    Miseq index reads missing

    Hello, I was just wondering if anyone has had this problem before. I recently tried to sequence a library containing 4 barcodes on the Miseq (using custom forward, reverse, and index primers and the 2x151 kit). On my sample sheet (below), I only put 8 Ns (the length of my barcode) under "Index" since I was just going to manually demultiplex later. However, it looks like only 302 cycles completed since there are only 302 folders in the BaseCalls/L001 folder. Does this mean that the index run never completed? If not, can I still recover those reads?

    Using Illumina RTA 1.13.56

    [Header]
    Investigator Name D
    Project Name t
    Experiment Name t2
    Date 11/1/12
    Workflow Resequencing
    Assay TruSeq DNA/RNA
    Description
    Chemistry Default
    [Reads]
    151
    151
    [Settings]
    CustomRead1PrimerMix C1
    CustomIndexPrimerMix C2
    CustomRead2PrimerMix C3
    OnlyGenerateFASTQ 1
    [Data]
    Sample_ID Sample_Name Sample_Plate Sample_Well Sample_Project Index Description GenomeFolder
    t2 t2 NA NA t NNNNNNNN first_try_at_t C:\Illumina\MiSeq Reporter\Genomes\PhiX\Illumina\RTA\Sequence\Chromosomes

    I couldn't seem to find the answer to this question elsewhere, but sorry if I missed it and thank you for your help.
  • hoisinjl
    Member
    • Jun 2009
    • 23

    #2
    Can you look at the RunInfo.xml file (in the runfolder). If the software correctly parsed your samplesheet, it should read:

    <Reads>
    <Read NumCycles="151" Number="1" IsIndexedRead="N" />
    <Read NumCycles="7" Number="2" IsIndexedRead="Y" />
    <Read NumCycles="151" Number="3" IsIndexedRead="N" />

    Comment

    • Number6
      Member
      • Feb 2009
      • 19

      #3
      The MiSeq treats barcode reads differently than the HiSeq and one way to do this is to modify the MiSeq Reporter.exe.config file to include the "CreateFastqForIndexReads" setting. This is described in the MiSeq Reporter User Guide in the "MiSeq configurable settings" section.

      This will give you the barcode reads as a separate fastq file, instead of being in the sequence header, like the HiSeq.

      You can modify this file, restart the MiSeq Reporter service, and reanalyze your run with MSR.

      Comment

      • scotoma
        Junior Member
        • Nov 2012
        • 2

        #4
        Thank you both for your help. Unfortunately, RunInfo.xml looks like this:

        <Reads>
        <Read NumCycles="151" Number="1" IsIndexedRead="N" />
        <Read NumCycles="151" Number="2" IsIndexedRead="N" />
        </Reads>
        <FlowcellLayout LaneCount="1" SurfaceCount="1" SwathCount="1" TileCount="12" />


        so it appears MSR parsed the sample sheet incorrectly and the index read was skipped. Regardless, I'll try Number6's strategy of configuring Miseq Reporter to output a separate fastq for index reads (if not for this run, at least for the next). Also, I guess that in the future I'll input the actual index sequences. Is there anything else I should try? Thanks again.

        Comment

        • hoisinjl
          Member
          • Jun 2009
          • 23

          #5
          I am also going to try the "CreateFastqForIndexReads" as that will solve some of the post-analysis work I do to generate the three files.

          When starting a run, after you upload the SampleSheet, the next screen shows how the SampleSheet is parsed. I would add that you should double-check that the # cycles for read1, index and read2 are correctly read.

          Comment

          Latest Articles

          Collapse

          • SEQadmin2
            Beyond CRISPR/Cas9: Understand, Choose, and Use the Right Genome Editing Tool
            by SEQadmin2



            CRISPR/Cas9 sparked the gene editing revolution for both research and therapeutics.1 But this system still showed severe issues that limited its applications. The most prominent were the heavy reliance on PAM sequences, delivery limitations, double-stranded breaks that prompt unintended edits and cell death, and editing inefficiency (both in targeting and in knock-in reliability).

            Despite this, “CRISPR helped turn genome editing from a specialized technique into
            ...
            07-31-2026, 11:01 AM
          • SEQadmin2
            Proteomic Platforms: How to Choose the Right Analytical Strategy to Improve Detection and Clinical Applications
            by SEQadmin2


            Proteomics platforms are evolving rapidly, with advances in mass spectrometry and affinity-based approaches expanding what researchers can detect and at what scale. As the field moves toward deeper proteome coverage and clinical applications, scientists face an increasingly complex landscape of tools. This article will explore how researchers are navigating these choices to find the right platform for their work.

            The systematic characterization of the human proteome has
            ...
            07-20-2026, 11:48 AM

          ad_right_rmr

          Collapse

          News

          Collapse

          Topics Statistics Last Post
          Started by SEQadmin2, 08-13-2026, 12:22 PM
          0 responses
          31 views
          0 reactions
          Last Post SEQadmin2  
          Started by SEQadmin2, 08-11-2026, 10:35 AM
          0 responses
          24 views
          0 reactions
          Last Post SEQadmin2  
          Started by SEQadmin2, 08-06-2026, 07:41 AM
          0 responses
          38 views
          0 reactions
          Last Post SEQadmin2  
          Started by SEQadmin2, 08-03-2026, 10:13 AM
          0 responses
          51 views
          0 reactions
          Last Post SEQadmin2  
          Working...