    Originally posted by huguesparri View Post
    At the moment, 150pb read are not supported by Illumina on the Hiseq2000.
    I think it's still possible to run it on the machine but there is a strong probability that your error rate explode at the end of the read. And I don't think it will be suitable for your application.
    Unofficially, our FSE told us that a "correct" error rate for a 100nt read was below 2% (I think it was V2 kits a that time, it may have vary with V3 kits). I won't be surprise if your reach 4 to 5% error rate with a 150nt read.
    Here is a plot of the error rate (phiX) for our most recent HiSeq2000 run:



    Just for the top surface. Bubbles drive the error rate up higher for the bottom surface. Even so the error rate at 100 bases is generally below 0.5%.

    The Sequence Analysis Viewer (SAV) shows >80% of the reads with no errors (counting both surfaces). Again, this is based on aligning phiX reads spiked into 7 lanes of the flowcell to phiX by the instrument software.

    --
    Phillip

    Comment

    • huguesparri
      Member
      • May 2008
      • 97

      #17
      I've checked our last 2*100nt run. On the PhiX control lane, error rate is between 0.8 and 1%. But for the spiked PhiX in our sample, it's just above 1%.
      :'(

      Comment

      • JackieBadger
        Senior Member
        • Mar 2009
        • 385

        #18
        MiSeq is now capable of 2x250b PE

        Comment

        • NextGenSeq
          Senior Member
          • Apr 2009
          • 482

          #19
          Also the upgraded MiSeq gives 16 million reads.

          Comment

          • snetmcom
            Senior Member
            • Oct 2008
            • 159

            #20
            Originally posted by NextGenSeq View Post
            Also the upgraded MiSeq gives 16 million reads.
            16million clusters or reads?

            Comment

            • JackieBadger
              Senior Member
              • Mar 2009
              • 385

              #21
              Reads....there are many more clusters... Without the upgrade MiSeq produces up to 5 million reads.

              Comment

              • kmcarr
                Senior Member
                • May 2008
                • 1181

                #22
                Originally posted by JackieBadger View Post
                Reads....there are many more clusters... Without the upgrade MiSeq produces up to 5 million reads.
                No, with the upgrade the output is 15-17 million CLUSTERS. This makes 30-34 million PE reads.

                Comment

                • trauba
                  Junior Member
                  • May 2016
                  • 3

                  #23
                  What exactly is it that limits the read length anyways?
                  I understand imperfect cleavage of fluorescence reduces the # of extending oligos, lowering the signal + adding noise.
                  Also, eventually some oligos get "out of phase" and then muddy the signal. Is this somewhat caused by insufficient purging of reagents from previous cycles?

                  Thanks.
                  Jim

                  Comment

                  Latest Articles

                  Collapse

                  • SEQadmin2
                    Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
                    by SEQadmin2



                    Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

                    Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
                    ...
                    10-14-2026, 09:55 AM
                  • SEQadmin2
                    New Genomics Technologies Take Aim at Long-Standing Limits
                    by SEQadmin2


                    Researchers using sequencing and genomics tools often have to make trade-offs. They can choose between speed or scale, short reads or long-range information, or targeted panels or a view of the whole transcriptome. New technologies that have been released this year are built to address those tough choices.

                    We asked six companies the same four questions to learn about their latest products. The new technologies bring a lot to the table, including rethinking sequencing
                    ...
                    09-28-2026, 10:25 AM

                  ad_right_rmr

                  Collapse

                  News

                  Collapse

                  Topics Statistics Last Post
                  Started by SEQadmin2, 10-22-2026, 11:13 AM
                  0 responses
                  26 views
                  0 reactions
                  Last Post SEQadmin2  
                  Started by SEQadmin2, 10-22-2026, 10:48 AM
                  0 responses
                  20 views
                  0 reactions
                  Last Post SEQadmin2  
                  Started by SEQadmin2, 10-15-2026, 10:52 AM
                  0 responses
                  36 views
                  0 reactions
                  Last Post SEQadmin2  
                  Started by SEQadmin2, 10-14-2026, 05:39 AM
                  0 responses
                  43 views
                  0 reactions
                  Last Post SEQadmin2  
                  Working...