)

    Hi everyone!

    I am doing Amplicon Sequencing on a 454 GS FLX Titanium Platform (so I am following the emPCR Method Manual – Lib-A SV) and I have a question regarding the calculation of the % bead enrichment
    (on the protocol: 3.7 Sequencing Primer Annealing - step 6).

    For some unexplained reason, the enrichment values are very high, from 40% to even 70% (except for only 3 samples which are around 20%). Do you think that something went wrong with the emulsions? Certainly they did not seem broken.

    When I calculated the % Bead enrichment, for the total input beads I used as input value the number of recovered beads that I had previously counted from the % Bead Recovery (on the protocol: 3.6.1 Preparation for Indirect Enrichment - step 8), instead of the proposed by the protocol (4.0 × 10^6 for 1×Amp SVE) in order to be on the safe side. Do you think this is the correct way to calculate the % bead enrichment? I should mention that the % Bead Recovery rates were below 65% in most cases (which is considered typical) but there were around 62%.

    The beads are counted with a Z1 Beckman Coulter and the DNA was quantified at a QuantiFluor, with the Quant-iT PicoGreen dsDNA Assay Kit, as proposed by Roche.

    I am waiting for your suggestions!!
  • Christina85
    Junior Member
    • Aug 2011
    • 7

    #2
    I forgot to mention that I have performed the emulsion titration with 2 x 10^6 Capture Beads per emulsion tube and 4 x 10^6 Capture Beads per emulsion tube, to check which one will yield better % Bead enrichment, but there were barely any differences between the different emulsions.

    Comment

    • RCJK
      Senior Member
      • May 2009
      • 156

      #3
      Hi Christina,

      What cpb values did you use for your inputs? Have you tried repeating the reactions with lower cpb values? In my experience, I generally have to use a lower input amount for amplicon projects. I would also say to closely check for any broken or slightly broken emulsions. Those would be the two factors that would most likely result in such high enrichments. Hope this helps!

      Comment

      • Christina85
        Junior Member
        • Aug 2011
        • 7

        #4
        I used 2 cpbs and 4 cpbs. 2 cpbs seemed to have worked just fine the previous times I run FLX (for 16S amplicons).
        This time, I was not certain for the optimal cpbs because I am sequencing a different gene, so I decided to do titration for 2 and 4 cpbs.
        I think I might try lowering the cpbs, even though 2 is the lowest Roche proposes.
        The emulsion didn't seem broken, but based on the posts I 've checked so far they might have just as well been broken without any obvious change.

        Comment

        • ajthomas
          Senior Member
          • Mar 2010
          • 167

          #5
          You might also have a contamination problem. If any of the components of the emPCR or the workspace where you set it up have been contaminated with the waste from a previous emPCR (especially from the emulsion breaking or first melt), you will get high enrichment values that don't seem to change much with different cpb ratios.

          Comment

          • pmiguel
            Senior Member
            • Aug 2008
            • 2328

            #6
            Originally posted by Christina85 View Post
            I used 2 cpbs and 4 cpbs. 2 cpbs seemed to have worked just fine the previous times I run FLX (for 16S amplicons).
            This time, I was not certain for the optimal cpbs because I am sequencing a different gene, so I decided to do titration for 2 and 4 cpbs.
            What is the typical ratio of beads to micro reactors? Anyone? That is, you add your template to the aqueous phase of a cocktail that becomes all the non-mock micro reactors. What percentage of those actually have a bead in them?

            If every non-mock reactor had a bead in it, then you would want 0.1 to 0.2 cpb -- to get that 10 -20% enrichment value. But if most of the non-mock reactors are empty, that would explain why a higher ratio of template molecules to beads is needed.

            --
            Phillip

            Comment

            Latest Articles

            Collapse

            • SEQadmin2
              Beyond CRISPR/Cas9: Understand, Choose, and Use the Right Genome Editing Tool
              by SEQadmin2



              CRISPR/Cas9 sparked the gene editing revolution for both research and therapeutics.1 But this system still showed severe issues that limited its applications. The most prominent were the heavy reliance on PAM sequences, delivery limitations, double-stranded breaks that prompt unintended edits and cell death, and editing inefficiency (both in targeting and in knock-in reliability).

              Despite this, “CRISPR helped turn genome editing from a specialized technique into
              ...
              07-31-2026, 11:01 AM
            • SEQadmin2
              Proteomic Platforms: How to Choose the Right Analytical Strategy to Improve Detection and Clinical Applications
              by SEQadmin2


              Proteomics platforms are evolving rapidly, with advances in mass spectrometry and affinity-based approaches expanding what researchers can detect and at what scale. As the field moves toward deeper proteome coverage and clinical applications, scientists face an increasingly complex landscape of tools. This article will explore how researchers are navigating these choices to find the right platform for their work.

              The systematic characterization of the human proteome has
              ...
              07-20-2026, 11:48 AM

            ad_right_rmr

            Collapse

            News

            Collapse

            Topics Statistics Last Post
            Started by SEQadmin2, 08-06-2026, 07:41 AM
            0 responses
            15 views
            0 reactions
            Last Post SEQadmin2  
            Started by SEQadmin2, 08-03-2026, 10:13 AM
            0 responses
            31 views
            0 reactions
            Last Post SEQadmin2  
            Started by SEQadmin2, 07-31-2026, 02:55 AM
            0 responses
            41 views
            0 reactions
            Last Post SEQadmin2  
            Started by SEQadmin2, 07-24-2026, 12:17 PM
            0 responses
            26 views
            0 reactions
            Last Post SEQadmin2  
            Working...