    Library invisible on gel, despite Nanodrop results. Gel extraction problem?

    I'm having some difficulty with my very first Illumina library prep, and was hoping someone might point out any obvious errors I'm committing in my ignorance. I'm using a protocol that involves size selection via agarose gel, then adapter ligation, then PCR amplification. Everything look a-OK when things are run out on the gel, stained with GelRed. There's a nice, bright smear from which I cut a band, and purify it using the Qiagen MinElute gel extraction kit. I ligate on the adapter and do the PCR. I run it out on a gel to check if it worked and... nada. The ladder is crisp and clear but the PCR product and template are invisible. Where did the DNA go?

    OK, so I'm messing something up, probably with the gel extraction if I had to guess. But weirdly, when I analyze the template and PCR product with a Nanodrop, I get the following:

    Template:
    Nucl. Acid Concentration: 11.7 ng/ul
    A260: 0.235
    A280: 0.075
    260/280: 3.15
    260/230: 3.09

    PCR:
    Nucl. Acid Concentration: 720.2 ng/ul
    A260: 14.405
    A280: 9.906
    260/280: 1.45
    260/230: 0.61

    There seems to be a decent amount of DNA in both, and it's much greater in the PCR. Does that mean the PCR is working, despite my not being able to see it on the gel? Is some contaminant from the gel purification to blame for these seemingly high DNA concentrations?

    I definitely messed a few things up, such as using anhydrous ethanol rather than molecular grade, and not letting the MinElute columns equilibriate to room temperature, and not doing enough washes during the gel extraction. But what do you make of the Nanodrop results?
  • nitut
    Junior Member
    • Sep 2012
    • 1

    #2
    Hello,
    Can you please check if there is an gel extraction to be carried out before Adapter ligation?
    If I am not wrong after the Adapter ligation, the protocols says to do AMPure XP Bead purification and then gel extraction.

    Comment

    • ulz_peter
      Senior Member
      • Feb 2010
      • 219

      #3
      The high DNA concentration after PCR might also be caused by dNTPs only, since Nanodrop can't distinguish between unbound nucleotides and a PCR product.

      Comment

      • ingrid@illumina
        Junior Member
        • Sep 2012
        • 1

        #4
        Your protocol should look like this:
        gDNA -> fragment -> repair ends -> adenylate 3' ends -> ligate adapters -> clean-up using AMPure XP beads -> size selection (gel extraction) -> PCR amplification -> validation -> normalization & pooling

        Try using Picogreen or Qubit to validate your starting material and a BioAnalyzer for the best validation of your final libraries.

        Comment

        • luc
          Senior Member
          • Dec 2010
          • 469

          #5
          thdybwf,

          your PCR Nanodrop data look fishy (for example the 260/280 ratio). Likely, your PCR did not work (because the ligation did not work, because ... ?).
          Please post the detailed protocol that you were following. How much DNA did you start out with?

          Comment

          • pmiguel
            Senior Member
            • Aug 2008
            • 2328

            #6
            Originally posted by thdybwf View Post
            But weirdly, when I analyze the template and PCR product with a Nanodrop, I get the following:

            Template:
            Nucl. Acid Concentration: 11.7 ng/ul
            A260: 0.235
            A280: 0.075
            260/280: 3.15
            260/230: 3.09

            PCR:
            Nucl. Acid Concentration: 720.2 ng/ul
            A260: 14.405
            A280: 9.906
            260/280: 1.45
            260/230: 0.61

            There seems to be a decent amount of DNA in both, and it's much greater in the PCR.
            UV spectrophotometry is often misleading. For the reasons mentioned by others in this thread and a bunch of other ones, including some I detail here.

            The nanodrop is easy to use, and the UV spectrum it plots can give you useful information -- if you bother to look at it. But whenever you perform an assay, you want to know its limits of sensitivity as well as anything that might confound it reading only (or mostly) what you are interested in (DNA concentration).

            ingrid@illumina gives good about as good advice as can be packed into a couple of sentences. However Illumina (along with many other companies) does its level best to "black box" its protocols. That is, they tend to hide salient details in a misguided attempt to prevent others from reverse-engineering their protocols. That puts them firmly in the "part of the problem" camp in my opinion.

            But, hey, I can't entirely blame Illumina because nobody seems to have taught you the basics of what a UV spectrophotometer detects. To a first approximation the answer is "lots of stuff", including DNA and RNA, but also nucleotides, acetic acid or salts thereof, salts of guanidium, phenol, DTT, etc. etc. etc. Don't feel too bad about it personally, more than 90% of bench scientists in molecular biology I meet are in the same boat. (Well, I am at a University, so these scientists are largely students...)

            Then there is the issue of sensitivity. Can you detect a 10 nM library on a nanodrop? Worthwhile to keep a couple of "rules of thumb" in mind as you navigate a protocol. Some details here. But to a first approximation:

            1 ul of a 1 nM solution contains 1 billion molecules (2nd approx.: ~60% of that.)

            1 ng of a 1 kbp fragment is 1 billion molecules (2nd approx.: ~93% of that.)


            Therefore if your library is 10 nM, that would be about 6 billion molecules/ul. If these molecules are 1 kbp, then that is about 6 ng/ul. But they are probably closer to 1/2 that length. So 3 ng/ul.

            Can you reliably detect DNA at 3 ng/ul on a nanodrop? Not on ours. Below 10 ng/ul, things get pretty sketchy. How about on an agarose gel? Too many variables to say for sure, but probably not unless your library is in a very tight size range or you have a very sensitive system set up.

            Hence the wisdom embedded in "ingrid@illuminas" advice -- go with an Agilent Bioanalyzer to QC your library. Using a high sensitivity chip you are good down into the pg range. Or fluorimetry/qPCR -- except you don't see the size of your products that way.

            --
            Phillip

            Comment

            Latest Articles

            Collapse

            • SEQadmin2
              Proteomic Platforms: How to Choose the Right Analytical Strategy to Improve Detection and Clinical Applications
              by SEQadmin2


              Proteomics platforms are evolving rapidly, with advances in mass spectrometry and affinity-based approaches expanding what researchers can detect and at what scale. As the field moves toward deeper proteome coverage and clinical applications, scientists face an increasingly complex landscape of tools. This article will explore how researchers are navigating these choices to find the right platform for their work.

              The systematic characterization of the human proteome has
              ...
              07-20-2026, 11:48 AM
            • SEQadmin2
              Advanced Sequencing Platforms Tackle Neuroscience’s Toughest Genomics Problems
              by SEQadmin2



              Genomics studies in neuroscience face a special challenge due to the brain’s complexity and scarcity of samples. Mapping changes in cell type and state using conventional next-generation sequencing methods remains challenging. Advances in technologies like single-cell sequencing, spatial transcriptomics, and long-read sequencing have opened the door to deeper studies of the brain and diseases like Alzheimer’s, amyotrophic lateral sclerosis (ALS), and schizophrenia.
              ...
              07-09-2026, 11:10 AM
            • SEQadmin2
              Cancer Drug Resistance: The Lingering Barrier to Rising Survival
              by SEQadmin2



              Cancer survival rates have significantly increased in the last few decades in the United States, reaching a combined 70% 5-year survival rate by 2021. Behind this number, there are years of research to find new therapies, drug targets, and early detection methods. But there is one core challenge that keeps slowing down these advances, and it’s about drug resistance.

              There is no single reason why many patients don’t respond to treatment as expected. Cancer is...
              07-08-2026, 05:17 AM

            ad_right_rmr

            Collapse

            News

            Collapse

            Topics Statistics Last Post
            Started by SEQadmin2, 07-20-2026, 11:10 AM
            0 responses
            14 views
            0 reactions
            Last Post SEQadmin2  
            Started by SEQadmin2, 07-13-2026, 10:26 AM
            0 responses
            32 views
            0 reactions
            Last Post SEQadmin2  
            Started by SEQadmin2, 07-09-2026, 10:04 AM
            0 responses
            43 views
            0 reactions
            Last Post SEQadmin2  
            Started by SEQadmin2, 07-08-2026, 10:08 AM
            0 responses
            29 views
            0 reactions
            Last Post SEQadmin2  
            Working...