    MiSeq stopped run

    Hi all,

    I set up 2 x 251 paired end run over weekend, which started normally and proceeded almost to the end of the first run. but stopped Saturday morning at cycle 232 with error message "Object reference not set to an instance of an object". Obviously, tech support is closed until Monday; User Guide has no information on this error, so I am asking if anyone have seen that happened? It is upsetting as the run was going beautifully with cluster density >1300, 85% passing filter and a lot of data promised. Can the run be resumed somehow? If not, can the data from the fisrt read be be salvaged at least?
  • yaximik
    Senior Member
    • Apr 2011
    • 199

    #2
    Well, I figured out at least how to salvage data. I am posting it here if anyone runs in a similar problem. Although the run is done as nothing I can do at this point with my experience beyond proceeding with a postrun wash. But the problem was how to get usable data, which was about 5 GB out of expected 5.3 (10.6 paired). MiSeq reporter could not start as the run has not been completed, so it was happily idling in background. All I needed was to modify three files in Data/Illumina/MiSeqAnalysis/<RunNAme>/

    Adding RTAComplete.txt - I just copied one from the previous successful run and modified date to when this run has stopped

    RunInfo.xml - deleted line about read 2 and modified line about read 1 to match the number of completed cycles minus one, that is 231 (232-1)

    SampleSheet.csv - deleted line under [Reads] referring to the second set of 251 cycles, modified the line about the first set to match the actual cycles minus one (231)

    The latter two can be modified without triggering MiSeq Reporter to run, which immediately engages once the RTAComplete.txt is added. If MSR terminates with error, the job can be requeued after modifications. Voila! I got my fastq files.
    Last edited by yaximik; 10-20-2012, 06:02 PM.

    Comment

    • pmiguel
      Senior Member
      • Aug 2008
      • 2328

      #3
      Hi yaximik,
      Many thanks for your protocol. May well come in useful in the future.
      One pre-run intervention that I always try to do before starting a HiSeq run is a full console computer shutdown, start up. Or at least a reboot. I don't always do it before a MiSeq run but I think I will start doing so. Costs only a few minutes and should give you as close to a clean slate as you can get without reimaging the drive before a run.
      --
      Phillip

      Comment

      • MoritzF
        Member
        • Sep 2012
        • 10

        #4
        I encountered the same problem today...any idea how that error came about?

        Comment

        • yaximik
          Senior Member
          • Apr 2011
          • 199

          #5
          Sorry about that. Tech Support said it is likely happened because of too high cluster density (it was >1300) and intensity profile was weird going over maximum with a sharp decline before abort. I have to take that as explanation, but pmiguel doubted this is the reason ( in another thread). I hope that close to the recommended density will prevent that from happening.

          Comment

          • GenoMax
            Senior Member
            • Feb 2008
            • 7142

            #6
            It may still be possible to recover data (albeit partial depending on there the run failed) from such runs.

            You will need access to a working install of CASAVA and someone who is familiar with illumina runs since some settings files have to be edited manually.

            PS: I did not realize that Yaximilk had provided directions on how to do this on the MiSeq itself in a previous post. I will leave this in as an alternate option.
            Last edited by GenoMax; 11-01-2012, 07:33 AM.

            Comment

            • MoritzF
              Member
              • Sep 2012
              • 10

              #7
              Mine was 1257....I still don't get it though....I could understand if the results were bad when the clusters are too many, but exiting the run with a strange error message doesn't seem like a problem with cluster density to me. I contacted tech support, let's see what they say.

              Thanks fors answering anyway. I am just now trying to recoup the reads, following the protocoll you supplied. Thanks!

              Comment

              • TonyBrooks
                Senior Member
                • Jun 2009
                • 303

                #8
                Cluster density estimation will top out eventually. You may actually have >2000k/mm2, but the MiSeq will still report 1300.

                How did the Q scores look? What % of bases did you get >Q30? In my experience, this is usually a better estimate of cluster density. If Illumina are right with the over clustering, then you should have pretty terrible Q's around the time the run crashed.

                I saw something similar recently (stopped MiSeq run) where someone attempted to do a 151PE run on an insert of 60bp. R1 began fine, then deteriorated towards the end. The index read was great, but then R2 was horrible. It stopped at cycle 295.
                Cluster density was 1260, so I don't know if the problem was the over-read or the higher cluster density.
                We used yaximik's fix to ressurect the R1, although using CASAVA with the --use-bases y*,i*,n* should also work (as GenoMax stated).

                Comment

                • MoritzF
                  Member
                  • Sep 2012
                  • 10

                  #9
                  Base quality was excellent until base 80-100, at which point it dropped dramatically.

                  I also think that maybe the fragments are too small, which is very strange. We used Nextera XT library prep, and the fragments looked too large on the Agilent control chip.

                  When I checked last night, Quality score was 85% better than Q30 (at which point the machine was still before base 80) and today, before the run stopped, it was 25% better than Q30, which was when the Forward Read was complete and it began reading the reverse reads.

                  Still, I don't understand how this can cause the machine to stop completely. Bad data, bad reads, I would understand, but stopping with an incomprehensible error message wouldn't have made me think that there was a problem with the library?

                  Comment

                  • GenoMax
                    Senior Member
                    • Feb 2008
                    • 7142

                    #10
                    Originally posted by MoritzF View Post
                    Mine was 1257....I still don't get it though....I could understand if the results were bad when the clusters are too many, but exiting the run with a strange error message doesn't seem like a problem with cluster density to me. I contacted tech support, let's see what they say.
                    Is this an upgraded MiSeq or a v.1 machine? With the v.2 hardware upgrade Illumina is putting in a beefier computer (core i7 CPU with 16 GB of RAM). I do not remember if the original MiSeq had a lesser spec computer (e.g. 8 GB of RAM).

                    Could the error possibly be due to RTA (or some other software component) running out of available memory/disk space? I had noticed with a MiSeq run (where RTA did not start normally) the disk (D drive) filled up rapidly and would probably have run out of space if I had not noticed that and kicked off RTA manually.

                    Comment

                    • TonyBrooks
                      Senior Member
                      • Jun 2009
                      • 303

                      #11
                      I think our RTA ran fine. We had BCL files generated.
                      I think the problem we had is we ran out of fragments to sequence. We had probably sequenced our insert and right through the adapter. This meant we were still running cycles without being able to incorporate bases. I think this resulted in focus problems (hence terrible Q scores after about 100 bases). When it happened on R2, it go so bad that the run just aborted.

                      Our error message was as follows;

                      Best focus is too near edge of range: Autofocus best Z NaN - offset -0.0007 = NaN, which is outside soft limits -0.2950 .. -0.0050 mm: LEDSnapCommand.PrepareCaptureAndGetOffsetFpgaZPosLEDDAC (Illumina.Instrument.Bolt.Logic.SoftLimitsException)

                      Comment

                      • MoritzF
                        Member
                        • Sep 2012
                        • 10

                        #12
                        Our machines has an i3 CPU, it has not been hardware upgraded. Which is, according to the manual, not required to run 2x251 bp runs. It would be unfortunate if this was the case...

                        Anyway, the hard drives are pretty empty, so this shouldn't have caused the crash. Whether the RAM was full or not at the time the MiSeq crashed, I can't find out anymore.

                        The errors in our log file begin with
                        12-11-01 13:35:14.078 +01 000000031 INF: Y Motor: Position error before settling = -1 encoder counts (0.000 mm)
                        And then there are more of a similar kind. I hope it is not a hardware problem with a motor inside the machine.
                        All in all it seemed more like a software problem to me, though.

                        Comment

                        • mcnelson.phd
                          Senior Member
                          • Jul 2011
                          • 162

                          #13
                          Thanks for putting this out there, now I can get some data from my failed run on Saturday. My FAS was having me try to re-run RTA but that just hung at the last valid base, #502. Would have really sucked if I couldn't get any data considering the run almost completed.

                          As for the vague-ness of the error. I talked with an Illumina support person and they agreed that it's a pretty useless error message. He had me fully power down and then restart the system and said that a lot of these types of errors also come along with issues that show up when MCS is trying to re-initialize. Fortunately we didn't have any problems there but I'm still not pleased that I don't know what caused this error since this is the second time in a month that it's caused a run failure.

                          Edit: to get read 2 I had to also edit runParameters.xml and CompletedJobInfo.xml to have read 2 be only as many cycles as listed in the edited sample sheet and RunInfo.xml files.
                          Last edited by mcnelson.phd; 12-03-2012, 02:31 PM.

                          Comment

                          • yaximik
                            Senior Member
                            • Apr 2011
                            • 199

                            #14
                            That was the only purpose - if someone ever needs this. I also posted further development in a separate thread - may be someone finds this useful. I certainly did.

                            For your failed run - could you post BioAnalyzer trace of your libary, cluster density, the full intensity curve and a few tumbnail images at cycles 30, 250 and 500? I am gathering information as to what may be reasons for stopped runs.

                            Comment

                            • creeves
                              Member
                              • Jul 2012
                              • 26

                              #15
                              MiSeq stopped at start

                              I have twice had that error message you mention. For me, both times it occurred right at the beginning of the run before cluster generation had even started. When it happened the first time Illumina told me to power off and back on again, which I did, and then restart the run, which required a post-run wash first, since the MiSeq thought it had completed the previous run that never actually started. After the wash (where I switched back to the previous flow cell), the run went smoothly. I am now doing the same thing all over again on Sunday morning when there is no chance of getting help from Illumina. I think it may have something to do with space on the D: drive since I had to delete some image files before starting the first time.

                              Comment

                              Latest Articles

                              Collapse

                              • SEQadmin2
                                From Collection to Sequencing: Why Sample Preparation and Preservation Define Sequencing Data
                                by SEQadmin2


                                Data variability is still an issue in sequencing technologies despite the advances in reproducibility and accuracy of these platforms. But the problem does not originate in the sequencing itself, but in the previous steps, before the sample reaches the sequencer.


                                The first step is collection, followed by preservation and sample preparation for analysis. Most scientists overlook those steps, but not being careful might just be skewing the experiment’s results.
                                ...
                                06-02-2026, 10:05 AM
                              • SEQadmin2
                                Single-Cell Sequencing at an Inflection Point: Early Impacts of New Platforms and Emerging Trends
                                by SEQadmin2


                                With the launch of new single-cell sequencing platforms in 2026, the field stands at an exciting inflection point. This article surveys the most impactful advances in the field and discusses how they’re reshaping research in cancer, immunology, and beyond.


                                Introduction

                                Single-cell sequencing technologies have undergone remarkable advances over the past decade, transitioning from low-throughput experimental approaches to highly scalable platforms capable of...
                                05-22-2026, 06:42 AM
                              • SEQadmin2
                                Environmental Genomics in the Age of NGS: From Microbes to Conservation Strategies
                                by SEQadmin2

                                Studying ecosystems means dealing with complex, multi-species communities that are hard to observe at scale. This complexity, however, hides many important questions to be answered, from how biogeochemical cycles work and how climate change can affect species distribution to how conservation strategies can work best.


                                Genomics, particularly since the expansion of NGS, has transformed ecosystem ecology. By sequencing environmental DNA, we can now assess biodiversity without direct...
                                05-06-2026, 09:04 AM

                              ad_right_rmr

                              Collapse

                              News

                              Collapse

                              Topics Statistics Last Post
                              Started by SEQadmin2, Today, 08:59 AM
                              0 responses
                              10 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 06-02-2026, 12:03 PM
                              0 responses
                              21 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 06-02-2026, 11:40 AM
                              0 responses
                              17 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 05-28-2026, 11:40 AM
                              0 responses
                              31 views
                              0 reactions
                              Last Post SEQadmin2  
                              Working...