    TruSeq / SPRIworks?

    Is anyone else using the SPRIworks system with Illumina's TruSeq kits? We are noticing some unfortunate changes. We started using SPRIworks with the old adapters and didn't have much trouble. Now that we are using TruSeq, suddenly the adapter dimer situation is much worse. Also, some libraries are running higher in size than they should be after PCR - for instance, a 200-400 bp size selection is resulting in some libraries with an avg size of 500 bp.

    It seems like the size selection is inconsistent.

    I'd love to hear any feedback or experiences from other groups.
  • ashchin
    Member
    • Dec 2010
    • 19

    #2
    Hi ,
    If you overload the pcr reaction you get fragments greater than 600bp along with 200 to 400 bp.

    Comment

    • HESmith
      Senior Member
      • Oct 2009
      • 512

      #3
      The TruSeq adapters are substantially different from the earlier ones, with a much larger region of unpaired sequence, that causes the libraries to migrate anomalously during gel electrophoresis. Illumina recommends using Sybr Gold staining to mitigate the problem.

      Comment

      • jlove
        Member
        • Apr 2011
        • 49

        #4
        Originally posted by ashchin View Post
        Hi ,
        If you overload the pcr reaction you get fragments greater than 600bp along with 200 to 400 bp.
        Ashchin - do you mean overload the PCR with template? I don't *think* that's happening -this is ChIP-Seq and thus I am using very little input. It actually doesn't seem to happen as much with greater input, such as with genomic DNA sequencing.

        Thanks for the suggestion, I'd like to hear more!

        Comment

        • jlove
          Member
          • Apr 2011
          • 49

          #5
          Originally posted by HESmith View Post
          The TruSeq adapters are substantially different from the earlier ones, with a much larger region of unpaired sequence, that causes the libraries to migrate anomalously during gel electrophoresis. Illumina recommends using Sybr Gold staining to mitigate the problem.
          Thanks HESmith - I am actually using SPRIworks, which does the size selection using beads, so SYBR staining isn't an option here... but the question is, why are they migrating anomalously? I am under the impression that they are slipping through the size selection because they are just under the cutoff size of 200-400 bp. Do you think they are also clumping together and running larger? This might make sense, if so, since they are showing up in the clusters in larger numbers than the bioA leads me to believe when I look at the 125 bp peak.

          Comment

          • vtosha
            Member
            • May 2010
            • 36

            #6
            Do you run size selection after adapter ligation? As I can suppose some fragments could ligate to each other in spend of adapter ligation, and so we get chimera libraries with length approximately twice than we wait. Try to take more adapters.

            Comment

            • jlove
              Member
              • Apr 2011
              • 49

              #7
              Originally posted by vtosha View Post
              Do you run size selection after adapter ligation? As I can suppose some fragments could ligate to each other in spend of adapter ligation, and so we get chimera libraries with length approximately twice than we wait. Try to take more adapters.
              vtosha- Thank you for your reply. We do run the size selection after ligation, so I don't know if this is happening or not. But it's worth looking into, thank you.

              Comment

              • genbio64
                Member
                • Dec 2009
                • 42

                #8
                @jlove, did you figure out the SPRI works problem?

                Comment

                • jlove
                  Member
                  • Apr 2011
                  • 49

                  #9
                  Hi GenBio64- We stopped using SPRIworks altogether, and I think the reason for our troubles had to do with the fact that we were doing ChIP-Seq (with tiny amounts of input), which does not size select very well on beads. It works fine with larger inputs, generally, but I don't have any experience with it using TruSeq, since it's been so long since I tried using it. I just do them manually.

                  Comment

                  • pmiguel
                    Senior Member
                    • Aug 2008
                    • 2328

                    #10
                    Ampure beads work well with ng levels of DNA. However the rest of the SPRIworks reagents are likely calibrated for much higher levels of inserts. Might have worked to dilute the adapters 10x to maintain a reasonable molar ratio of insert:adapters.

                    --
                    Phillip

                    Comment

                    • jlove
                      Member
                      • Apr 2011
                      • 49

                      #11
                      Phillip- you're right, I do ChIP manually with bead purifications and it works very well. The reaction volumes (and bead amounts) were not compatible with the small inputs in the SPRI cartridges.

                      Comment

                      Latest Articles

                      Collapse

                      • SEQadmin2
                        Proteomic Platforms: How to Choose the Right Analytical Strategy to Improve Detection and Clinical Applications
                        by SEQadmin2


                        Proteomics platforms are evolving rapidly, with advances in mass spectrometry and affinity-based approaches expanding what researchers can detect and at what scale. As the field moves toward deeper proteome coverage and clinical applications, scientists face an increasingly complex landscape of tools. This article will explore how researchers are navigating these choices to find the right platform for their work.

                        The systematic characterization of the human proteome has
                        ...
                        Today, 11:48 AM
                      • SEQadmin2
                        Advanced Sequencing Platforms Tackle Neuroscience’s Toughest Genomics Problems
                        by SEQadmin2



                        Genomics studies in neuroscience face a special challenge due to the brain’s complexity and scarcity of samples. Mapping changes in cell type and state using conventional next-generation sequencing methods remains challenging. Advances in technologies like single-cell sequencing, spatial transcriptomics, and long-read sequencing have opened the door to deeper studies of the brain and diseases like Alzheimer’s, amyotrophic lateral sclerosis (ALS), and schizophrenia.
                        ...
                        07-09-2026, 11:10 AM
                      • SEQadmin2
                        Cancer Drug Resistance: The Lingering Barrier to Rising Survival
                        by SEQadmin2



                        Cancer survival rates have significantly increased in the last few decades in the United States, reaching a combined 70% 5-year survival rate by 2021. Behind this number, there are years of research to find new therapies, drug targets, and early detection methods. But there is one core challenge that keeps slowing down these advances, and it’s about drug resistance.

                        There is no single reason why many patients don’t respond to treatment as expected. Cancer is...
                        07-08-2026, 05:17 AM

                      ad_right_rmr

                      Collapse

                      News

                      Collapse

                      Topics Statistics Last Post
                      Started by SEQadmin2, Today, 11:10 AM
                      0 responses
                      7 views
                      0 reactions
                      Last Post SEQadmin2  
                      Started by SEQadmin2, 07-13-2026, 10:26 AM
                      0 responses
                      29 views
                      0 reactions
                      Last Post SEQadmin2  
                      Started by SEQadmin2, 07-09-2026, 10:04 AM
                      0 responses
                      38 views
                      0 reactions
                      Last Post SEQadmin2  
                      Started by SEQadmin2, 07-08-2026, 10:08 AM
                      0 responses
                      25 views
                      0 reactions
                      Last Post SEQadmin2  
                      Working...