  • pauji
    Junior Member
    • Jun 2011
    • 3

    Library on gel but not on Bioanalyzer

    I'm having a strange problem with my library prep for Illumina. Unfortunately I have no been able to find any information on this, not here nor from Agilent, so i have to assume that i'm the only one with this problem...
    maybe i can get some feedback from you guys

    My post-PCR RAD libraries (i.e. dsDNA 300-500bp) are clearly visible on my agarose gels
    but hardly or mostly not at all on the DNA1000 Bioanalyzer chips

    The Bioanalyzer runs look fine, i.e. the ladder is excellent and I cannot detect any problems with the lower and higher marker peaks, i.e. the matrix and peak detection seems to work fine. Between those peaks though the electrogram is completely flat, no DNA at all by now I have at least 20 different PCR products that I can't visualize on the BioAnalyzer. So my PCR apparently works but without the Bioanalyzer I can't sequence

    Has anybody seen/heard of this before? any idea what I can do?? I guess the problem is most likely with the Bioanalyzer but...
    any comments appreciated!!
    cheers
    P
  • vkkolla
    Junior Member
    • May 2012
    • 1

    #2
    Hi P, try it on DNA High sensitivity chip.
    Cheerz..

    Comment

    • pmiguel
      Senior Member
      • Aug 2008
      • 2328

      #3
      There are lots of possibilities. Here is a list in order of what seems most likely to me:
      1. You are diluting your DNA before loading it on the BioAnalyzer DNA chip. If so, reduce the dilution.
      2. Your agarose gel is more sensitive than the BioAnalyzer chip. This could happen if your BA reagents are expired, or near their expiration date, or if you gel electrophoresis parameters are tuned well and you have a good image documentation set up. Hard to say though because you gave absolutely none of your gel electrophoresis parameters. If you know the mass of your MW ladder for the gel, you could estimate the concentration of your PCR products. Then see if what you loaded is within the detection limits of a DNA chip.
      3. Your agarose gel loading dye or PCR mix is contaminated with something you are interpreting as your PCR fragment. Try running a "mock" lane, where you set up a PCR reaction, with no template DNA. Then, with no thermal cycling, prep that the same way you do real samples and load it on the gel. Use the same tips and pippeters! (They may be the source of this contaminating band.)
      4. Your PCR products are single-stranded. This could happen if one of your primers ran out before the other. All subsequent amplification cycles would linearly increase the concentration of only one of the strands. Single-stranded molecules migrate more slowly on BioAnalyzer chips than with gel electrophoresis. You could test this by running some of your samples on an RNA chip.
      5. Okay, this is the most crazy one, and almost certainly not what you are seeing. Years ago I loaded some uncut plasmid preps on a DNA chip. I could see these clearly on an agarose gel. But saw nothing on the DNA chip. Maybe the supercoiling caused them not to bind any of the dye? Anyway, if your bands are actually somehow became circular and supercoiled... This is testable, but so far-fetched I am not going to comment further.

      Comment

      • pauji
        Junior Member
        • Jun 2011
        • 3

        #4
        Thanks for your input guys, i really appreciate!

        as of your comments:
        a) will do! couldn't do it so far because we don't have them in my lab and agilent was unwilling to give me a sample...


        b) lots of good ideas, thanks!
        1. no dilution.

        2. yep, my agarose gel is more sensitive than the BA chip! i just don't get why
        the DNA1000 is supposed to detect 0.5ng/ul, Etbr-agarose gels usually require ~10ug... according to my guesstimates based on ladder intensity i have plenty of DNA to detect with the BA but i don't! i considered old reagents but the beautiful marker peaks and BA ladder seemed to indicate that everything is working fine... still this is the most parsimonious explanation for me.
        positive gels were obtained with very different gel parameters, i.e. 1-2%, 100-200V, 30-60min, different loading dyes...

        3. yep, i considered that too but that'd mean tons of contamination at just the right size + my negative controls are fine (i.e. empty) + i should detect the contamination on the BA as well + no clear contamination pattern etc. etc.

        4. i read about ssDNA BA problems but i though PCR products are by definition dsDNA... single-stranded PCR products seemed like an oxymoron too me! definitely a cool idea, thanks! i obviously can't exclude that but it seems unlikely given that a) the band is just the right size on the gel (i.e. not faster) and i have nothing on the BA (i.e. no big unexplained band)

        5. also a very cool idea that i hadn't considered!

        as of now, i looks like i'm having a fairly rare problem that might cannot be easily explained.
        i think i'll ask a friend in another lab to run a chip for me on another BA with other reagents... i.e. changing everything and hope that everything changes :-) i'll report back!

        cheers!
        P

        Comment

        • Susanne
          Member
          • Aug 2009
          • 33

          #5
          In case you are dealing with a smear, your sensitivity of 0.5 ng/µl for the DNA1000 doesn't hold true, as these are defined to be in a (sharp) peak.
          For a smear, the sensitivity will be lower (maybe ~10fold).

          Comment

          • miss_shell
            Junior Member
            • Mar 2012
            • 4

            #6
            I agree with Vkkolla. We run all of our post-PCR samples on a high sensitivity chip. Should solve your problem.

            Comment

            • pauji
              Junior Member
              • Jun 2011
              • 3

              #7
              hey guys

              looks like the price goes to... vkkolla & miss_shell for suggesting the high sensitivity chip... it showed beautiful libraries, which i accepted instantly

              after reading up a bit more here, i found out that most people do qc with BA chip but don't trust its quantification at all. subsequent quantifications of my libraries with the qubit system were 2-10x higher than BA quantifications and coincided much better with gel observations. I thus went with that and am now waiting for the results from sequencing.

              it doesn't all make sense to me but it doesn't have to as long as i works ;-) otherwise, i'll be back soon...

              wish me good luck :-)
              cheers
              P

              Comment

              Latest Articles

              Collapse

              • SEQadmin2
                From Collection to Sequencing: Why Sample Preparation and Preservation Define Sequencing Data
                by SEQadmin2


                Data variability is still an issue in sequencing technologies despite the advances in reproducibility and accuracy of these platforms. But the problem does not originate in the sequencing itself, but in the previous steps, before the sample reaches the sequencer.


                The first step is collection, followed by preservation and sample preparation for analysis. Most scientists overlook those steps, but not being careful might just be skewing the experiment’s results.
                ...
                Yesterday, 10:05 AM
              • SEQadmin2
                Single-Cell Sequencing at an Inflection Point: Early Impacts of New Platforms and Emerging Trends
                by SEQadmin2


                With the launch of new single-cell sequencing platforms in 2026, the field stands at an exciting inflection point. This article surveys the most impactful advances in the field and discusses how they’re reshaping research in cancer, immunology, and beyond.


                Introduction

                Single-cell sequencing technologies have undergone remarkable advances over the past decade, transitioning from low-throughput experimental approaches to highly scalable platforms capable of...
                05-22-2026, 06:42 AM
              • SEQadmin2
                Environmental Genomics in the Age of NGS: From Microbes to Conservation Strategies
                by SEQadmin2

                Studying ecosystems means dealing with complex, multi-species communities that are hard to observe at scale. This complexity, however, hides many important questions to be answered, from how biogeochemical cycles work and how climate change can affect species distribution to how conservation strategies can work best.


                Genomics, particularly since the expansion of NGS, has transformed ecosystem ecology. By sequencing environmental DNA, we can now assess biodiversity without direct...
                05-06-2026, 09:04 AM

              ad_right_rmr

              Collapse

              News

              Collapse

              Topics Statistics Last Post
              Started by SEQadmin2, Yesterday, 12:03 PM
              0 responses
              17 views
              0 reactions
              Last Post SEQadmin2  
              Started by SEQadmin2, Yesterday, 11:40 AM
              0 responses
              13 views
              0 reactions
              Last Post SEQadmin2  
              Started by SEQadmin2, 05-28-2026, 11:40 AM
              0 responses
              29 views
              0 reactions
              Last Post SEQadmin2  
              Started by SEQadmin2, 05-26-2026, 10:12 AM
              0 responses
              31 views
              0 reactions
              Last Post SEQadmin2  
              Working...